Test & Measuring Equipment - Statistical studies to calibration determine frequency

R

Ron Dooley

#1
This question is a result of a survey by one of our customers.

The question:

Control of Test & Measuring Equipment.

"Requirement by our customer.
Calibration results to be analyzed using statistical techniques to establish frequency".

I know of no requirement for ISO or QS that has any language concerning statistical studies for calibrations to determine frequency.

Would appreciate any help. Thanks
 
Elsmar Forum Sponsor
D

David Mullins

#2
Good Management!
When you open the book that came with the equipment, what is the manufacturers recommendation on calibration frequency. Determining this frequency yourself can be an expensive and time consuming operation. Additionally, I know in Oz there are already numerous standards that provide this sort of guidance.
At the end of the day, the stats to determine frequency relate to the average time taken to exceed the acceptable degree of risk of the accuracy of the equipment.

------------------
 

Marc

Hunkered Down for the Duration
Staff member
Admin
#3
I don't know that there is a specific requirement for statistical techniques but I would be ready to use cal records as evidence of stability during your cal interval. I would also be ready to say you evaluated the need for statistical analysis of calibration data and found it to not be value added based upon instrument stability. You don't need a graph to tell when the instrument is wearing or such - its stability will fall with use. If heavily used this could be a day or a week. Or it could be a month or a year. Lot's of possibilities.
 
Thread starter Similar threads Forum Replies Date
x-files ISO 9001 Clause 7.6 - Test as Measuring Equipment... ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
W Moving M&TE (Measuring & Test Equipment) to New Facility General Measurement Device and Calibration Topics 3
R ISO 17025 M&TE (Measuring & Test Equipment) Calibration Requirements ISO 17025 related Discussions 4
J Measuring and Test Equipment that must undergo MSA Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 8
N Control of Measuring and test Equipment - Flowchart wanted ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
G ANSI/NCSL Z540.3 (Requirements for the Calibration of Measuring and Test Equipment) General Measurement Device and Calibration Topics 1
K Measurement & Test - What is the definition of devices? Measuring equipment is clear. General Measurement Device and Calibration Topics 1
A Calibration of inspection, measuring and test equipment Inspection, Prints (Drawings), Testing, Sampling and Related Topics 12
A Method of Disposition on Damaged Inspection,Measuring Test Equipment Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
R Test & Measuring Equipment again ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
A Test Probes for measuring ohms per square on shielded plastic surfaces Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
D Interlaboratory Comparisons Test of CMM (coordinate measuring machine) General Measurement Device and Calibration Topics 12
K Is Test Software considered a Monitoring and Measuring Device? General Measurement Device and Calibration Topics 1
E MSA for Push and Pull test Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 0
J Confusing on Harmonic Emissions/Voltage Fluctuations/Flicker test for CISPR 11 Class A Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Routine production tests - dielectric voltage withstand test IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
D Test Method Validation Qualification and Validation (including 21 CFR Part 11) 4
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Accelerated Aging - Creating test samples - Implantable medical device Question Other Medical Device Related Standards 4
E Mentor for Test Method Validation (TMV) Design and Development of Products and Processes 2
L ASQ's Biomedical Auditor Course Test ASQ - American Society for Quality 1
W Regarding Requirement for Hi pot & Leakage Current test Station Set-up IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R Regulatory Affairs Certification (RAC) prep - Practice test versus actual exam Professional Certifications and Degrees 0
R SAT (System Accuracy Test) temperature according to AMS2750E AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
R MSA for ATE (Automatic Test Equipment Embedded Software) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
D Test summary report example for design validation wanted - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
S Documenting Design Verification Test Results (ISO 9001) Design and Development of Products and Processes 1
A Earth leakage current test for Permanently installed Large scale MEE IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
A Bench test report translation to English for 510(k) submission Other US Medical Device Regulations 2
J Production Line Test Brasil - High Voltage Dielectric Strenght Test Other Medical Device Regulations World-Wide 5
W Seeking Guidance Verification Test Strategy for Class B Medical Devices IEC 62304 - Medical Device Software Life Cycle Processes 1
J 510(k) for a control kit for an external IVD test kit 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
K Cord anchorage pull test IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Minitab assistant: 1-Sample % Defective Test Using Minitab Software 1
N Is capability applicable for a destructive test? Capability, Accuracy and Stability - Processes, Machines, etc. 9
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Test method validation - Is MSA (MSA1, MSA2, MSA3 and linearity) a good solution? Medical Device and FDA Regulations and Standards News 1
B Operator protection - When to apply table 7, Dielectric strength test voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Medical Device Airworthiness, DO-160 Test Planner IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J Audit Finding For Not Retaining Test Results ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
W ASTM F1929 dye penetration test - Validation for in-house testing ISO 13485:2016 - Medical Device Quality Management Systems 13
S IVF medical devices standards and test methods Medical Device and FDA Regulations and Standards News 4
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
N Heavy tare to simulate high load in a routine sensitivity test of a floor scale? General Measurement Device and Calibration Topics 3
bryan willemot Looking for a certified test report for aerospace parts AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
A Information about Medical Device Test Laboratories Other Medical Device and Orthopedic Related Topics 4

Similar threads

Top Bottom