Test Method Validation of Vision System
msk92
I work in the pharma/med industry. We recently introduced an automated vision system. I am looking for some resources/templates that could help me draft and perform a test method validation using the vision system. The vision system will have pass/fail criteria so we need to validate the test method to make sure the system is consistently performing and we have a baseline.
Any help would be greatly appreciated.
Take a look at this document under the categorical data section.
Since you are medical systems I would recommend the McNemer approach as it is statistically more appropriate than the agreement or even Kappa approach.
  • MSA V and V.pdf
msk92
Thank you so much for sharing! This is very helpful. Can we use a similar system like CMM or SmartScope to create master samples and then validate/calibrate our vision system using the master samples?