Testing a Low Current (Battery) DC Medical Device

markl368

Involved In Discussions
We're looking into testing a medical device that is operated using a battery (DC) and running at a very low current level; <100 uA. Are there any standards that outline safety and risk considerations and tests to confirm this for such devices??
 

Pads38

Moderator
IEC 60601-1 (plus collaterals and particulars) applies to all electrical medical devices. They may be mains powered, battery powered or both.

There is no distinction made for a low supply current but some parts (thermal risks, fire) would be less relevant due to the limited total energy available from your cell.
 

Marcelo

Inactive Registered Visitor
IEC 60601-1 (plus collaterals and particulars) applies to all electrical medical devices.

Just a quick comment to avoid confusion. IEC 60601 applies to all "medical electrical equipment"as defined by the standard. There's several electrical medical devices which are not medical electrical equipment.

Anyway, even if the device from the OP is not a medical electrical equipment, he could use the requirements of IEC 60601 to show compliance with electrical safety requirements (even more so if there's no standard related to his device).
 
Thread starter Similar threads Forum Replies Date
C Hardness Testing - Low Cpk's using Minitab Misc. Quality Assurance and Business Systems Related Topics 10
M IEC 60601-1 Testing required for low-voltage devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P Hot water cylinder parts - Testing per the Low Voltage Directive CE Marking (Conformité Européene) / CB Scheme 1
J Testing Potential New Hires for a Low Level Part Inspection Position Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
H Regarding 47 CFR 15c testing CE Marking (Conformité Européene) / CB Scheme 0
O ESD Testing IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
Stoic Which medical device companies are using Recombinant Factor C (rFC) instead of Limulus Amebocyte Lysate (LAL) for endotoxin testing? Sustainability, Green Initiatives and Ecology 0
S Testing & Validation Phase Design and Development of Products and Processes 1
I Design Verification Testing on Non-Validated Equipment ISO 13485:2016 - Medical Device Quality Management Systems 2
R Quality Agreements with external testing laboratories now mandatory? EU Medical Device Regulations 16
A New biocompatibility testing? US Medical Device Regulations 1
S Guidance on Factory Acceptance Testing for injection moulds please Qualification and Validation (including 21 CFR Part 11) 2
C Biocompatability Testing - Electric Breast Pumps , Silicone Component EU Medical Device Regulations 3
L Proficiency Testing Statistical Analysis ISO 17025 related Discussions 0
T Changes to product IFU & Summative Testing IEC 62366 - Medical Device Usability Engineering 1
A Chemotherapy Resistance Testing for EU EU Medical Device Regulations 1
N FDA mechanical bench testing Manufacturing and Related Processes 5
E Microbial Barrier Testing for Medical Devices Qualification and Validation (including 21 CFR Part 11) 2
A Safety testing of Medical devices ISO 13485:2016 - Medical Device Quality Management Systems 2
C When should I take endotoxin testing? EU Medical Device Regulations 3
A Safety testing for Medical devices to be sold in India ISO 13485:2016 - Medical Device Quality Management Systems 0
M Test Lab Recommends ANSI C63.27 Wireless Coexistence Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
P Biocompatibility testing ISO-10993 for FDA submission Other Medical Device Related Standards 7
R Tips on transitioning out of testing lab Career and Occupation Discussions 0
T Aircraft GAPP Software Testing Compliance Question AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
E Usability for Near Patient Testing Test Kit Human Factors and Ergonomics in Engineering 2
J Single Fault Testing for Secondary Circuits (60601-1 section 13.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
P Medical device testing lab/service standard Other Medical Device Related Standards 1
T Cytoxtoxicity Testing Dilution Other Medical Device Related Standards 1
M Altitude testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
H IEC 62368-1 Scope of Testing CE Marking (Conformité Européene) / CB Scheme 3
H EMC & Safety Testing of the product with the same Sr. No. for EC Declaration CE Marking (Conformité Européene) / CB Scheme 3
A Non-medical device testing in the medical system IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
H Is testing according to IEC 60601-1 & IEC 60601-1-2 mandatory for a device being registered under MDR? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W CQI-9 Testing Fluids Manufacturing and Related Processes 0
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
T Testing for proving food safety on USP class vi material RoHS, REACH, ELV, IMDS and Restricted Substances 0
J Risk Analysis for Proficiency Testing Reliability Analysis - Predictions, Testing and Standards 1
Rene Minassian Control of (TMME) Testing and Measurement Equipment Oil and Gas Industry Standards and Regulations 2
M Is complete testing required as per ISO 10993 for materials used in orthopedic implants or is literature review route possible Other Medical Device Related Standards 5
B A testing lab within an organization ISO 17025 related Discussions 10
B Package integrity Testing failure ISO 13485:2016 - Medical Device Quality Management Systems 7
R Basic Lab Practices - related to ISO 17025 testing labs ISO 17025 related Discussions 1
N Microbial testing on food packaging materials Food Safety - ISO 22000, HACCP (21 CFR 120) 3
A Stability Testing of a device with sterile fluid pathway claim Other Medical Device and Orthopedic Related Topics 1
M Cosmetic Shelf Life Testing Misc. Quality Assurance and Business Systems Related Topics 10
S Safety testing IEC 60601 series Medical Device and FDA Regulations and Standards News 2
M EE's - Laptops/Printer Change and Emissions Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
keya1tos EU to UK plug adapter correct testing standard CE Marking (Conformité Européene) / CB Scheme 9
D IEC 60601-1-2: Is EMC immunity testing required for a device without essential performance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 25