Testing ME equipment powered by Battery

EMengineer

Involved In Discussions
Hi,

Could someone provide any information how leakage current measurements are performed when ME equipment is battery powered?
Battery voltage is up to 3.7 VDC and there is no possibility to use it on patient when it's in charging mode (because input is the same for charging and as patient input connection).
All measurement diagrams (in standard) are made for ME equipment which is powered by supply mains. Or should I measure everything like it was intended to be powered from supply mains?
Thank you for your answers in advance!

EDIT: found the answer in cl. 8.7.4.2 - ME equipment leakage measurement is the same no matter what the supply is (mains or internal).
 
Last edited:
Thread starter Similar threads Forum Replies Date
I Design Verification Testing on Non-Validated Equipment ISO 13485:2016 - Medical Device Quality Management Systems 2
Rene Minassian Control of (TMME) Testing and Measurement Equipment Oil and Gas Industry Standards and Regulations 2
M Do i need to have equipment validation if 100% testing is completed? Qualification and Validation (including 21 CFR Part 11) 6
E How do you know your test equipment is good? Functional testing and Golden Boards ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
K Is a 3V torch classified as medical electrical equipment needing EMI/EMC testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Testing an equipment for MRI Compatibility Other US Medical Device Regulations 7
E Rich O2 Environment Testing Laboratories (Clause 11.2.2 ME EQUIPMENT and ME SYSTEMS) IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M ISO17025 Testing Laboratory Equipment Calibration Requirements ISO 17025 related Discussions 4
C Is there a standard for testing equipment at a single point? General Measurement Device and Calibration Topics 13
T Is Testing Equipment Re-Validation required if the Test Equipment is Relocated? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
A Minimum Testing Requirements for a Clear Protective Equipment Barrier (Class II 510K) Other US Medical Device Regulations 1
P Testing to clause 15.4.6 - Actuating parts of controls of Medical Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Testing Medical Equipment used in conjunction with Oxygen Rich Environments IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
F Help with Electrical Safety Testing - Medical Equipment Regulations IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
6 Fuel Testing Laboratory Test Equipment - Methods and Calibration General Measurement Device and Calibration Topics 2
K Where do you draw the line on what fixtures/testing equipment require calibration General Measurement Device and Calibration Topics 36
M GR&R (Gage R&R) for Automated Electrical Testing Equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
S Calibration of equipment in an A2LA accredited testing laboratory General Measurement Device and Calibration Topics 7
D Sealed bag testing equipment Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
D Does waste water equipment need calibration if we also outsource testing? Miscellaneous Environmental Standards and EMS Related Discussions 12
I Procedure for Calibration of my Load Cell Testing Equipment - Needed General Measurement Device and Calibration Topics 5
Y Is NDC needed for gage used for ATE? Automated Test Equipment IC Testing Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
D Looking for 100% sort testing equipment for overall length, flatness, and parallelism General Measurement Device and Calibration Topics 1
A Measurement standards - Testing equipment for lubricant characteristics - Grease, Oil General Measurement Device and Calibration Topics 2
H Regarding 47 CFR 15c testing CE Marking (Conformité Européene) / CB Scheme 0
O ESD Testing IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
Stoic Which medical device companies are using Recombinant Factor C (rFC) instead of Limulus Amebocyte Lysate (LAL) for endotoxin testing? Sustainability, Green Initiatives and Ecology 0
S Testing & Validation Phase Design and Development of Products and Processes 1
R Quality Agreements with external testing laboratories now mandatory? EU Medical Device Regulations 16
A New biocompatibility testing? US Medical Device Regulations 1
S Guidance on Factory Acceptance Testing for injection moulds please Qualification and Validation (including 21 CFR Part 11) 2
C Biocompatability Testing - Electric Breast Pumps , Silicone Component EU Medical Device Regulations 3
L Proficiency Testing Statistical Analysis ISO 17025 related Discussions 0
T Changes to product IFU & Summative Testing IEC 62366 - Medical Device Usability Engineering 1
A Chemotherapy Resistance Testing for EU EU Medical Device Regulations 1
N FDA mechanical bench testing Manufacturing and Related Processes 5
E Microbial Barrier Testing for Medical Devices Qualification and Validation (including 21 CFR Part 11) 2
A Safety testing of Medical devices ISO 13485:2016 - Medical Device Quality Management Systems 2
C When should I take endotoxin testing? EU Medical Device Regulations 3
A Safety testing for Medical devices to be sold in India ISO 13485:2016 - Medical Device Quality Management Systems 0
M Test Lab Recommends ANSI C63.27 Wireless Coexistence Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
P Biocompatibility testing ISO-10993 for FDA submission Other Medical Device Related Standards 7
R Tips on transitioning out of testing lab Career and Occupation Discussions 0
T Aircraft GAPP Software Testing Compliance Question AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
E Usability for Near Patient Testing Test Kit Human Factors and Ergonomics in Engineering 2
J Single Fault Testing for Secondary Circuits (60601-1 section 13.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
P Medical device testing lab/service standard Other Medical Device Related Standards 1
T Cytoxtoxicity Testing Dilution Other Medical Device Related Standards 1
M Altitude testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
H IEC 62368-1 Scope of Testing CE Marking (Conformité Européene) / CB Scheme 3

Similar threads

Top Bottom