Testing ME equipment powered by Battery
EMengineer
Could someone provide any information how leakage current measurements are performed when ME equipment is battery powered?
Battery voltage is up to 3.7 VDC and there is no possibility to use it on patient when it's in charging mode (because input is the same for charging and as patient input connection).
All measurement diagrams (in standard) are made for ME equipment which is powered by supply mains. Or should I measure everything like it was intended to be powered from supply mains?
Thank you for your answers in advance!
EDIT: found the answer in cl. 8.7.4.2 - ME equipment leakage measurement is the same no matter what the supply is (mains or internal).
E How do you know your test equipment is good? Functional testing and Golden Boards ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
K Is a 3V torch classified as medical electrical equipment needing EMI/EMC testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Testing an equipment for MRI Compatibility Other US Medical Device Regulations 3
E Rich O2 Environment Testing Laboratories (Clause 11.2.2 ME EQUIPMENT and ME SYSTEMS) IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M ISO17025 Testing Laboratory Equipment Calibration Requirements ISO 17025 related Discussions 4
C Is there a standard for testing equipment at a single point? General Measurement Device and Calibration Topics 13
T Is Testing Equipment Re-Validation required if the Test Equipment is Relocated? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
A Minimum Testing Requirements for a Clear Protective Equipment Barrier (Class II 510K) Other US Medical Device Regulations 1
P Testing to clause 15.4.6 - Actuating parts of controls of Medical Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Testing Medical Equipment used in conjunction with Oxygen Rich Environments IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
F Help with Electrical Safety Testing - Medical Equipment Regulations IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
6 Fuel Testing Laboratory Test Equipment - Methods and Calibration General Measurement Device and Calibration Topics 2
K Where do you draw the line on what fixtures/testing equipment require calibration General Measurement Device and Calibration Topics 36
M GR&R (Gage R&R) for Automated Electrical Testing Equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
S Calibration of equipment in an A2LA accredited testing laboratory General Measurement Device and Calibration Topics 7
D Sealed bag testing equipment Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
D Does waste water equipment need calibration if we also outsource testing? Miscellaneous Environmental Standards and EMS Related Discussions 12
I Procedure for Calibration of my Load Cell Testing Equipment - Needed General Measurement Device and Calibration Topics 5
Y Is NDC needed for gage used for ATE? Automated Test Equipment IC Testing Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
D Looking for 100% sort testing equipment for overall length, flatness, and parallelism General Measurement Device and Calibration Topics 1
A Measurement standards - Testing equipment for lubricant characteristics - Grease, Oil General Measurement Device and Calibration Topics 2
E 60601-1 - Tilt testing - Tensile safety factor IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
V Setup for testing against ISO14708 clause 16 (protection of the patient from herms caused by heat) Other Medical Device Related Standards 0
gramps What do you think about automated QA testing For software app industry? Misc. Quality Assurance and Business Systems Related Topics 5
B In house NIOSH pre Testing accepted by NIOSH? US Food and Drug Administration (FDA) 1
M Bacteriostasis/Fungistasis Testing Other Medical Device and Orthopedic Related Topics 6
P Sample Size for Distribution Simulation Testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
N EN 813, EN 12277, EN 1497 - Testing some harness prototypes to an EN standard Various Other Specifications, Standards, and related Requirements 0
D Essential performance and EMC immunity testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
J Biocompatbility testing on Class 1 device requirements Other US Medical Device Regulations 12
Q Summative Usability Evaluation Testing: prior or during Clinical Investigation? Human Factors and Ergonomics in Engineering 6
B NIOSH Approval for Surgical N95 Respirators - Required testing US Food and Drug Administration (FDA) 2
M ECG lead leakage currents - How to specify ECG leads during electrical safety testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C Stress / Challenge Conditions for Design Verification Testing to Reduce Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
D CFR Title 14: Aeronautics and Space PART 120—DRUG AND ALCOHOL TESTING PROGRAM Federal Aviation Administration (FAA) Standards and Requirements 3
lanley liao Purchase Acceptance Criteria - Tensile testing Oil and Gas Industry Standards and Regulations 2
M Device mounted at IV pole - what about mechanical stability testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J Cochlear Implant Testing 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
A Class I (exempt) testing requirements Other Medical Device Related Standards 0
JoCam Electrical Testing for Japan, PSE or CB Scheme Other Medical Device Regulations World-Wide 0
M Who are the go to companies for non-destructive hardness testing? General Measurement Device and Calibration Topics 3
P Testing cloud-based backups IT (Information Technology) Service Management 7
I IATF Lab Scope Testing Qualification and Competency Documentation IATF 16949 - Automotive Quality Systems Standard 3
N Chemical Testing on Medical Devices - Solutions in a container closure system (bag) EU Medical Device Regulations 1
M Comparing data from destructive testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
T Flammability testing Reliability Analysis - Predictions, Testing and Standards 0
E Manufacturers should develop a testing device for covid19 Service Industry Specific Topics 0