Tests IEC 60601-1 - Grounding Resistance, Leakage Current and Dielectric Rigidity in an automated way
Tiago Alfenas
Good Morning!
We are an electromedical manufacturer and would like to ask you a question, which device do you use to test the 60601 (Grounding Resistance, Leakage Current and Dielectric Rigidity) in an automated way? Example only one model of ours each lot is about 1,000 units, and according to we interpret we have to do 100% of the lot, the tests take about 40 minutes per device, we would like to shorten this time
One example here:
The Rigel 288+ Electrical Safety Analyzer from Rigel Medical
(Disclosure - no connections)
It can be set up with a pre-programmed set of tests and limits, can record results to computer etc. and can be quite quick.
It is not entirely 'hands free'; typically, you would still need to use a probe for earth bond tests.
Also note that it does not do the dielectric part of the tests - you'll need another machine for that.
I believe that other manufacturers do similar equipment - for example Fluke, but I do not have experience of those.
C ABS plastic : which IEC 60601-1 tests to redo? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
N ECoG/EEG electrodes - Required IEC 60601-1 tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
O Is Patient Cable Exempted from EFT Tests? (IEC 60601-1-2) CE Marking (Conformité Européene) / CB Scheme 1
S Clarification regarding tests in IEC 60601-2-25 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
T List of Tests - IEC 61010-1, 3rd edition and IEC 60601-1, 3.1 edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
K Interpretation of IEC 60601-1-2 Electromagnetic Compatibility – Requirements & tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
G Sample size for IEC 60601-1-11 environmental tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P PSU Transformer Tests Approved to IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 60601 - List of required tests - Medical Device Class IB IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
B IEC 61010-1 - Voltage / Dielectric strength tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
W CPU BIST IEC 62304 - Embedded code has CPU instruction tests IEC 62304 - Medical Device Software Life Cycle Processes 2
W Non Sterile Medical Device Environmental Tests Other Medical Device Related Standards 4
S Required tests for Surgical gown US Food and Drug Administration (FDA) 1
M Sterile packing validation tests to be performed and protocol Other Medical Device and Orthopedic Related Topics 1
E Routine production tests - dielectric voltage withstand test IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
Q Old products new class - Dental Devices - Choosing tests EU Medical Device Regulations 2
G Supplier flowdown verification of tests and revisions Supplier Quality Assurance and other Supplier Issues 0
O Is it mandatory for Medical Devices to work properly after dielectric strength tests? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
K Changes in process dispersion, and location hypotheses tests Using Minitab Software 3
M MDR requires the NB to perform "tests in order to check that the quality management system" EU Medical Device Regulations 3
Z EN 868-8 "required tests" - Rigid sterilization containers Other Medical Device Related Standards 0
M Define voltage and frequency to perform tests 61010-1 and 61326-1 for CE certification CE Marking (Conformité Européene) / CB Scheme 4
M Informational US FDA Final Guidance – Coronary, Peripheral, and Neurovascular Guidewires – Performance Tests and Recommended Labeling Medical Device and FDA Regulations and Standards News 0
J GLP accreditation for mechanical tests of medical device? Other Medical Device and Orthopedic Related Topics 1
M Routine testing of medical electrical systems - What specific electrical safety tests should be performed? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
A Touch current in single fault conditions test and earth leakage current in normal conditions test, are they really different tests? IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
O Statistical justification of sampling size in V&V tests ISO 13485:2016 - Medical Device Quality Management Systems 5
John Predmore A Balanced view of statistical tests Statistical Analysis Tools, Techniques and SPC 3
D API 6A Chokes - PR1 / PR2 Tests Oil and Gas Industry Standards and Regulations 12
E 2 participants in inter-laboratory comparison tests ISO 17025 related Discussions 3
R Room requirements for electrical tests in devices ISO 13485:2016 - Medical Device Quality Management Systems 2
M EN60601-1-2 4th edition requirements - Tests for devices produced in two options IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
F Sample Size for Biocompatibility Tests Other Medical Device Related Standards 4
Gamula Biocompatibility GLP conform tests outside China China Medical Device Regulations 1
W Equipment for Oven Calibration, Uniformity and System Accuracy Tests AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
T ISO 17025 Section 5.9.1 - Assuring Quality of Tests - Requirements ISO 17025 related Discussions 4
K VDA 2 Requirements of Tests for PPA VDA Standards - Germany's Automotive Standards 2
K How do you select the right grips for your tensile tests? Imported Legacy Blogs 0
Q Standardized Anodize Coating Tests Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
D Should Biocompatibility Tests be performed on Production Samples? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
R Medical Device Storage Temperature Tests and Requirements EU Medical Device Regulations 14
R Is pre-conditioning supplier component before incoming inspection tests allowed? Quality Manager and Management Related Issues 16
K Mandatory Electrical Safety Tests - Regular Production Line IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
A Non-replicable tests in MSA (Measurement Systems Analysis) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
C Sampling plan for destructive tests on small lots ISO 13485:2016 - Medical Device Quality Management Systems 20
somashekar The Land of Tests ... Book, Video, Blog and Web Site Reviews and Recommendations 2
V Are microbial limit tests required for sterile products? Qualification and Validation (including 21 CFR Part 11) 3
P Humidity Pre-Conditioning before leakage current and dielectric strength tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
J Biocompatibility tests in AAA LAC Laboratory ISO 13485:2016 - Medical Device Quality Management Systems 6
J Plastic Compatibility tests with the contact of mucous membranes Other Medical Device Regulations World-Wide 3
