dear all,
I have a confusion on mensument of patient auxiliary current decribed in 8.7.4.8 and figure E.4
8.7.4.8 Measurement of the PATIENT AUXILIARY CURRENT
ME EQUIPMENT with an APPLIED PART is tested according to Figure 19, using an appropriate
measuring supply circuit unless the ME EQUIPMENT has only a single PATIENT CONNECTION.
The PATIENT AUXILIARY CURRENT is measured between any single PATIENT CONNECTION and all
other PATIENT CONNECTIONS, either connected directly together or loaded as in NORMAL USE
(see also Annex E).
1# opinion is :
the patient auxiliary current is only applicable and testd between different patient connections within the single apppied part with multipel patient conncections
2# opinion is :
the patient auxiliary current is should be testd between different patient connections with different applied parts ?
i think 1# opinion is right. can anybody help me to clarrify this ? thank you .
I have a confusion on mensument of patient auxiliary current decribed in 8.7.4.8 and figure E.4
8.7.4.8 Measurement of the PATIENT AUXILIARY CURRENT
ME EQUIPMENT with an APPLIED PART is tested according to Figure 19, using an appropriate
measuring supply circuit unless the ME EQUIPMENT has only a single PATIENT CONNECTION.
The PATIENT AUXILIARY CURRENT is measured between any single PATIENT CONNECTION and all
other PATIENT CONNECTIONS, either connected directly together or loaded as in NORMAL USE
(see also Annex E).
1# opinion is :
the patient auxiliary current is only applicable and testd between different patient connections within the single apppied part with multipel patient conncections
2# opinion is :
the patient auxiliary current is should be testd between different patient connections with different applied parts ?
i think 1# opinion is right. can anybody help me to clarrify this ? thank you .