As far as I know, there is no direct requirement in IEC 60601-1 that the ME EQUIPMENT has a SECONDARY CIRCUIT. Am I right?
To be more specific let's imagine the flyback converter with D2Dn LED array as a load (see the picture). Capacity of C1 and the charge voltage are quite big, say, 10mF and 1000V. The medical effect is accomplishing by applying the light from D2Dn to a skin or something like that. The device is CLASS I and it has PROTECTIVELY EARTHED ENCLOSURE. Whole the enclosure is an ACCESSIBLE PART. There are clearances and creepages between the inside circuit and the enclosure, which provide 1MOP . There aren't MOPs between different sides of transformer TR1.
Does such ME DEVICE have the right to exist? Would there be any special requirements to the power cord?
As far as I know, there is no direct requirement in IEC 60601-1 that the ME EQUIPMENT has a SECONDARY CIRCUIT. Am I right?
To be more specific let's imagine the flyback converter with D2Dn LED array as a load (see the picture). Capacity of C1 and the charge voltage are quite big, say, 10mF and 1000V. The medical effect is accomplishing by applying the light from D2Dn to a skin or something like that. The device is CLASS I and it has PROTECTIVELY EARTHED ENCLOSURE. Whole the enclosure is an ACCESSIBLE PART. There are clearances and creepages between the inside circuit and the enclosure, which provide 1MOP . There aren't MOPs between different sides of transformer TR1.
Does such ME DEVICE have the right to exist? Would there be any special requirements to the power cord?