Thermo-Hygrometer - Measurement of 10DegC @10%RH
Hi ,
I am using a thermo-hygrometer(with external sensor)in Climatic testing in climatic chamber , where my product is tested at 10DegC with 10%RH , but my thermo-hygrometer is calibrated in 10%RH , 50%RH , 95%RH @25DegC and at temperature values of 5DegC , 25DegC 50DegC without humidity control . Can i use this thermo-hygrometer for measurement of 10DegC @10%RH. or i need to calibrate the instrument at 10DegC 10RH?
dwperron
Your thermo-hygrometer uses separate sensors for temperature and humidity. Because of this, if the humidity sensor was calibrated at 10% the result should be valid over the full temperature range.
But I still stuck in a thought, the humidity is different in different temperatures so how a calibration of sensor at 3 points at one temperature is enough.
dwperron
Typically, the thermistors used in these devices are stable performers and prove to be much more reliable than the humidity sensors.
A 15°C deviation from the calibration point should not be considerable.
M How to read a calibration certificate for thermo hygrometer General Measurement Device and Calibration Topics 2
A Calibration/Verification of thermo-hygrometer in storage areas ISO 13485:2016 - Medical Device Quality Management Systems 9
S Medical Device Labeling - Is there a Thermo Disinfector/Washer Symbol Other Medical Device Related Standards 7
G Definition ASTM D624 - Test Method for Tear Strength of Vulc. Rubber & Thermo. Elastomers Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 1
F Calibratable Thermometer / Hygrometer for Quality Control Room Calibration Frequency (Interval) 1
T Uncertainty in Hygrometer Calibration Certificate General Measurement Device and Calibration Topics 4
D How much Deviation is Acceptable when Calibrating a Hygrometer? General Measurement Device and Calibration Topics 3
N Interpreting Hygrometer Calibration Uncertainty - Temperature Coefficient Measurement Uncertainty (MU) 1
Q Would you Calibrate this Digital Hygrometer / Thermometer? General Measurement Device and Calibration Topics 17
H Hygrometer Calibration - Best way to verify accuracy of a 'reference' Hygrometer? General Measurement Device and Calibration Topics 11
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
I QMS monitoring, measurement, analysis and evaluation requirement - Template ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
F Physical Part Measurement General Measurement Device and Calibration Topics 6
B ISO 6508 and portable hardness measurement instruments General Measurement Device and Calibration Topics 0
Ron Rompen Measurement of residual magnetism General Measurement Device and Calibration Topics 2
M Surgical angle measurement guide device with an application software Medical Device and FDA Regulations and Standards News 1
M Disabling measurement data during fault conditions IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J Most versatile measurement device General Measurement Device and Calibration Topics 2
P Surface scratch measurement General Measurement Device and Calibration Topics 1
hogheavenfarm GDT Flatness measurement question Inspection, Prints (Drawings), Testing, Sampling and Related Topics 10
K Software Validation for Measurement Tools used in Process Validation ISO 13485:2016 - Medical Device Quality Management Systems 2
K Best Measurement Systems Demos in California? General Measurement Device and Calibration Topics 1
K Process Mapping - Inputs/Outputs/Detail Activities/Control points/Measurement? Process Maps, Process Mapping and Turtle Diagrams 4
N Program or application for standard time measurement in industry from video surveillance? Human Factors and Ergonomics in Engineering 1
M Defining frequency of measurement tools callibration Calibration and Metrology Software and Hardware 3
T How to “validate/qualify” a new measurement equipment? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Bev D Verification and Validation of Measurement Systems Misc. Quality Assurance and Business Systems Related Topics 0
F How do you add accuracies for tools with different measurement devices? General Measurement Device and Calibration Topics 1
M Viable count Measurement ISO Standard Other Medical Device Related Standards 1
W Air Quality Measurement Hardware and Software General Measurement Device and Calibration Topics 11
O Dimension Measurement Tool Recommendation General Measurement Device and Calibration Topics 21
Marco Bernardi Plane measurement with a 3D touch machine like a CMM. Calibration and Metrology Software and Hardware 3
Marco Bernardi Diameters measurement with 3D touch machine like a CMM Calibration and Metrology Software and Hardware 4
R Measurement Uncertainties Budget for Thermometer Measurement Uncertainty (MU) 6
O Performance Measurement ISO 9001: 2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
O Monitoring performance - How do I determine performance measurement basis within my organization? Misc. Quality Assurance and Business Systems Related Topics 4
G What information to put on measurement Dimensional Results APQP and PPAP 7
D O Ring capability and measurement - What is the automotive 'norm' for capability studies on O Rings? General Measurement Device and Calibration Topics 3
F Electrical Metrology (Source/Generation and Measurement) General Measurement Device and Calibration Topics 1
F Measurement Audit and ILC for ISO 17025 Clause 7.7.2 - Comparison with results of other laboratories ISO 17025 related Discussions 0