To Exclude Or Not - ISO 13485 7.5.3.2.2 and 8.2.4.2 - Plastic Pellets

NikkiQSM

Quite Involved in Discussions
#1
Hello Cove People,

Hoping I can get some advice on implementing our system from ISO 9001 to 13485 - A little background on our company is that we are a plastics manufacturer mainly focused on medical compounds. We do NOT make medical devices, we just make the plastic pellets that eventually become the medical devices (i.e. catheters)

Now, I understand I can exclude anything in section 7 of the standard as long as I have a good reason.

I was going to exlude clause 7.5.3.2.2 Particular requirements for active implantable medical devices and implantable medical devices This clause basically states that we will have records regarding materials used and environment conditions. And it states that our distributors (truck deliver vendors) are to maintain records of when and where they deliver our product.

Although we technically already have this clause covered, I was going to exclude it mainly because, again, we do NOT make medical devices, just the material that eventually becomes the medical device.

Now I have come across this clause: 8.2.4.2 Particular requirements for active implantable medical devices and implantable medical devices "The organization shall record the indentity of personnel performing any inspection or testing."

Again, we have this covered and would adhere to this requirement. I am just a little confused because I would think I would either have to include or exclude both, not just one. And of course, I don't think that I can exclude anything besides a requirement in section 7.

So should I just include both, even though the title of the clause states implantable devices, even though we do NOT make devices?

Thanks for any help!!!

Nikki
 
Elsmar Forum Sponsor
T

treesei

#2
Re: To Exclude Or Not... That is the ?

By theory, a material manufacturer needs not include these clauses. However, if it has customers that use its material in such special devices, the customers may ask the supplier for some extra support.:2cents:
 
#3
Re: To Exclude Or Not - ISO 13485 7.5.3.2.2 and 8.2.4.2

I would suggest that if your materials are used to create implantable products then you should have full batch traceability. So for a particular batch of plastic pellets you would keep records of the constituent chemicals (manufacturer, type, batch, expiry dates etc) plus processing conditions, plus records of who has "signed of" the record of this information.

So, in effect, you are complying with those clauses.
 

Marcelo

Inactive Registered Visitor
#4
The main poblem regarding the application of ISO 13485 in these cases is: ISO 13485 was created to be applied to the manufacturer (meaning, the company responsible) for the medical device. It does not use these terms because it tried to avoid problems in harmonization due to different legal terms.

With this in mind, you can exclude anything you want and do everything you want.

If you are under some kind of cettification sheme which requires you to comply wwith ISO 13485, then things is a little diofferent.


Anyway, just to make some points clear:

Now, I understand I can exclude anything in section 7 of the standard as long as I have a good reason.

No you can't, not i ISO 13485. The way to read ISo 13485 is:

If regulatory requirements (applicable to the "manufacturer", that was the idea) permit, you can exclude design controlst - 7.3 - from the system. This was done to align with the FDA QSR, which does not require design controls to class I devices.

You cannot exclude anything besides 7.3 and again, just if applicable regulatory requirements prmit it, to claim compliance with ISO 13485.

There's another case, which is non-applicability of a requirement. Again, this is only for Cluse 7, and this is related mainly to things like the "active implantable" and "sterile" requirements, and also some process requirements such as service.

In these case, meaning, depending on the nature of the devie, you can deem the requirement non applicable. For example, if the device isn't sterile ou implantable, you can eem the related clauses non-applicable. If service is not a requirement, you can deem the clause non-applicable too.
 

somashekar

Staff member
Super Moderator
#5
we just make the plastic pellets that eventually become the medical devices (i.e. catheters)
Your products are far away from being any medical device. Application of the ISO13485 by you falls very close to what the ISO9001 requires + some specific requirements of ISO13485.
Please do nothing about what ISO13485 specifically calls for certain types of medical devices. You have a very natural justification of non applicability of all such clauses.
Keep it simple and good luck .....
 
R

Roland Cooke

#6
1) No

2) Very no


1) You can only EXCLUDE design control, and then only if applicable regulations permit you to do that. Now you can argue that a clause is NON-APPLICABLE, provided you give a robust justification. In this case, that shouldn't be an issue.

2) You can only deem sub-sections of Clause 7 (Product Realisation) non-applicable. So 8.2.4.2 will apply to you.
However since you don't make implants you won't ever need to record the identity of anyone involved in inspection or testing. However doing that anyway might not be a bad idea for your organization (you decide).
 
C

cartman1417

#7
Since I am new to the Quality Regulation, I have question about how to exclude stuff from the quality manual?

For example, on similar lines to the topic of this thread, we are a medical device company and we are going to manufacture Class IIa diagnostic device for vaginal examination.

So the clauses about implantable devices do not apply to our company.

So my question is, how to I exclude stuff from the quality manual?

Do I just delete the sections in 7 and 8 that do not apply? Or should I keep the headings and just write "Not Applicable" in the paragraph as content?

If I delete section 8.2.4.2, should I give justification at the beginning in the exclusion section?
 

NikkiQSM

Quite Involved in Discussions
#8
This is only my opinion, but its worked for me:

Note any exclusions, with reason for exclusion, in the Scope of your Quality Manual.

I would still leave the heading of the clause you are excluding in the correct spot, in the manual, and write - "Excluded - See Scope" under it.

Again, just what has always worked for me.
 

somashekar

Staff member
Super Moderator
#9
So the clauses about implantable devices do not apply to our company.

So my question is, how to I exclude stuff from the quality manual?
You exclude stuff by stating in the quality manual under a separate table which you may call "Non-applicability and justification"
Quote the ISO13485:2003 clause number and its heading text and state the same very reason why you say diagnostic device for vaginal examination are not applicable to the requirements as said in the stated clause text.
Since it is your QM aligning to the ISO 13485:2003 standard, be specific and clear about your justification concerning to your medical device.
When you justify for below....
7.5.3.2.2 Particular requirements for active implantable medical devices and implantable medical devices
The same stands for this also...
8.2.4.2 Particular requirement for active implantable medical devices and implantable medical devices
 
Last edited:

RCW

Quite Involved in Discussions
#10
There's another case, which is non-applicability of a requirement. Again, this is only for Cluse 7, and this is related mainly to things like the "active implantable" and "sterile" requirements, and also some process requirements such as service.
2) You can only deem sub-sections of Clause 7 (Product Realisation) non-applicable. So 8.2.4.2 will apply to you.
However since you don't make implants you won't ever need to record the identity of anyone involved in inspection or testing. However doing that anyway might not be a bad idea for your organization (you decide).
Okay, so my company does not manufacture active implantable medical devices. I can claim that 7.5.3.2.2 is non-applicable but I can't claim 8.2.4.2 because that is not in Clause 7.
Therefore I need to make the provision to record the identity of personnel performing any inspection or testing of active implantable medical devices even though my company does not produce them.

Is that as stupid as it sounds? How do I explain how I do something when I don't have a requirement to do it other than one sentence out of a specification?!
 
Last edited:
Thread starter Similar threads Forum Replies Date
qualprod ISO 9001:2015 Clause 8.3 - Is it possible to exclude one product but include it in other? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
S Can we exclude our subsidiary company from our ISO 9001 certification? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
A ISO 9001 Clause 7.5.5 Preservation of Product - Apply or exclude? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
B Can we neglect (exclude) the ISO 9001 Calibration clause? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
D Can I exclude Customer Property? ISO 9001 Clause 7.5.4 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 36
G Can a testing laboratory exclude section 7.3 of ISO 9001? Quality Management System (QMS) Manuals 2
L Can you exclude ISO 9001 Clause 7.3 completely from QMS? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
Casana Auditor changing ISO 9001 AUDIT scope midstream - Cannot exclude all of 7.3 anymore General Auditing Discussions 26
Q In ISO 9001:2000 Can We Exclude Some Customers and Products from our registration? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 25
A What To Clauses Exclude - AS9100D AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
a_bardi Is it possible to exclude internal audit from qms scope? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
J Not Design Responsible - Can we exclude all or parts of AS9100 clause 8.3 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
D Can we exclude purchasing and supplier evaluation - ISO9001:2008 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
I Can we exclude 7.3 Clause of TS 16949 Design and Development of Products and Processes 5
S PFMEA Assumption to Exclude Latent Failures FMEA and Control Plans 6
Q Can we exclude design if we do not physically design our product? 7.3 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
A Developing Technology and Equipment to Manufacture Clients Product - Can exclude 7.3? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
Fender1 Management Wants to Exclude a Customer Product Line from Registration Scope ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
E Is it Possible to Exclude Top Management (CEO, Chairman) in the QMS? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
H Planning of Product Realization - AS9100 Clause 7.1 - Can we exclude 7.1f? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
S Design Review: Can I exclude Custom Products ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
L Can this Contract Manufacturer exclude Clause 7.3? Design and Development ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
P Which aspects of section 7 can I justifiably exclude? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
B Can I exclude 7.5.1 & 7.5.2 (Control of Production & Validation) from our QMS? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
J Can I exclude Service even though we do some service? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
N Scope of Certification - Can we exclude Distribution? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
Z Can we Exclude the Validation of Process Clause? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
N Contract Review Question - China manufacturing site to exclude 7.2.2? Contract Review Process 6
L Audit Nonconformance - NCR because we exclude Design and Development ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
K Can we Exclude Design and Development clause in ?Industrial Design? Company? Design and Development of Products and Processes 12
J New TL 9000 Released - Can Exclude more than Clause 7 Various Other Specifications, Standards, and related Requirements 0
I AS9100 - Would like to know if I can exclude the design Design and Development of Products and Processes 6
D Design of Services - Can a Heat Treating service exclude parts of 7.3? Design and Development of Products and Processes 7
J Can We Exclude ISO9000:2000 Clause 7.3.7 (Control Of Design And Development Changes)? Design and Development of Products and Processes 7
C Service - Must we exclude in the Scope? Service Industry Specific Topics 5
B My company makes product to order - Are we design responsible? Exclude 7.3? Design and Development of Products and Processes 17
S Need ISO 15189:2012 Documentation toolkit. Document Control Systems, Procedures, Forms and Templates 0
chris1price Archiving of paper records - ISO 9001 7.5.3.1b Records and Data - Quality, Legal and Other Evidence 4
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
D Common practices in ISO 9001 deployment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
Q ISO 9001-2015 Internal audit finding Internal Auditing 12
B ISO 17025:2017 risk management Risk Management Principles and Generic Guidelines 0
P Audit check for IT company (ISO 9001) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Label Making & Printing Standards ISO / ASTM ISO 13485:2016 - Medical Device Quality Management Systems 5
Sidney Vianna Interesting Discussion Should ISO 9004 be changed from a guidance standard to a requirements standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
Ed Panek ISO 13485:2016 Section 5.5.3 ISO 13485:2016 - Medical Device Quality Management Systems 3
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
ebrahim QMS as per ISO 13485, Clause 4.2 Requirements for regulatory purposes for Medical Devices Authorized Representatives. ISO 13485:2016 - Medical Device Quality Management Systems 3
S ISO 2768-mk print call out Other ISO and International Standards and European Regulations 11
T ISO 17024, clauses 4.3.8. and 5.1.1. Other ISO and International Standards and European Regulations 4