Alright I was shown the basics on how to perform certain in-house calibrations/verifications when I took over my job...

We measure/verify our pin gage set (rarely used) in-house with a Digital Micrometer (flat/flat) 0.00005" resolution.

What SHOULD the tolerance be for shop grade plus size pins (generic brand)?

I'm getting numbers similar to last year's results (previous QA mgr verified them), but I don't know what the tolerance SHOULD be. None have been over 0.00050 beyond their stated sizes yet...

We have two sets:
.061 - .250
.251 - .500

We/I currently verify them annually, is anyone waiting longer?
 
Elsmar Forum Sponsor

Mike S.

Happy to be Alive
Trusted Information Resource
#2
Bob_M said:
Alright I was shown the basics on how to perform certain in-house calibrations/verifications when I took over my job...

We measure/verify our pin gage set (rarely used) in-house with a Digital Micrometer (flat/flat) 0.00005" resolution.

What SHOULD the tolerance be for shop grade plus size pins (generic brand)?

I'm getting numbers similar to last year's results (previous QA mgr verified them), but I don't know what the tolerance SHOULD be. None have been over 0.00050 beyond their stated sizes yet...

We have two sets:
.061 - .250
.251 - .500

We/I currently verify them annually, is anyone waiting longer?
I'm not an expert on this, but I have never had a set of + gage pins that were more than + .0002". If they are .0005" over I'd be suspect of them -- maybe there is rust or contamination on them?

How specific does your measurement need to be ?

Cal interval should take into consideration their usage level, previous cal history, and mfr. recommendations. For pins we use often we do them every 3 months, of less often used pins we go annually, but you gotta watch for rust/corrosion.

We have some really good cal experts on the Cove who may be able to answer better.
 
J

Jimmy Olson

#3
Most likely the gages you have are class ZZ, in which case the tolerance is 0.0002. If you are running into gages that are 0.0005 I would recomend cleaning them and remeasuring.

If 0.0005 is acceptable for what you are using them for then you could special cal them at that tolerance as long as you state that on the cert.
 
B

Bob_M

#4
Richard Olson said:
Most likely the gages you have are class ZZ, in which case the tolerance is 0.0002. If you are running into gages that are 0.0005 I would recomend cleaning them and remeasuring.

If 0.0005 is acceptable for what you are using them for then you could special cal them at that tolerance as long as you state that on the cert.
At the moment we're rarely using them (that I know of). I think our die maintainance uses them more often than production.

We have no current products with tolerances beyond 3 places, and the majority of our "manufactured" holes are in thin sheet metal at right around the nominal sizes. Larger holes, even using a caliper are within spec for all current parts. (Smaller Punched holes for our parts are not typcially critical dimensions).

I think we just have the pin as "just in case" or to help verify a hole size during setup...
----------
Most of what i'm reading is between +.00000 and +.00025 with a few as high as .00040.
At the moment its not a major concern...
Should it be? Or should I spend more time on the pins when we actually start using them on a regular basis?
----------
 

Mike S.

Happy to be Alive
Trusted Information Resource
#5
If a pin .0004" oversize is still okay for the accuracy you need, that's fine, just state this in the cal paperwork.

You might also consider, if you rarely use them, to save you lots of work, make them "calibrate before each use". Teach the users how to calibrate them when they use them. If the users will follow instructions, this can make your cal workload much lighter so you can concentrate on value-added work.
 
B

Bob_M

#6
Mike S. said:
If a pin .0004" oversize is still okay for the accuracy you need, that's fine, just state this in the cal paperwork.

You might also consider, if you rarely use them, to save you lots of work, make them "calibrate before each use". Teach the users how to calibrate them when they use them. If the users will follow instructions, this can make your cal workload much lighter so you can concentrate on value-added work.
Like I said we barely use the pins and our tolerances are not very tight, so I assume .0004 is OK for NOW.

The calibrate before use not going to happen right now.
 

CarolX

Super Moderator
Super Moderator
#7
same business

Bob,
You and I are in the same business. I have a full set of gages and I don't calibrate them in the classic sense, we put a micrometer with 4 place resolution on them before each use. We, too, do not use them very frequently, we use our calipers more often. They do get a bit of use when we run a part with pressed in hardware, since those tolerances are usually +.003/.000 (and can give you a real headache if you are out of tolerance...LOL).

Just another way to skin the cat....in other words....if they are not used...why check them????

Can they be marked "for reference only" and removed from the system?

CarolX
 
B

Bob_M

#8
CarolX said:
Bob,
You and I are in the same business. I have a full set of gages and I don't calibrate them in the classic sense, we put a micrometer with 4 place resolution on them before each use. We, too, do not use them very frequently, we use our calipers more often. They do get a bit of use when we run a part with pressed in hardware, since those tolerances are usually +.003/.000 (and can give you a real headache if you are out of tolerance...LOL).

Just another way to skin the cat....in other words....if they are not used...why check them????

Can they be marked "for reference only" and removed from the system?

CarolX
At the moment I'm keeping them in the calibration loop so they can be used without on the spot calibration. Switching to operator-based calibration is not practical at the moment. At times I barely know how to calibrate certain things. I'm not about to try to teach others how to calibrate. The only exception might be the tool maintenance guy.

Since its only a few hours per year "wasted" its not a big deal yet... I was more worried about what the tolerance should be. *shrug*

So much to learn, too little time, too little money...
 
R

Ryan Wilde

#9
Hi all,

Calibrating little things like these is what I used to do for a living (now I do electronics more), so I'll chime in.

The tolerance of Class ZZ Plus pin gages is -0 / + 0.0002". You cannot adequately prove that tolerance with a micrometer, but then again, with your stated use, you really shouldn't need to hold that tolerance.

The basic tolerance that you CAN adequately prove is -0.0001 / +0.0003. To prove that tolerance, you will have to hold firm on -0 / +0.0002 on your measurement tolerance. The reason being is that 0.00005" resolution micrometers have a possible error of 0.0001" (0.00005" linearity error + 1 count quantizing error).

I'd go with a wider tolerance for my pins if I were you. It's safe, and it makes sense for your use.


Ryan
 
B

Bob_M

#10
Ryan Wilde said:
Hi all,

Calibrating little things like these is what I used to do for a living (now I do electronics more), so I'll chime in.

The tolerance of Class ZZ Plus pin gages is -0 / + 0.0002". You cannot adequately prove that tolerance with a micrometer, but then again, with your stated use, you really shouldn't need to hold that tolerance.

The basic tolerance that you CAN adequately prove is -0.0001 / +0.0003. To prove that tolerance, you will have to hold firm on -0 / +0.0002 on your measurement tolerance. The reason being is that 0.00005" resolution micrometers have a possible error of 0.0001" (0.00005" linearity error + 1 count quantizing error).

I'd go with a wider tolerance for my pins if I were you. It's safe, and it makes sense for your use.

Ryan
Thanks for the very technical answer :D
I've already finished check all of them. The highest I got was +.00040
using our .00005 resolution digital mic.
At the moment that is not a big concern.

Althought people have assumed Class ZZ (which at the moment is Greek to me), but looking at the set, they look very generic, no brand name, no serial number, just a wooden box and numbered + pins...
I have no idea what quality other than
"shop grade +" they really are...

Thanks for the info.
 
Thread starter Similar threads Forum Replies Date
N Drawing tolerance vs. Measurement device General Measurement Device and Calibration Topics 4
I Control Plan (Product/Process specification/ Tolerance) acceptance FMEA and Control Plans 27
J Mechanical inspection techniques of close tolerance parts Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
Proud Liberal Small radius with tight tolerance on a torus General Measurement Device and Calibration Topics 4
L GRR for a tolerance that has changed Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
T How to justify this High %Tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
D Tolerance definition based on expected Cp/cpk Reliability Analysis - Predictions, Testing and Standards 14
N % Tolerance - Type 1 study on the gages, then a gage R&R (ANOVA) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
S Tolerance limits for micro balance General Measurement Device and Calibration Topics 1
J Determining SPC tolerance Statistical Analysis Tools, Techniques and SPC 21
Q % Study variation low, % tolerance high - GR&R Interpretation help Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
R Uncertainty in measurement larger than tolerance Measurement Uncertainty (MU) 2
WEAVER Electronic Weighing Scale Calibration Tolerance Manufacturing and Related Processes 1
I Determining Calibration Tolerance of a Measurement Device General Measurement Device and Calibration Topics 2
Ronen E Tolerance intervals (?) question - Flow Rate Probability Range Statistical Analysis Tools, Techniques and SPC 6
D Tolerance when verifying an analytical balance General Measurement Device and Calibration Topics 3
E Discussion between co-worker on tolerance and uncertainty and how to apply it. Thoughts? 17025 ISO 17025 related Discussions 1
C Thread Tolerance - Can I use a 3A Ring on a part that the drawing calls out for a 2A? Manufacturing and Related Processes 5
T GRR based on part tolerance or process variation. Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
A Excel OOT (Out of Tolerance) Formula Excel .xls Spreadsheet Templates and Tools 3
N Tolerance Interval plots in Minitab Statistical Analysis Tools, Techniques and SPC 0
S Question about a basic additive variance/tolerance example Quality Tools, Improvement and Analysis 2
G Assigning a calibration tolerance - An x-y coordinate machine - Uncertainty as my verification tolerance General Measurement Device and Calibration Topics 4
R Feeler Gauge Calibration - What tolerance is used for calibrating feeler gauges? General Measurement Device and Calibration Topics 4
S SPC (Statistical Process Control) for Unilateral Tolerance - Questions Statistical Analysis Tools, Techniques and SPC 6
B Gage R&R Acceptable (10-30%), deduct Total Variation from Tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
Y How to fix weight tolerance for plastic injection molded part? Does it vary with material groups? Manufacturing and Related Processes 1
F Position tolerance check gage manufacturing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
M GD&T tolerance or band? Using a symbol like parallelism // Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
J What is the difference between Process Variation and Tolerance? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
J Gage R&R without tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
A Calibration of Pressure Gauge, How to Determine Calibration Tolerance General Measurement Device and Calibration Topics 1
V Using K-Factor(Tolerance Interval) Analysis for Design Verification Statistical Analysis Tools, Techniques and SPC 3
C Acceptable NDC for %GR&R part inspection to Tolerance (%Tolerance) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
O Cpk for Unilateral Tolerance Capability, Accuracy and Stability - Processes, Machines, etc. 7
S Doubts about SPC taken in Machining - Part has +-0.01 Tolerance Statistical Analysis Tools, Techniques and SPC 5
A Calibration tolerance for data logger General Measurement Device and Calibration Topics 3
N Straightness Gage Tolerance General Measurement Device and Calibration Topics 0
G Calculating Tolerance of a Panel Meter with a 2 degrees of an Arc General Measurement Device and Calibration Topics 0
Hami812 Cpk Setting Tolerance - Cart before the Horse? (Wifi Routers) Capability, Accuracy and Stability - Processes, Machines, etc. 3
kedarg6500 Excel spreadsheet for Unilateral Tolerance Study wanted Excel .xls Spreadsheet Templates and Tools 2
U Control Charting a Feature with Bonus Tolerance Quality Tools, Improvement and Analysis 10
J Process Capability - How much percentage of tolerance? VDA Standards - Germany's Automotive Standards 3
B Gage R&R MSA - Variation, Tolerance, Variance by different methods IATF 16949 - Automotive Quality Systems Standard 23
R Fixing of tolerance limit - Electronic weighing balance with the capacity of 150 Kgs General Measurement Device and Calibration Topics 5
bobdoering Unilateral Tolerance Capability Calculations Capability, Accuracy and Stability - Processes, Machines, etc. 1
bobdoering Unilateral Tolerance Capability Calculation Imported Legacy Blogs 0
H Tolerance, decimal places, and measuring equipment capability Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
G Out of Dimensional Tolerance - MRB Audit Issue Nonconformance and Corrective Action 11
P Profile Tolerance and CPK Manufacturing and Related Processes 3

Similar threads

Top Bottom