Hi,
We have a new product being evaluated and this clause has come up in discussions with our NB.
The product is powered from the mains by an off-the-shelf 60601-compliant desktop PSU; the 24V DC output is through a shielded cable and a metal-bodied plug.
The shield and plug body are connected to PE within the PSU. The PSU is Class II, so the connection to PE is purely for EMC reasons and is classed as a functional earth.
The product is in a metal housing. The PSU DC output is floating with respect to PE, but the negative terminal is connected to the housing at the DC inlet.
The 24V DC is distributed throughout the product and several accessories, all connected using the same type of metal-bodied plug. The insulation between the 24V and the housing (PCBs, connector internals etc) is classed as 1 MOOP.
When the touch current is measured in normal conditions, as the housing has a connection to PE, the current is close to zero. Similarly, when the touch current is measured in SFC with the PE connection broken, we get a small AC current from the primary-secondary capacitance, well below the 500 uA limit. So far, so good.
The NB then point us to clause 8.1, which states
" NORMAL CONDITION includes all of the following simultaneously:
[...]
- short circuit of any or all insulation that does not comply with the requirements of 8.8;
– short circuit of any or all CREEPAGE DISTANCES or AIR CLEARANCES that do not comply with
the requirements of 8.9;
– open circuit of any or all earth connections that do not comply with the requirements of
8.6, including any functional earth connection."
So the NB are arguing that the touch current measurement should be done with the 24V shorted to the housing. Fine, I said, in this case the PSU output will be short-circuited and it will just current-limit for the duration of the test. I tried this myself in the lab and got a DC touch current of a few tens of microamps. Similarly, if the functional earth connection in the PSU output cable is broken, there is no return path for the touch current.
The NB are now arguing that 8.1 requires them to find the least favourable combination of shorts / opens in all the functional earth, DC + and DC - connections throughout the system, and measure the touch current in that configuration. Which, with the resulting 24V DC between the housing and earth and a 1K measuring device impedance, it will fail.
We have several similar products, going back many years, all using the same power topology, that have all passed evaluation by the same NB, so this is a bit of a new one on me. They are refusing to be drawn on the previous product evaluations, not surprisingly.
Another similar issue relates to 8.4.2(c) which states
"The limits specified in b) above do not apply to the following parts if the probability of a
connection to a PATIENT, either directly or through the body of the OPERATOR, through which
a current exceeding the allowable TOUCH CURRENT could flow, is negligible in NORMAL USE,
and the instructions for use instruct the OPERATOR not to touch the relevant part and the
PATIENT simultaneously:
– accessible contacts of connectors
[...]
For such parts, the voltage to earth or to other accessible parts shall not exceed 42,4 V
peak a.c. or 60 V d.c. in normal condition or in single fault condition. The d.c. limit of
60 V applies to d.c. with not more than 10 % peak-to-peak ripple. If the ripple exceeds that
amount, the 42,4 V peak limit applies. The energy shall not exceed 240 VA for longer than
60 s"
We do have such a warning in the IFU, as the probability of the user disconnecting the DC plug, poking their finger into it to touch the +ve pin, and simultaneously touching the patient, is indeed considered negligible. But the NB then point us to the 240VA limit. The PSU has a 250W rating, and the max continuous power before it goes into limiting will be somewhat higher than this.
Comments welcome.
We have a new product being evaluated and this clause has come up in discussions with our NB.
The product is powered from the mains by an off-the-shelf 60601-compliant desktop PSU; the 24V DC output is through a shielded cable and a metal-bodied plug.
The shield and plug body are connected to PE within the PSU. The PSU is Class II, so the connection to PE is purely for EMC reasons and is classed as a functional earth.
The product is in a metal housing. The PSU DC output is floating with respect to PE, but the negative terminal is connected to the housing at the DC inlet.
The 24V DC is distributed throughout the product and several accessories, all connected using the same type of metal-bodied plug. The insulation between the 24V and the housing (PCBs, connector internals etc) is classed as 1 MOOP.
When the touch current is measured in normal conditions, as the housing has a connection to PE, the current is close to zero. Similarly, when the touch current is measured in SFC with the PE connection broken, we get a small AC current from the primary-secondary capacitance, well below the 500 uA limit. So far, so good.
The NB then point us to clause 8.1, which states
" NORMAL CONDITION includes all of the following simultaneously:
[...]
- short circuit of any or all insulation that does not comply with the requirements of 8.8;
– short circuit of any or all CREEPAGE DISTANCES or AIR CLEARANCES that do not comply with
the requirements of 8.9;
– open circuit of any or all earth connections that do not comply with the requirements of
8.6, including any functional earth connection."
So the NB are arguing that the touch current measurement should be done with the 24V shorted to the housing. Fine, I said, in this case the PSU output will be short-circuited and it will just current-limit for the duration of the test. I tried this myself in the lab and got a DC touch current of a few tens of microamps. Similarly, if the functional earth connection in the PSU output cable is broken, there is no return path for the touch current.
The NB are now arguing that 8.1 requires them to find the least favourable combination of shorts / opens in all the functional earth, DC + and DC - connections throughout the system, and measure the touch current in that configuration. Which, with the resulting 24V DC between the housing and earth and a 1K measuring device impedance, it will fail.
We have several similar products, going back many years, all using the same power topology, that have all passed evaluation by the same NB, so this is a bit of a new one on me. They are refusing to be drawn on the previous product evaluations, not surprisingly.
Another similar issue relates to 8.4.2(c) which states
"The limits specified in b) above do not apply to the following parts if the probability of a
connection to a PATIENT, either directly or through the body of the OPERATOR, through which
a current exceeding the allowable TOUCH CURRENT could flow, is negligible in NORMAL USE,
and the instructions for use instruct the OPERATOR not to touch the relevant part and the
PATIENT simultaneously:
– accessible contacts of connectors
[...]
For such parts, the voltage to earth or to other accessible parts shall not exceed 42,4 V
peak a.c. or 60 V d.c. in normal condition or in single fault condition. The d.c. limit of
60 V applies to d.c. with not more than 10 % peak-to-peak ripple. If the ripple exceeds that
amount, the 42,4 V peak limit applies. The energy shall not exceed 240 VA for longer than
60 s"
We do have such a warning in the IFU, as the probability of the user disconnecting the DC plug, poking their finger into it to touch the +ve pin, and simultaneously touching the patient, is indeed considered negligible. But the NB then point us to the 240VA limit. The PSU has a 250W rating, and the max continuous power before it goes into limiting will be somewhat higher than this.
Comments welcome.