Touch Current for Exposed Charging Contacts on Bodyworn Medical Device
#1
We have a body worn device in development that will use a 4V rechargeable battery, and will recharge in a cradle through surface charging contacts.
In normal use the charging contacts (on the device, connected to the internal battery) will likely make contact with the user's skin. I know that clause 8.4.2(c) is typically used in cases with removable batteries or other exposed contacts that can be avoided and cautioned against, but in this case it would be hard to argue that.
Either we recess the contacts such that they are no longer accessible, or we design some method of protection that would not allow the 100/500uA in NC/SFC. This that correct?
Is there some other way we can approach this?
Elsmar Forum Sponsor
H
hcrook
#2
Is the body-worn device intended to be charged while the patient/operator is wearing it?
You're correct about make the contacts inaccessible or adding a MOP.
#3
Thanks HC.
The device will be removed from the body and then charged in a cradle.
The cradle will store energy in a lithium ion battery of its own (allowing recharging away from a line voltage source) and will need to be recharged via line voltage (using a wall adapter and usb cable).
Thread starter Similar threads Forum Replies Date
A Touch current in single fault conditions test and earth leakage current in normal conditions test, are they really different tests? IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
A IEC 60601-1 SIP/SIO Isolation and Touch current requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-2-49 - Touch Current (Enclosure Leakage Current) requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W Touch Current Measurement and Limits of 42.4V peak or 60Vdc IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
R Should we scrape off the enamel when measure the touch current? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Are Adapter Plugs Tested for Touch Current IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W 'Touch Current' test - Seeking clarity about clause 8.7.4.6 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
B IEC 61010, Touch currents related to burns at higher frequencies (cl. 6.3, A.3) Other Medical Device Related Standards 0
B Accessible parts per 8.4.2 b) - Touch currents in combination with MOOP or MOPP? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Marco Bernardi Plane measurement with a 3D touch machine like a CMM. Calibration and Metrology Software and Hardware 3
Marco Bernardi Diameters measurement with 3D touch machine like a CMM Calibration and Metrology Software and Hardware 4
A IEC 60601-1 Impact test on glass touch controller IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Touch Protected Plugs used for Peripheral Nerve Stimulation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Hershal Now for a touch of lunacy Coffee Break and Water Cooler Discussions 0
R ESD issue with capacitive touch screen IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
M Zeiss U-Soft versus Uni-Touch CMM Software Calibration and Metrology Software and Hardware 7
Wes Bucey Touch Pad on Laptop - Make Me Smart After Work and Weekend Discussion Topics 6
M California Proposition 65 Compliant (no more than 600 ppm touch area) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J Power switch requirement and on/off touch panel marking IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Marc An iMac "Touch" in the future? After Work and Weekend Discussion Topics 2
R Starrett AV612 (combo vision and touch probe) Galileo Instrument Software Validation General Measurement Device and Calibration Topics 3
T Touch Screen Capable Software suggestions for SPC (Statistical Process Control) Quality Assurance and Compliance Software Tools and Solutions 4
9 Another Touch on Management Commitment - Open Corrective Actions Quality Manager and Management Related Issues 6
Sidney Vianna Microsoft Surface techno gadget - Multi-Touch Table After Work and Weekend Discussion Topics 3
C APQP Craftsmanship requirements - What customers see, touch, hear, smell, & how used APQP and PPAP 5
F Operator Instructions and Lot Control - Touch Screens? Document Control Systems, Procedures, Forms and Templates 1
W FMEA - Current control and occurrence rating FMEA and Control Plans 3
W Regarding Requirement for Hi pot & Leakage Current test Station Set-up IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
S Forced ServiceNow validation - No change in our current user and functional requirements IT (Information Technology) Service Management 4
W Need for current design or process control FMEA and Control Plans 2
A Earth leakage current test for Permanently installed Large scale MEE IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
E Best way to keep up with pharmaceutical quality knowledge and current affairs Service Industry Specific Topics 1
T Iec 60601 Impedance and current-carrying capability IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Informational TGA – Current status of breast implant products in Australia Medical Device and FDA Regulations and Standards News 0
F Proficiency Testing for ISO 17025: AC Current Ranges ISO 17025 related Discussions 1
T Tests IEC 60601-1 - Grounding Resistance, Leakage Current and Dielectric Rigidity in an automated way IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Calibration - Isolated current meter - Measuring patient auxiliary leakage current Calibration and Metrology Software and Hardware 4
Q Raising A CAPA on Current CAPA system US Food and Drug Administration (FDA) 3
A SPC for Eddy Current Testing Statistical Analysis Tools, Techniques and SPC 28
A Adding Remote Site to Current IATF 16949 Certification IATF 16949 - Automotive Quality Systems Standard 2
S Leakage Current test and Dielectric strength for Battery operated ECG device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M Leakage current measurement with non-frequency-weighted device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A QMS - Medical Device - New company under a current one ISO 13485:2016 - Medical Device Quality Management Systems 3
E IEC 60335-1 leakage current measurement Other ISO and International Standards and European Regulations 3
K Questions on Technical File to show it is current EU Medical Device Regulations 5
H IS there any database for list of CE marked Medical devices and there current status EU Medical Device Regulations 8
B Mains fuses and OVER-CURRENT RELEASES IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P Adding an additional distributor to a current Mexico Medical Device Registration Other Medical Device Regulations World-Wide 0
M IEC 60601-1 and Active Current Protection Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M Testing a Low Current (Battery) DC Medical Device Other Medical Device Related Standards 4
Similar threads
Top Bottom