I have created this thread in response to a suggestion by Jane_Ackerman. Thanks for the idea, as this is a great topic to explore.
Her note in the pressure gage thread:
Hershal,
If the calibration provider is indeed accredited, and you have their cert and scope on file, why would the NIST numbers they provide, not be valid to prove their traceability?
Isn't that part of the "credibility" of being "accredited" (ie: aren't they audited for that by their AB)?
(May need to be a separate forum topic)
Jane
My response follows.....
According to the NIST Policy on Traceability, Section II.B.4, the "NIST numbers" are job numbers.....
"Test report numbers issued by NIST are used solely for administrative purposes. Although they often uniquely identify documents that bear evidence of traceability, test report numbers themselves do not address the issues listed in II.B.1 above, and should not be considered as the sole evidence of traceability."
NCSLI also has a position on the numbers.....
"Test report numbers issued by the National Institute of Standards and Technology (NIST) of the United States Department of Commerce are intended to be used solely for administrative purposes. Although they are often used to uniquely identify documents which bear evidence of traceability, test report numbers should not be used nor required as proof of the adequacy or traceability of a test or measurement."
All of which means that NIST numbers may be used as a COMPONENT of the traceability path, but then it MUST be tied to a SPECIFIC calibration at a specific time/place, under specific conditions, for a specific customer.
To use the NIST numbers then means you will need a copy of the individual report that goes with that specific calibration of the instrument. And if the report does not document the specific calibration of the specific instrument, then it is not valid for traceability, any more than a generic calibration procedure is.
Hope this helps.
Hershal
Her note in the pressure gage thread:
Hershal,
If the calibration provider is indeed accredited, and you have their cert and scope on file, why would the NIST numbers they provide, not be valid to prove their traceability?
Isn't that part of the "credibility" of being "accredited" (ie: aren't they audited for that by their AB)?
(May need to be a separate forum topic)
Jane
My response follows.....
According to the NIST Policy on Traceability, Section II.B.4, the "NIST numbers" are job numbers.....
"Test report numbers issued by NIST are used solely for administrative purposes. Although they often uniquely identify documents that bear evidence of traceability, test report numbers themselves do not address the issues listed in II.B.1 above, and should not be considered as the sole evidence of traceability."
NCSLI also has a position on the numbers.....
"Test report numbers issued by the National Institute of Standards and Technology (NIST) of the United States Department of Commerce are intended to be used solely for administrative purposes. Although they are often used to uniquely identify documents which bear evidence of traceability, test report numbers should not be used nor required as proof of the adequacy or traceability of a test or measurement."
All of which means that NIST numbers may be used as a COMPONENT of the traceability path, but then it MUST be tied to a SPECIFIC calibration at a specific time/place, under specific conditions, for a specific customer.
To use the NIST numbers then means you will need a copy of the individual report that goes with that specific calibration of the instrument. And if the report does not document the specific calibration of the specific instrument, then it is not valid for traceability, any more than a generic calibration procedure is.
Hope this helps.
Hershal