Hi Joe,
Welcome to the Cove
If you refer to the Note of ISO 13485, point 6.2.2, it states the following -
NOTE National or regional regulations might require the organization to establish documented procedures for identifying training needs.
Thus with regards to regulatory aspects, this requirement has been aligned to FDA's 21 CFR Part 820.25 which states the following -
820.25 Personnel.
(a) General. Each manufacturer shall have sufficient personnel with the necessary education, background, training, and experience to assure that all activities required by this part are correctly performed.
(b) Training. Each manufacturer shall establish procedures for identifying training needs and ensure that all personnel are trained to adequately perform their assigned responsibilities. Training shall be documented.
(1) As part of their training, personnel shall be made aware of device defects, which may occur from the improper performance of their specific jobs.
(2) Personnel who perform verification and validation activities shall be made aware of defects and errors that may be encountered as part of their job functions.
Thus FDA is bit more detailed. Hope this clarifies.