Transferring Calibration Certificates

K

KShannon

#1
Typically we calibrate all of our manufactured products in-house to known traceable standards. We are developing a new product that would have extensive capital costs if we were to purchase the equipment to calibrate it ourselves. We have an equally expensive option to send out each product to be calibrated by a third party. The supplier of the sensor which is the main component in the product can supply us a with a calibration certificate. If we can prove the inherent design of the product in no way alters the performance of the sensor, can we simply translate the sensor calibration data to our own certificate. Incidentally, we will have the capability to check each sensor against other known standards that are approximately as accurate. Unfortunately these other standards do not meet our corporate target of a 4:1 ratio.
 
Elsmar Forum Sponsor
K

KShannon

#3
Sorry for the confusion ... what I want to know is simply whether or not I can transfer calibration test results from the manufacturer of a part onto my own certificate of calibration without duplicating the actual test.
 
D

D.Scott

#4
IMHO I wouldn't think so. First, if the certificate were from your company, you would be expected to have all the test data and be able to meet all the requirements of 17025 pertaining to the calibration. Your certificate would have to list the name and address of the lab who did the calibration and the location where the calibration was performed.

There could be a way around this but I can't see it. It would seem easier to me to simply supply the sensor supplier calibration with the product. If the sensor is the part that gets calibrated and it is proven by your expensive testing that the calibration isn't changed by your inclusion in a product, why would you re-calibrate it? If there is a doubt as to the state of calibration after you do your part, work with the supplier to establish a method of verification (not a calibration) before the item is shipped.

Down the road when the end user needs to re-calibrate the sensor, who will he call, you or the supplier?

I trust the pros here will put us on the right track if I am wrong here.

Dave
 
R

Ryan Wilde

#5
Actually, 17025 is very clear in that you must state on your certificate that these tests were subcontracted.

But since you are a manufacturer calibrating product, I doubt that ISO 17025 is what you are trying to comply with, rather ISO or QS 9000 (correct me if I'm wrong). Neither of these standards has any direction involving this particular situation (ISO 9000 defers to ISO 10012, QS9000 requires that any calibration supplier be ISO 17025 [Guide 25] accredited). So I guess what you are stuck with is, if you are:

ISO 9000 registered, prove that it has no effect on the calibration, put it in writing, keep it on file, check it every so often, and document what you are doing;

QS9000 registered, make sure your supplier is either accreditied to ISO 17025 or has a handy dandy OEM waiver that AIAG is so keen on, and again, prove no degradation in writing;

ISO 17025 accredited/compliant/whatever then go through the same proof process, and make sure your supplier is accredited as well, and state on the certificate that "these tests were subcontracted to XXX Inc.", or possibly "these tests were performed by the manufacturer of the sensor", or any wording that clearly states that the calibration was not performed by your company. The reasoning is supposedly to do with traceability.

Hope this helps a bit.

Ryan
 
K

KShannon

#6
Thanks for the great input. It's nice to see that a common sense approach is potentially allowable (our Engineering staff will be shocked!) For the record, our company is currently seeking ISO9001 registration and will be looking at ISO Guide 25 next.
 
Thread starter Similar threads Forum Replies Date
V Going out of business - related obligation to transferring to another manufacturer EU Medical Device Regulations 0
P Transferring medical data from a device (Sec 201(h)): regulatory implications US Medical Device Regulations 3
R Transferring a calibrated counting scale to a sister company Calibration Frequency (Interval) 2
M Transferring AS9100 to new Legal Entity AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 12
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
H Transferring an ISO 13485 certificate to a new CB - Do I have to start from scratch? ISO 13485:2016 - Medical Device Quality Management Systems 2
B ISO 13485 Transferring a product from manufacturing facility 1 to manufacturing facility 2 ISO 13485:2016 - Medical Device Quality Management Systems 6
W Transferring epoxy from container to dispenser - Labeling AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
E Pallets need to be changed when raw materials are transferring from WH to MFG area? Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 6
Q Transferring EPA Registrations - Timeline? Various Other Specifications, Standards, and related Requirements 3
K Transferring Authorized Representative (AR) in Malaysia Other Medical Device Regulations World-Wide 7
D Transferring Paper Records to Electronic Record Archival Records and Data - Quality, Legal and Other Evidence 4
harrysons Automotive product transferring process what requirement to comply? IATF 16949 - Automotive Quality Systems Standard 3
kisxena Transferring a validated software platform to a virtual server Software Quality Assurance 8
K Process Validation - Production Line Transferring to a Different Facility Qualification and Validation (including 21 CFR Part 11) 16
J Help me make a judgement about transferring this process Six Sigma 6
S Netsuite Preventive Maintenance and Calibration Manufacturing and Related Processes 0
D Roughness specimen calibration Calibration and Metrology Software and Hardware 5
dhuot1990 General Calibration SOP Manufacturing and Related Processes 3
B Gage calibration Gonogo (calcul of cgk) General Measurement Device and Calibration Topics 0
Vader22 As-Found FAILED calibration ISO 14001:2015 Specific Discussions 16
M Gravimetric blender for resin in automotive industry - periodical calibration via external lab with certification Reliability Analysis - Predictions, Testing and Standards 0
N Calibration Method for Scales and Flowmeters Reliability Analysis - Predictions, Testing and Standards 7
T Calibration Tracking ISO 13485:2016 - Medical Device Quality Management Systems 4
Q Considering addition of mass calibration to ISO17025 scope ISO 17025 related Discussions 6
Z Resolution in calibration General Measurement Device and Calibration Topics 1
J Run Down Adapters Calibration Processes General Measurement Device and Calibration Topics 4
S Looking for accredited calibration lab - Onsite up to 650A and 1 kV. General Measurement Device and Calibration Topics 3
CaliperJim Screw Thread Micrometer Anvil Calibration General Measurement Device and Calibration Topics 7
Strange Juan HP 85054B Standard Mechanical Calibration Kit Calibration Procedure General Measurement Device and Calibration Topics 1
H Air Pressure Gage Calibration General Measurement Device and Calibration Topics 6
T Calibration on AVL AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
C Determining an as received OOT condition during calibration does not affect measurements taken prior to calibration. General Measurement Device and Calibration Topics 16
Sam.F Do heat treatment ovens need calibration? General Measurement Device and Calibration Topics 22
J Selecting reference standards for calibration General Measurement Device and Calibration Topics 3
C IATF 16949:2016 Major NC pertaining to 7.1.5.2.1 Calibration/verification records IATF 16949 - Automotive Quality Systems Standard 15
E Calibration Records during AS9100 Audit AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
John Predmore Calibration Determination decision Calibration Frequency (Interval) 15
J ISO 17025 Calibration Other Medical Device Related Standards 2
A Electronic Weighing scale calibration Calibration Frequency (Interval) 4
M Bourdon gauge calibration procedure failure ISO 17025 related Discussions 3
Ed Panek Calibration question Manufacturing and Related Processes 4
S Calibration/Worn gages Manufacturing and Related Processes 9
A Settings rings calibration General Measurement Device and Calibration Topics 4
R 7.1.5.2 -- Calibration AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
C CALIBRATION CERTIFICATE REQUIREMENTS General Measurement Device and Calibration Topics 11
S Calibration Label issues General Measurement Device and Calibration Topics 21
H Calibration / Capability in a PCP FMEA and Control Plans 1
R First Time Managing Calibration and Measurement System. Need Help. General Measurement Device and Calibration Topics 19
M CE marked medical device - do they need UKAS certified calibration provider ISO 17025 related Discussions 4

Similar threads

Top Bottom