Transformer Overload Test - Need help

EMengineer

Involved In Discussions
#1
cl. 15.5.1.3 (Overload test) says that, if first protection activates, You need to shunt it. After that you make a little smaller current, which activated protection, and then test the transformer or wait till second protection device enables. The question is: why to use smaller current? If I use smaller current than that which enabled first protection device, what is the point of smaller current? It shouldn't activate second protection device because it's value is even smaller than that current which activated first protection device.
Could someone explain this? I would appreciate for help.
P.S. First protection device is electrical protection from short circuit, second would be fuse
P.S.S. Sorry for my English (not native language), hope You'll understand :)
 
Elsmar Forum Sponsor

Peter Selvey

Staff member
Super Moderator
#2
The second protective device is usually not relevant.

If it is, it is usually because the first protective device is current based while the second is thermal based.

For example electronic protection is set at 1.00A.

So, the test is performed at 0.99A (just below).

The transformer and switching parts slowly rise in temperature and eventually trips a thermal based device e.g. a thermistor on the heatsink for the main switching transistor, or a thermal fuse in the transformer. That would be a secondary protective device. It would stop the test.

In such a case, the secondary protective device becomes the real "protective device" and should be controlled as a safety critical part.
 

mehrnaz

Starting to get Involved
#3
Dear Mr. Selvey
I have a question regarding the short circuit test of transformer output winding. if there is a transformer which has 2 output voltage of 12 and 24 volts. according to clause 15.5.1.1:
Transformers of ME EQUIPMENT shall be protected against overheating in the event of short circuit or overload of any output winding.
Each winding is tested, in turn........
This means that there should be protection in the case of short circuit of the 12v winding at once and in the other test short circuit of 24v winding shall be tested?
Is this necessary to check the short circuit between 12 v and 24 v, which are beside each other in the output terminal of transformer?
Because if I short circuit 12v and 0 volt at once or 24v and 0 volt in another test the transformer passes the test, but by short circuiting 12v and 24v the protection does not react to this test and the temperature of output wires exceeds...

Thanks in advance...
 

Peter Selvey

Staff member
Super Moderator
#4
In a way this is two different subjects; windings, as an energy source that supplies secondary circuits, which need to be rigorously tested for overload and short circuit by virtue of being connected to a complex circuit that no body really wants to delve into the billions of possible fault combinations that could occur in that circuit. Just add some protection and make sure it is effective no matter what, and everyone can sleep.

Separately we the possibility that the 12V circuit can short to the 12V circuit. That's a valid concern but the fault permutations are just one (1). In that case it's worth to investigate the fault directly. For example, inspecting the creepage and clearance, dielectric strength. If these don't meet the requirements for 2MOP, short away. If the short fails the temperature limits, it's a fail result.
 
Thread starter Similar threads Forum Replies Date
S How to overload a linear transformer with multiple protective devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
B Transformer Requirements - IEC 60601-1 vs. IEC 61558-2-15 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
G Where to buy Medical Grade Transformer IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
N Triple insulated wire for transformer isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E Creepage distance accross insulation of step up transformer EU Medical Device Regulations 2
A MOPs needed after Isolation Transformer IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Transformer Compliance Testing, overheat then humidity treat, then dielectric test? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
alonFAI Gaps between the Bobbin and the Ferrite Core in a Transformer Manufacturing and Related Processes 1
R Question on IEC 60601 clause 15.5.1.1 (transformer overheating) IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S How to implement protection for multi-winding transformer? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
P PSU Transformer Tests Approved to IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W Supply Voltage Condition for Transformer O/L and S/C (open & short circuit) testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
D Transformer Oil Analysis - Help with a a Duval's Triangle Preventive Action and Continuous Improvement 3
P Test Voltages for a Pulse Transformer used to Isolate the Patient Connection IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
L Quality Assurance Manual for Transformer Oils Service Industry Specific Topics 2
E Ethernet Isolation Transformer for Medical Grade PCB IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
R How to determine the reference voltage for three-phase transformer? IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
G 50 Turn Transformer Coil Calibration Technical question General Measurement Device and Calibration Topics 10
D UL Certified Isolation Transformer Various Other Specifications, Standards, and related Requirements 3
L looking for a forum related to motor/transformer lamination manufacturing Book, Video, Blog and Web Site Reviews and Recommendations 2
G What differentiates a "bridging transformer" from an ordinary "isolation transformer" General Measurement Device and Calibration Topics 5
N Are Overload Test Results expected to be maintained through out Life of Product? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R Windings Temperature exceeds the Limit - Overload Test of Switching Power Supply IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Marc Information Overload World News 3
E Test report to certify compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 3
E Tensile test machine, tinius olsen 602 Manufacturing and Related Processes 1
M Test failure Root cause not found Customer Complaints 10
Moumen H Variations between ASTM A29 Standard for steel bars and Mill test certificates specs Manufacturing and Related Processes 1
D Tip test 10° -> 5° IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Sponsor name presented in the test report EU Medical Device Regulations 4
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
S OEM full range calibrations vs Validation special test points Medical Device and FDA Regulations and Standards News 0
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Shelf Life Test for IFU and Implant Card Other Medical Device Related Standards 1
S Mechanical Test Under FDA Freedom of Information Act Medical Device and FDA Regulations and Standards News 5
S Critical characteristic on manufacturing operations test procedure Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S Knee Implant (Femoral -Cobalt chrome)-Sub chronic toxicity test (ISO 10993-11)choice of root Medical Device and FDA Regulations and Standards News 2
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
N Is there a need for clinical test of Class IIa products (for MDR)? EU Medical Device Regulations 2
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Non parametric test for semi-quantitative data. Statistical Analysis Tools, Techniques and SPC 5
Ed Panek Disaster Recovery Test IT (Information Technology) Service Management 3
A How much does a complete biocompatibility test package cost? Other ISO and International Standards and European Regulations 1
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Z Is Op Amp Common-Mode Rejection Ratio (CMRR) test applicable to battery operated devices? Other Medical Device Related Standards 0
E When to perform accelerated aging studies and transport test(OQ or PQ) Qualification and Validation (including 21 CFR Part 11) 4
U Does *anyone* know a lab that will test to EN 455-4 Medical Gloves shelf life determination? EU Medical Device Regulations 1
A Test requirements for marketing in US when RATED values are not of US IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
H Secondary Circuits, Voltage Test and Clearance/Creepage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0

Similar threads

Top Bottom