#1
Hello all,

Recently our lab decided to start mixing our own 5% salt solution for our two salt fog chambers. We installed a DiH2O water system which uses mixed bed tanks and monitor the resistivity levels daily to make sure they are within the specs of ASTM D1193. We purchased Morton Culinox 999 salt because it gave us the purity required. After mixing solution, we check the specific gravity of the final product with a hydrometer to make sure it's within the 4-6% range outlined in B117 and I will use NaOH or HCl to target pH 6.8.

Now, mixing a 5% solution is simple enough on paper but we have run into a few issues that maybe someone can help us troubleshoot. We have seen the pH in our collection tubes fluctuate up and down and sometimes land below the 6.5-7.2 range. Currently the 5% solution in our tank is pH 6.8 and spec grav 1.034, but the collected solution is pH 6.5 and spec grav 1.039. Based on B117 I was under the impression that the pH should increase slightly after atomization. What can we do control the pH better? Is it normal to expect the values to fluctuate? Also, what could lead to the specific gravity changing that much? We have taken apart the atomizer to inspect/clean it and we have replaced the filters in the tower and the reservoir tank, but nothing changed.

We are still in the R&D phase of this process, so any insights or resources are welcomed and appreciated. Thanks in advance.


Regards,

Kane
 
Elsmar Forum Sponsor

Miner

Forum Moderator
Staff member
Admin
#2
I would start with verifying the R&R of your measurement instruments. Are you certain that the fluctuations aren't simply measurement error?
 

kane

Registered
#3
We have been checking our pH probe against certified calibration fluids, so I generally feel good about the measurement reading. I had a technician use a second probe to compare the numbers and they were the same.
 
Thread starter Similar threads Forum Replies Date
A Call Center troubleshooting guides Medical Device and FDA Regulations and Standards News 5
A Plastic Injection Molding Process Troubleshooting Guide Manufacturing and Related Processes 1
D Parts removed for troubleshooting reason EASA and JAA Aviation Standards and Requirements 2
A Unreleased software for troubleshooting and clinical use ISO 13485:2016 - Medical Device Quality Management Systems 6
G Question concerning process troubleshooting and process FMEA usage in job shops FMEA and Control Plans 5
Marc Troubleshooting and Information Site - A Recommendation After Work and Weekend Discussion Topics 8
T HP3325A Function Generator - Troubleshooting Modulation on Output General Measurement Device and Calibration Topics 8
Marc Things to consider in a Beer FMEA - Beer Troubleshooting FMEA and Control Plans 7
T DSC Troubleshooting Guide-controlled document or not? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 15
B Guidance Required for Examining 168 Hour Salt Spray Test Samples Manufacturing and Related Processes 5
Q Salt spray test on anodic coating B117 vs NES M0140 Reliability Analysis - Predictions, Testing and Standards 2
R Salt spray per Mil-DTL-5541 coupons required 5 or 10 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
A Salt Bath Probe Style SAT requirement Waiver per AMS 2750 Rev E Section 3.4.7 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
J Salt Spray Failure Question (ASTM B 633 and B 117) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
S Measurement Uncertainty for Neutral Salt Spray test to BS EN ISO 9227:2006 General Measurement Device and Calibration Topics 4
S How to maximize the use of a Calcium Salt Dust Collector Food Safety - ISO 22000, HACCP (21 CFR 120) 3
S Template for Product Specifications for Calcium Salt with the Chemical Composition Food Safety - ISO 22000, HACCP (21 CFR 120) 3
T Control Chart - Salt Spray Test Hours - Powder Coated Parts Statistical Analysis Tools, Techniques and SPC 3
S Salt spray test on nickel? Parts were plated to QQ-N-290 Manufacturing and Related Processes 5
J Salt Spray Test on 3 year old Zinc Die Casting Chromed part per ASTM B 117-03 Manufacturing and Related Processes 2
J 200 hour Salt Spray per ASTM B 117 - What should be reported? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
P Found Blue Green Rust in Bolt after Salt Spray Testing Reliability Analysis - Predictions, Testing and Standards 11
C An Explanation of Salt Spray (Corrosion) Testing based on ASTM B117 Design and Development of Products and Processes 1
P Salt spray testing - Comparative corrosion data of 302 S26 vs Galvanised carbon steel Manufacturing and Related Processes 6
C Grammar Question - Please, pass the salt Coffee Break and Water Cooler Discussions 15
Z Salt Spray Procedure to ASTM B 117-90 - Seeking Example Various Other Specifications, Standards, and related Requirements 8
P What are GM's requirements for Salt Spray Testing? General Measurement Device and Calibration Topics 1
A Uncertainty in Test - Durability test, High Performance, Salt Spray chambers, etc Measurement Uncertainty (MU) 5
B The Salt Lake City Olympics Coffee Break and Water Cooler Discussions 1
V Salt spray testing procedure and training requirements - ASTM B-117? QS-9000 - American Automotive Manufacturers Standard 4
B Improper R&R Test? Downdraft Velocity in Automotive Spray Booths Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
C Advise on Oral Spray Pad Printing and Compliance 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
M Quality issues in a manufacturing plant - High-pressured solvent spray cans Quality Manager and Management Related Issues 2
T Supplier Spray Process Audit - Fish Eye, Spray Dust and Fiber Dust Defects Manufacturing and Related Processes 8
L Dry Lube Application (Ever lube 620) - Manual Application vs. Automated Spray Process Manufacturing and Related Processes 1
T Spray Paint Manufacturing Processes and related Information Manufacturing and Related Processes 3
R Spray Drier Equipment Optimization - How to start Statistical Analysis Tools, Techniques and SPC 6
S Insecticidal spray and ISO 14001 ISO 14001:2015 Specific Discussions 11
V Is spray painting allowed for non-blood contact medical devices? ISO 13485:2016 - Medical Device Quality Management Systems 3
C SPC In a Manual Spray Paint Application Statistical Analysis Tools, Techniques and SPC 7
W Plastic Part Painting/Spray Failure Mode - Sample PFMEA or control plan Manufacturing and Related Processes 7
E Spray Finishing Booth - Fire Resistant Walls Requirement (Why not a curtain?) Various Other Specifications, Standards, and related Requirements 4

Similar threads

Top Bottom