Trusting Gage Manufacturer Calibration Certifications

Michael_M

Trusted Information Resource
#1
We are not AS9100 certified yet (working towards it), but we do work for aerospace:

A little over 1 year ago we got a part that has Cadmium plate (.0002/.0005 thick). Per our standard method of manufacturing, we buy pre-plate thread ring gages to check the thread acceptance during the manufacture of the part changing the pitch diameters of the go and no-go based on the stated plate thickness. We also verify the pitch diameter with a thread micrometer a few times during the run (as a reference).

Now, 1 year later, I send our gages out to be calibrated and they come back with the pitch and minor diameter measuring per machinist hand-book as a normal thread. The gages themselves list the pre-plate pitch diameter and that they are pre-plate gages. The gage certifications I got with the gage show they are 'pre-plate'.

This brings up an issue. How far can I trust gage manufacture certifications? I realized this is most likely an isolated incident but..... If we had gotten a rejection on the parts, I most likely would not have sent the gages out to verify size, I would verify the certifications (which show good).

If this happened to you, what would you do. Has this happened to you, if so, what did you do. I am sending the gages back to the manufacture to review, should I do something else?
 
Elsmar Forum Sponsor

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#2
Personally, I generally do not accept a manufacturer's calibration, except when issued under an accreditation, so that there is sufficient oversight.

Otherwise, send the gages out to get calibrated by an accredited calibration lab, with that in their scope, and importantly, make sure they give you the uncertainty. Uncertainty is error, which you need to know.

Hope this helps.
 

BradM

Staff member
Admin
#3
Michael, as a general rule for me, not very far.

Manufacture companies are really good at making their product; most of the time calibration is a secondary activity. Some of the worst certificates I've seen come from mfg.; they use some internal (proprietary) procedure, and it's not very clear what they verify.

Too, I have to watch mfg. Many times they provide some certificate of conformance or something; and did not ever verify the instrument that I have. They just did some lot acceptance test.

I'm in the same camp with Hershal. If they are not accredited, I will only use a mfg. calibration service if there is no other vendor that will calibrate it. Otherwise, send it to a competent calibration vendor.
 

DietCokeofEvil

Trusted Information Resource
#4
Did the calibration cert you received indicate how the rings were calibrated? Unless you sent them to the manufacturer or verified that the cal house had the correct setting plug- they were probably not done on a setting plug.

If that is the case, I recommend you purchase a setting plug. A set plug is much easier to verify and then set rings to.

On the topic, I never trust a mfr's certificate. Compliance certificates mean nothing, and as a technician who sees new gages come in all the time, the amount of gages I send back for rework would probably make your head hurt. I always urge our customers to let us check the gages when they purchase. It's not to make extra $$- it's to verify that the gages are correct for their own protection.
 
Thread starter Similar threads Forum Replies Date
M Trusting ISO 13485 Certification of a Supplier... A Sad Story ISO 13485:2016 - Medical Device Quality Management Systems 116
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
L Gage R&R TMV Acceptance Criteria Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
S MSA for attribute relation gage Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
E Zero part to part variation - Gage R&R project Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 15
M Gage R&R and right way to measure Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 16
L Gage RandR on automated equipment. IATF 16949 - Automotive Quality Systems Standard 7
B Gage calibration frequency, ISO and IATF - What are the requirements Calibration Frequency (Interval) 3
D Difference between Test Method Validation and Gage R&R Qualification and Validation (including 21 CFR Part 11) 18
R Determining Uncertainty from Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
C Gage Block Wringing General Measurement Device and Calibration Topics 3
B Gage R&R with NDC=1 Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
S Type 1 Gage R&R or something else? Reliability Analysis - Predictions, Testing and Standards 6
G Should I perform Gage R&R only at the beginning of a new project? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
V Thread Plug Gage Pitch GO Diameter out of spec AQL - Acceptable Quality Level 8
G Gage R&R - Where am I going wrong? Part of a FAIR submission (Aerospace) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
R Gage R&R Excel templates Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
K Gage R&R with more than 3 appraisers Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
Anerol C Gage R&R Template AIAG 4th Edition IATF 16949 - Automotive Quality Systems Standard 3
T Attributes SPC study - Attributive control (Go gage) Statistical Analysis Tools, Techniques and SPC 5
S Capability or Gage R&R Study for Leak Tester? Reliability Analysis - Predictions, Testing and Standards 15
N % Tolerance - Type 1 study on the gages, then a gage R&R (ANOVA) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
M Definitive answer on Type 1 vs Type 2 vs Type 3 Gage Study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
V Gage Management and Gage R&R Software General Measurement Device and Calibration Topics 1
D Gage R&R Study on Load Cells - Large chemical blending tanks IATF 16949 - Automotive Quality Systems Standard 1
B Minitab Type 1 Gage Study on True Position Question Measurement Uncertainty (MU) 1
S Gage R&R on alloy Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
8 Need Help - Runout - Function Gage Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
L Gage R&R studies for identical Devices - Need to confirm the requirement to perform them Reliability Analysis - Predictions, Testing and Standards 2
G Stab Pin Size For Check Gage Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
E Functional Gage for TOP applied at RFS Design and Development of Products and Processes 5
F No reproducibility Error in Gage R&R? Reliability Analysis - Predictions, Testing and Standards 4
M Template for Attribute Gage R & R wanted Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
G Warning about Deltronic Gage Pins - The "Certification of Accuracy" document General Measurement Device and Calibration Topics 13
A Gage type 1 study on CMM Capability, Accuracy and Stability - Processes, Machines, etc. 2
Q Gage R&R for Instant Measurement Machine Manufacturing and Related Processes 6
M Gage R&R with large inter part variance - Torque Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
D Do we need normal data for gage r&r studies? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
D Do I need part variation while doing Destructive Variable Gage R&R MSA study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
S Gage R&R Standard Deviations Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
J Informational Gage R&R: Swapping operators Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
S Gage R & R for large amount of inspections IATF 16949 - Automotive Quality Systems Standard 10
S Ring Gage Thermal Expansion General Measurement Device and Calibration Topics 4
L Gage R&R for test precision or uncertainty prediction? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
S Minor diameter of threaded ring gage General Measurement Device and Calibration Topics 1
A Does Torque Gage Need Validation? Reliability Analysis - Predictions, Testing and Standards 4
J Ignoring a Failed Gage R&R and upcoming external audit IATF 16949 - Automotive Quality Systems Standard 22
R Scaling a Gage R&R - 2 holding fixtures each with 10 nests Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
B Gage R&R Acceptable (10-30%), deduct Total Variation from Tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12

Similar threads

Top Bottom