Two PT100 readings

C

chriswhitehead

Hi. I have two reference PT100's. What would be an acceptable difference between the two probes in the same bath at the same temperature. Mine seem to be about 2mK apart all the time. Is this good or bad?
:deadhorse:
:bonk:
:agree1:
 

dwperron

Trusted Information Resource
It looks like you are getting very good readings there!

Your results will be based on multiple factors such as the temperature stability of the baths (5 mK is a fairly typical specification for a good bath), the accuracy of the probes (good probes have an accuracy around 15 mK and annual stability around 10 mK), the meter accuracy (5 mK for a good meter), factors such as immersion depth, etc.

All in all if I had two PT100 probes that are consistently 2 mK apart I would be very satisfied!
 

dgriffith

Quite Involved in Discussions
I would never be satisfied with 2mK ... but I would understand that it's pretty darn good... :tg:.

DW is right, many variables in stirred liquid bath, including temperature gradients in 3 dimensions. You might get better if you used the same display channel, probe in same location, each probe's Ro checked in a properly realized ice bath. But why bother-- unless you're getting paid to do it!
 
Thread starter Similar threads Forum Replies Date
S Why does GAGEpack drop the temperature and humidity readings out of the verification event card? Quality Assurance and Compliance Software Tools and Solutions 0
E Performing a GRR when you get multiple readings over time per part. Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
M Measuring FIM (TIR) - Two inside diameters - Conflicting readings between inspectors Manufacturing and Related Processes 1
R Laminate ESD Mats - Low resistance (RTT) readings Manufacturing and Related Processes 6
T Calibration Certificate Readings After Adjustment General Measurement Device and Calibration Topics 5
B Reducing variation through multiple readings Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
J Reducing Variation via Multiple Readings Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
C National/International Standard For Calibration "As Found" Readings General Measurement Device and Calibration Topics 4
J Method to determine the Uncertainty of the readings of a Laser Scanner Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
B ESD Mats on Workbenches and ESD Surface Resistivity Meter Readings Manufacturing and Related Processes 6
M Gage pin calibration - Actual calibration readings required on the calibration cert General Measurement Device and Calibration Topics 3
S Gage R&R using only 2 appraisers and 2 readings?? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 17
C Interpretation of Decimal Readings Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 13
P Criteria for operator's readings at setup - Seeking help to explain chosen limits. Statistical Analysis Tools, Techniques and SPC 4
J Calibration Data Before & After Readings (TS 16949 Requirement) General Measurement Device and Calibration Topics 8
C Reducing Variation through Multiple Readings - MSA 3rd edition Chapter 5d Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 8
B Gage R&R for Automatic Test Systems - No operator influence on readings Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
G Recording Monitoring and Measuring Readings - Record inspection gages we use? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
J Gage R&R - Consistent Readings, Inconsistent Results? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 8

Similar threads

Top Bottom