UDI (barcode, datamatrix) Verification Solutions

Elsmar Forum Sponsor

howste

Thaumaturge
Super Moderator
#12
:topic:

I realize that this thread is in the medical device regulations forum, but I thought I'd point out that barcode and Data Matrix requirements exist for defense/military hardware as well. You never know who will come across this thread in a search.

The verification requirements for defense/military are specified in MIL-STD-130N (with change 1) clause 5.2.7. "Marking quality conformance shall include validation and verification of mark content and quality." Acceptance criteria for linear bar codes are based on ISO/IEC 15416 . Acceptance criteria for data matrix can use ISO/IEC 15415, AIM DPM-1-2006, ISO/IEC TR 29158, or SAE AS9132. Calibration must be traceable to GS1 US Calibrated Conformance Standard Test Cards.
 
M

miken99

#13
In GS1 general spec
5.5.2.7.6 Symbol specification table 6 - Regulated healthcare non-retail consumer trade items not scanned in general distribution
Figure 5.5.2.7.6-1. GS1 system symbol specification table 6 Symbol(s) specified X-dimension mm (inches) Minimum symbol height for given X mm (inches) Quiet Zone Minimum quality specification
Minimum Target Maximum For minimum X-dimension For target X-dimension For maximum X-dimension Left Right
GS1- 128 0.170 (0.0067") 0.495 (0.0195") 0.495 (0.0195") 12.70 (0.500") 12.70 (0.500") 12.70 (0.500") 10X 10X 1.5/06/660
GS1 DataMatrix (ECC 200) (*) 0.254 (0.0100") 0.380 (0.0150") 0.495 (0.0195") Height is determined by X-dimension and data that is encoded 1X on all four sides 1.5/08/660
Sorry about the formatting but you willl get the idea - 1.5 for the average of 10 scans.

In HIBCC
5.0 Print Quality
5.1 Code 128 or Code 39
The bar code symbol quality for a Code 128 or Code 39 symbol in its final configuration shall be no lower than a
C/06/660 when measured according to ISO/IEC 15416 Information technology -- Automatic identification and data
capture techniques -- Bar code print quality test specification -- Linear symbols. Labelers should attempt to reach
B/06/660 or better at the time of printing.
5.2 Aztec Code, Data Matrix or QR Code
The bar code symbol quality for an Aztec Code, Data Matrix or QR Code symbol in its final configuration shall be no
lower
than a C/06/660 when measured according to ISO/IEC 15415 Information technology -- Automatic
identification and data capture techniques -- Bar code print quality test specification -- Two-dimensional symbols.
Labelers should attempt to reach B/06/660 or better at the time of printing.

In the spirit of the UDI rule, most understand that these barcodes must be able to be read, easily and accurately as the products move through the global supply chain. That will mean using laser scanners, imagers, from manufacturers in all parts of the world. Verification is both logical and required.
 

greif

Involved In Discussions
#14
A quick bit of info: I am told that GS1 will no longer be selling Calibrated standard test barcode cards. 1D and 2D cards are available at
AppliedImage.com (the 2D is just been released this week).
2D card PN is AI-CCS-DM-E

Looks like it is up on the website here:
Applied Image Barcode Calibration Standards

Scanner: reads the data encoded in a bar code
Verifier: reads the symbol and grades the quality on several parameters
The calibrated standard test cards are calibrated on a master verifier (10x or better than the typical industrial verifier) and supplied with documentation for photometric and linear distance traceability.


Full disclosure: I work at Applied Image and helped build the 2D calibration equipment (it was fun!)

*****************************************************
Moderator note: Caution - This post is under moderator discussion. There is no evidence that GS1 will no longer be selling Calibrated standard test barcode cards.
 
Last edited by a moderator:

howste

Thaumaturge
Super Moderator
#15
A quick bit of info: I am told that GS1 will no longer be selling Calibrated standard test barcode cards.
Do you have a source for this info that you can link to? Does this imply that there will no longer be a nationally traceable standard? :mg:
 

greif

Involved In Discussions
#16
I have only heard rumors that GS1 is getting out of the calibration card business, but their website store only has 2 types of cards listed (a 2D and a RSS)- which looks like getting rid of existing stock. They have been out of stock on 1D UPC cards for a year or more.
******************************
Moderator note: Caution - There is no evidence that GS1 will no longer be selling Calibrated standard test barcode cards.
 
Last edited by a moderator:
M

miken99

#17
I know that GS1 has built a 'super verifier' for linear barcode calibration cards. It was delivered fairly recently so it may just be an issue of starting up the system.
We did not build the system but we were involved and very familiar with the project.
 

greif

Involved In Discussions
#18
I knew they were working on a new 1D master verifier (I think the original delivery was supposed to be a year or two ago).

This is one of the reasons I wonder if they will pull out of the market for calibrated cards. Likely they will pull out, since they need a source of these 1D cards where all the symbols grade exactly in the center 1/3 of the grade range, and they are extremely difficult to make. (and it is a pretty small market for an outfit their size).
 

greif

Involved In Discussions
#19
Question; does Mil-std-130N restrict the testing of barcode verifiers only to GS1 test cards, or can any card that is referenced to nationally traceable standards be used?


Mil-std-130N says:
5.2.7.2.2 Calibration. Calibration processes and materials for reflectance criteria provided
by the verifier manufacturer are acceptable for printed labels when referenced to the nationally
traceable standard (GS1 US Calibrated Conformance Standard Test Cards)

Maybe this is a little ambiguous, but wouldn't any Primary reference test symbol that was calibrated to "the nationally traceable standard" be suitable? Do the parentheses in the Mil Std. have special significance?
 

howste

Thaumaturge
Super Moderator
#20
Question; does Mil-std-130N restrict the testing of barcode verifiers only to GS1 test cards, or can any card that is referenced to nationally traceable standards be used?

Mil-std-130N says:
5.2.7.2.2 Calibration. Calibration processes and materials for reflectance criteria provided by the verifier manufacturer are acceptable for printed labels when referenced to the nationally traceable standard (GS1 US Calibrated Conformance Standard Test Cards)

Maybe this is a little ambiguous, but wouldn't any Primary reference test symbol that was calibrated to "the nationally traceable standard" be suitable? Do the parentheses in the Mil Std. have special significance?
I don't claim to be an expert on this, but when I read MIL-STD-130N it seemed to imply that the GS1 cards are the ones they require. I'd welcome discussion on any other opinions.
 
Thread starter Similar threads Forum Replies Date
C UDI Label Verification - How do we ensure the barcode is correct and prove it? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
I UDI requirements vs Japanese Barcode Marking guideline Japan Medical Device Regulations 1
P UDI-PI requirements on reusable surgical device, do we need serialisation? ISO 13485:2016 - Medical Device Quality Management Systems 0
J UDI - Where to document it? EU Medical Device Regulations 8
N UDI: GS1 or HIBC ??? EU Medical Device Regulations 1
F UDI-PI required on packaging (MDR) EU Medical Device Regulations 2
C UDI Questions on Control Units Other US Medical Device Regulations 6
M Supplier requirements - Major supplier is a Non-Profit registered with ICCBBA (FDA UDI) Supply Chain Security Management Systems 12
K Two FDA questions regarding UDI (and potentially 803.52 MDR) Other US Medical Device Regulations 0
V UDI - OEM or OBL / Legal Manufacturer Responsibility CE Marking (Conformité Européene) / CB Scheme 0
A European UDI-DI CE Marking (Conformité Européene) / CB Scheme 2
N FDA UDI - Label vs. Labeling - Does the insert need to include UDI? Other US Medical Device Regulations 1
K Is UDI (Unique Device Identifier) symbol mandatory under EU MDR? EU Medical Device Regulations 3
M UDI procedure/ report 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
R Shall a new UDI-DI be required when stand-alone software device's version is updated? EU Medical Device Regulations 1
D Hospital Decontamination and the MDR - Technical Files, UDI and PMS EU Medical Device Regulations 3
F UDI - Unit of Use and Class I, single-use devices EU Medical Device Regulations 4
F UDI Registration - Who is responsible for Registration of the Device? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
M UDI Database and EUDAMED - Will there be two separate databases? EU Medical Device Regulations 4
Q Must product name be listed the same name in FURLS, UDI, GUDID and Company Website? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
Q LOT or Serial Number Symbol not used when the information is contained in the UDI? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
J Looking for a template SOP for UDI implementation for EU Medical Devices EU Medical Device Regulations 1
M Basic UDI and technical documentation EU Medical Device Regulations 1
A UDI and Design Controls - Labeling change via the Design Control process 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
M MDR Requirements - UDI/Barcodes/QR codes on IFUs EU Medical Device Regulations 4
M UDI transition period and implementation EU Medical Device Regulations 6
M UDI size on packaging EU Medical Device Regulations 3
R UDI for US - Third Party device Repackager Other US Medical Device Regulations 9
M UDI placement - Packaging levels EU Medical Device Regulations 2
L UDI Requirments Gap Analysis - EU MDR vs FDA CFR EU Medical Device Regulations 8
M DI format for UDI after we have registered it with GS1 Other US Medical Device Regulations 0
P UDI labeling for power module IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
DitchDigger UDI, Labeling Accessories, General Insanity, Etc. US Food and Drug Administration (FDA) 1
K China UDI Implementation for October 1st 2019 China Medical Device Regulations 4
A What is the difference between Basic UDI-DI and UDI-DI? EU Medical Device Regulations 6
M UDI-PI on a package that contains more devices EU Medical Device Regulations 3
M Informational EU – Unique Device Identification (UDI) System – FAQs Medical Device and FDA Regulations and Standards News 0
F EUDAMED UDI Medical Devices User's Guide Medical Device and FDA Regulations and Standards News 7
J Class 1 Medical Device - Using a UPC over the UDI? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
Z Do we still have to put UDI if we have DI and lot number seperately on our product ? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
M EU Basic UDI-DI Trigger Points - When is a new Basic UDI-DI required EU Medical Device Regulations 3
P UDI-PI and the MDR - Is the UDI-PI required for all devices/classifications? EU Medical Device Regulations 8
M Informational US FDA issued the first warning letter for UDI violations to help ensure compliance Medical Device and FDA Regulations and Standards News 0
K Hospital UDI requirements Other Medical Device and Orthopedic Related Topics 7
J FDA Class 1 Medical Device UDI requirements Other US Medical Device Regulations 19
J UDI Requirements - Products that all fall under the same family 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M Informational EU – MDCG 2018-1 v2 Guidance on basic UDI-DI and changes to UDI-DI Medical Device and FDA Regulations and Standards News 0
F MDR 2017/745 Basic UDI-DI, UDI-DI & UDI-PI EU Medical Device Regulations 0
M Informational EU – EUDAMED UDI Device Data Dictionary + data sets Medical Device and FDA Regulations and Standards News 0
M EU MDR - UDI requirements and responsibility guidance EU Medical Device Regulations 1