UDI needed when we submit to ON?

LAM2020

Starting to get Involved
#1
hELLO
we are a manufacturer of class IIb devices and we are looking to submit our technical files to an ON according to the mdr 2017/745. one thing I am not sure about is when we have to submit our file do we have to have the UDI on the Labels already in place. Or can we submit our labels without it on it as we do not have anything in place at the time.
Thank you
 
Elsmar Forum Sponsor

xcanals_tecno-med.es

Involved In Discussions
#3
Hi
You need to asign the basic UDI-DI and the UDI-DI to the devices which will be evaluated by the NB (from 26 may 2021)
And this information must be included in the technical documentation according annex II of MDR
The use of UDI-DI and UDI-PI in the labelling depends on the classificaction of the devices:
class III - from 26 may 2021
class IIa and IIb - from 26 may 2023
class I - from 26 may 2025
BR
Xavier Canals
 

LAM2020

Starting to get Involved
#4
Thanks a lot for your reply,
so the medical devcie will be assigned a udi-di and for the packaging of the device is the ud-di the same or do i have to assign a different one for the packaging knowing that it will be a package of 1. thanks
 

xcanals_tecno-med.es

Involved In Discussions
#5
In short, yes. only one device + packaging with one UDI-DI , same in both

but note that the medical devices is not only the "device" itself you must consider the total product = device + label + ifu + packaging
for example if you have 2 variants :

1. "device" + IFU English
2. "device" + IFU Spanish

they must have different UDI-DI
Think always like a supermarket cashier: different (final) product --> different barcode (UDI)

Xavier Canals
 
Thread starter Similar threads Forum Replies Date
J UDI-DI how should we interpret Device version or model to determine if a new UDI-DI is needed? EU Medical Device Regulations 0
M UDI barcode quality EU Medical Device Regulations 3
C How to place software version for SaMD product in HIBC secondary data structure (UDI-PI)? Other US Medical Device Regulations 0
C Issues with UDI-DI and LAA/LUA registration in Eudamed EU Medical Device Regulations 1
Watchcat Mobile app - Different UDI for different smartphones? Other US Medical Device Regulations 7
Watchcat One UDI per country? ISO 13485:2016 - Medical Device Quality Management Systems 3
Watchcat UDI Timeframes? Other US Medical Device Regulations 5
A UDI-PI for IVDs used in house EU Medical Device Regulations 5
G SRN, Basic UDI, Legal Manufacturer EU Medical Device Regulations 6
A Basic UDI and UDI-DI for hip implant systems EU Medical Device Regulations 0
T FDA UDI Question - Class II Medical Device Other US Medical Device Regulations 2
R UDI Number Identification EU Medical Device Regulations 5
I Eudamed UDI-Module EU Medical Device Regulations 9
L Configurable UDI mdr EU Medical Device Regulations 0
L UDI implementation in Saudi Arabi (KSA) Other Medical Device Regulations World-Wide 0
L UDI-DI - Device will be sold by itself in a box EU Medical Device Regulations 2
P Basic UDI-DI and Legacy Devices EU Medical Device Regulations 19
C UDI for consumable/replacement components under the MDR EU Medical Device Regulations 2
K UDI for two units with different S/N on the same package/case Other US Medical Device Regulations 10
C Validation of process for releasing the UDI EU Medical Device Regulations 4
A MDR requirement where unit of use packaging is too small for UDI carrier EU Medical Device Regulations 1
S Does a refurbished product required a new UDI? US Food and Drug Administration (FDA) 3
A UDI for product already packed up EU Medical Device Regulations 0
M Brand owner or manufacturer responsible for Basic UDI-DI and UDI-DI? EU Medical Device Regulations 6
Aymaneh UDI-PI Software CE Marking (Conformité Européene) / CB Scheme 0
P UDI-PI requirements on reusable surgical device, do we need serialisation? ISO 13485:2016 - Medical Device Quality Management Systems 3
J UDI - Where to document it? EU Medical Device Regulations 8
N UDI: GS1 or HIBC ??? EU Medical Device Regulations 1
F UDI-PI required on packaging (MDR) EU Medical Device Regulations 4
C UDI Questions on Control Units Other US Medical Device Regulations 6
M Supplier requirements - Major supplier is a Non-Profit registered with ICCBBA (FDA UDI) Supply Chain Security Management Systems 12
K Two FDA questions regarding UDI (and potentially 803.52 MDR) Other US Medical Device Regulations 0
V UDI - OEM or OBL / Legal Manufacturer Responsibility CE Marking (Conformité Européene) / CB Scheme 0
A European UDI-DI CE Marking (Conformité Européene) / CB Scheme 2
N FDA UDI - Label vs. Labeling - Does the insert need to include UDI? Other US Medical Device Regulations 1
K Is UDI (Unique Device Identifier) symbol mandatory under EU MDR? EU Medical Device Regulations 3
M UDI procedure/ report 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
R Shall a new UDI-DI be required when stand-alone software device's version is updated? EU Medical Device Regulations 1
D Hospital Decontamination and the MDR - Technical Files, UDI and PMS EU Medical Device Regulations 3
F UDI - Unit of Use and Class I, single-use devices EU Medical Device Regulations 5
F UDI Registration - Who is responsible for Registration of the Device? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 12
M UDI Database and EUDAMED - Will there be two separate databases? EU Medical Device Regulations 4
Q Must product name be listed the same name in FURLS, UDI, GUDID and Company Website? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
Q LOT or Serial Number Symbol not used when the information is contained in the UDI? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
J Looking for a template SOP for UDI implementation for EU Medical Devices EU Medical Device Regulations 1
M Basic UDI and technical documentation EU Medical Device Regulations 1
A UDI and Design Controls - Labeling change via the Design Control process 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
M MDR Requirements - UDI/Barcodes/QR codes on IFUs EU Medical Device Regulations 4
M UDI transition period and implementation EU Medical Device Regulations 6
M UDI size on packaging EU Medical Device Regulations 3