It is a 510(k) exempt device only subject to device listing and GMP requirements.
Oops... I seem to have missed that point...
Actually, it doesn't make too much of a difference. For the 510(k) exemption to apply, you need to identify a device in commercial distribution, under the same classification, with similar characteristics. This means you first need to have a clear definition or scope for your own device (ie, what does it include etc.), which you are able to compare with another given device. So, does your scope (as well as the equivalent device's) include rechargeable batteries and a charger, or not? This should determine how you need to address those items.
It seems like you are saying:
1) If they are accessories, they are subject to UDI requirements. Now, obviously we do not manufacture the batteries and the chargers ourselves, then do we have to request that the labeling be changed by our suppliers or do we just slap on UDI labels onto the batteries and the charger? This seems a bit excessive.
It might look excessive but you must realise that even if you don't make those items in-house, or have them made to your order, you are still their responsible manufacturer (regulatory-wise) as accessories to a medical device, because you're the one who assigns to them their intended purpose as such and you're the one who places them as an integral part inside the immediate packaging of your device (the briefcase). In that they cease to be "just plain batteries and charger" and become something slightly different that potentially entails different responsibilities.
If you have an option to have them labeled to your requirements by the OEM that would seem more streamlined from the regulatory perspective. If not, adding "just UDI labels" should be carefully examined against the applicable regulations (once they are identified).
2) If they are an integral part of the device and not apply UDI to the batteries and chargers, then am I also exempt of listing them on the GUDID as well?
You're not expected to put UDI on device components, nor to register them on GUDID. If you go down this path I suggest not placing a UDI directly on the feedback device, and instead placing the lowest-level UDI in this case on the briefcase. I would also make sure that there is no intermediate packaging (eg plastic bags, extra boxes etc.) between the items and the briefcase. That would help clarify that the entire collection of items inside the briefcase is a single medical device and that the briefcase is the primary retail packaging. Then you could register the briefcase and contents as a single medical device on GUDID.
The above is my current view of this situation. Will it fly? I don't know. I think hardly any of us knows for sure, because UDI is such a new topic and none of us has a lot of field experience with class II devices (having gone past the implementation date - 24.9.16 - only recently).
Cheers,
Ronen.