I'm trying to compile an uncertainty budget for the calibration of our product that uses an externally produced sensor module and need to know if the manufacturer's specifications (i.e. temperature sensitivity, non-linearity, etc.) for errors can be used in the uncertainty budget or if we need to carry out our own tests and calculate these errors with our own traceable equipment.
dwperron
There are two types of uncertainty contributors you will use in your budgets. Type A contributors are the ones where you perform a statistical analysis of data you have collected. Examples of Type A contributors would be repeatability, reproducability, etc. Type B contributors are ones that you have not derived, but have found from another source. This would include manufacturer's specifications, data from handbooks, results from previous tests, etc. You do not need to confirm the manufacturer's specifications yourself, you can use them.
Be sure that you determine if they are reporting a tolerance or an uncertainty. If it is an uncertainty you will need to determine if they are reporting an expanded uncertainty (such as k=2) and divide by the k factor to get the uncertainty. Otherwise just use the uncertainty. It they report a tolerance you will need to normalize it to use in your calculations. Typically you would apply a rectangular distribution, dividing the tolerance by √3.
R Uncertainty Budget for CMM ISO 17025 related Discussions 3
C Should resolution be included in uncertainty budget for digital caliper or micrometer calibration? Measurement Uncertainty (MU) 5
F Learning Uncertainty Budget for Novice - Fluke 5520A Measurement Uncertainty (MU) 1
G What degrees of freedom means in uncertainty budget and how do I set the value Measurement Uncertainty (MU) 2
G Uncertainty Budget Examples for Caliper, Micrometer and Dial Gauge Measurement Uncertainty (MU) 3
N Include homogeneity uncertainty into budget table for thermocouples Measurement Uncertainty (MU) 1
M Uncertainty Budget for a Micrometer Measurement Uncertainty (MU) 5
J Uncertainty budget for IR thermometry at high T - 1000 - 1500?C Measurement Uncertainty (MU) 1
I UUT Accuracy as part of Uncertainty Budget Measurement Uncertainty (MU) 6
S Gage R&R as a wider aspect of uncertainty budget Measurement Uncertainty (MU) 3
E Uncertainty Budget with a mixture of kg and % Reading Measurement Uncertainty (MU) 4
I Uncertainty Budget examples for Multimeter, Calibrator, DC Voltage wanted Measurement Uncertainty (MU) 2
C Optical Comparator Uncertainty Budget Measurement Uncertainty (MU) 2
Q Uncertainty Budget for Potable Water Metering System Measurement Uncertainty (MU) 2
S Calculating an Uncertainty Budget for an Optical Comparator General Measurement Device and Calibration Topics 4
J Uncertainty Budget for the Optical Microscope - Deriving Linearity Uncertainty Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
M CMM Best Uncertainty - Type specific CMM Uncertainty Budget Measurement Uncertainty (MU) 3
S Measurement Uncertainty Budget (MU) of Ultrasonic Thickness gage? Measurement Uncertainty (MU) 4
G Uncertainty budget for industrial temperature controllers Measurement Uncertainty (MU) 2
J Measurement Uncertainty Budget for Comparative Dimensional Measurements Measurement Uncertainty (MU) 8
D Uncertainty Budget - Gage Blocks Measurement Uncertainty (MU) 7
M Contributing Factors in a Supermicrometer Measurement Uncertainty Budget? Measurement Uncertainty (MU) 8
L Creating Meaurement Uncertainty (MU) Budget for Infrared Thermometers Measurement Uncertainty (MU) 1
CalRich Measurement Uncertainty for Adjustable Thread Rings but No Budget Made Measurement Uncertainty (MU) 4
L % Uncertainty Budgets - Auditor wants to see my uncertainty budget represented as a % Measurement Uncertainty (MU) 5
R Measurement Uncertainty (MU) Budget for a 0-1 inch Micrometer Measurement Uncertainty (MU) 3
S Mercury in Glass Thermometer Uncertainty Budget Measurement Uncertainty (MU) 1
N QUERIES On ISO 17025 AUDIT - Measurement Uncertainty Budget, etc. ISO 17025 related Discussions 1
K Calibration Report Uncertainty Budget - Discussion and Spreadsheets Attached Measurement Uncertainty (MU) 11
G Calc uncertainty of angle; by measuring x,y points Measurement Uncertainty (MU) 0
CaliperJim Uncertainty and Accuracy Requirement Deviation Provision General Measurement Device and Calibration Topics 2
AllTheThings High complexity uncertainty examples Measurement Uncertainty (MU) 4
F Uncertainty not Required Measurement Uncertainty (MU) 3
C Uncertainty budgets and meeting 4:1 TUR for digital calipers and micrometer Measurement Uncertainty (MU) 0
L Uncertainty calculation for an average Measurement Uncertainty (MU) 1
F Missing value of Uncertainty B Measurement Uncertainty (MU) 10
F Issues in Uncertainty Estimation Measurement Uncertainty (MU) 5
F Operator Manual and Type B Uncertainty Measurement Uncertainty (MU) 3
S Uncertainty Budgets - Phase Noise (SSB), Harmonics Measurement Uncertainty (MU) 2
M Informational US FDA Final Guidance – Consideration of Uncertainty in Making Benefit-Risk Determinations in Medical Device Premarket Approvals, De Novo Classificati Medical Device and FDA Regulations and Standards News 0
E Discussion between co-worker on tolerance and uncertainty and how to apply it. Thoughts? 17025 ISO 17025 related Discussions 1
G 3 Operator Std. Dev. and uncertainty calculation Measurement Uncertainty (MU) 2