Uncertainty of Measurement for Calibration Laboratory?
amanbhai
It maybe a simple question!
We are a calibration lab and working for ISO/ IEC 17025:2005.
Now question is: Do we need to calculate the uncertainty of measurement for each new calibration equipment (we calibrate for our customer) or we can calculate it once and mention it for each same equipment we calibrate e.g. thermometer.
Any feedbaks are appriciated!
qusys
Re: Uncertainty of measurement for Calibration lab?
It should be done for each equipment to be calibrated, since its characteristic
amanbhai
But sometimes it is very difficult to calculate a single thermometer since it takes a lot of time in repeatability and reproduceability.
dv8shane
In my opinion if the UUT contributors are more than 5% of system uncertainty, you should calibrate each type individually with a separate uncertainty budget. You could categorize several models i.e. Fluke 3.5 digit thermometers under 1 budget if the proofs show that they biggest contributors are the repeatability and resolution of 0.1 degrees you could possibly then express the uncertainty as +/- X least significant digit(s) for the category. This helps to cut down on development time and internal costs.
BradM
Good posts on these already.
I have seen the spectrum, from someone calculating it once and posting as Best Measurement Capability (changed only when standards are calibrated); to estimating every calibration that is performed.
I think it depends on how stable your processes are, and how significant the variance is each time you verify it, and your customer requirements (if applicable).
Personally, I think the data should explain your answer. If you have enough history from... say.. every day it the data demonstrates stability, then extend to every week.
matwelli
but without this data....you have not truly checked the calibre of the device under test.
LIG thermometers need bore uniformity and hysteresis of the glass checked
Fully supported PRT's need hysteresis checked
Thermocouples need estimations for inhomogenity
what sort of thermometers are you calibrating ? we may be able to sugest a method that speeds up teh process and allows an easy way to estimate the uncertainty each time
if you wanted to do a "blanket" or "type test" uncertainty, the number would be quite large as you would have to allow for worst case on alot of contributions to the budget.
chriswhitehead
ARRRRR!! I've spent two weeks of my own time gleaning as much information I can about the dreaded uncertainty measurements and lots of practising. I can now get a figure for my certificates that looks like the correct figure but how can I be sure its right.
Any courses in the UK?
Any other sites for more practise examples?
adamt
I feel your pain as I was in the same boat years ago. The problem is there is different ways to come up with an uncertainty for a UUT. My best suggestion is to try to think of anything that would affect the measurement and write it out, add it together and expect if you missed something your auditor will have you include it in your budget. Make sure you look at the new ILAC-p14/2010 there are some new requirements that you may have to include in your calculations. Hope this helps
C MU (Measurement Uncertainty) in an Internal Calibration Laboratory Measurement Uncertainty (MU) 1
S MU (Measurement Uncertainty) for Oscilloscope Calibration Measurement Uncertainty (MU) 2
K New Gauges calibration certificate shows deviations from the measurement uncertainty IATF 16949 - Automotive Quality Systems Standard 5
A Calibration Method and Measurement Uncertainty General Measurement Device and Calibration Topics 5
B Length Measure - Interpreting Calibration Results and Measurement Uncertainty (MU) Measurement Uncertainty (MU) 3
R Signal Conditioning Card - Calibration Measurement Uncertainty Measurement Uncertainty (MU) 5
T Universal Testing Machine - Calibration and Measurement Uncertainty General Measurement Device and Calibration Topics 1
M Measurement Uncertainty in Extensometer Calibration Measurement Uncertainty (MU) 10
B Calculating Measurement Uncertainty in Oven Calibration (using thermocouples) Measurement Uncertainty (MU) 4
howste Calibration Certificate Contents: How to report Measurement Uncertainty? Measurement Uncertainty (MU) 18
M Measurement Uncertainty for Electrical Calibration Measurement Uncertainty (MU) 4
J 17025 Para. 5.6.2 - Calibration certificates - Measurement results - Uncertainty ISO 17025 related Discussions 6
C External calibration provider does not issue uncertainty of measurement Measurement Uncertainty (MU) 2
Hershal Measurement Uncertainty.....stirring up dust - MU variance for any given calibration Measurement Uncertainty (MU) 21
J Internal Laboratory Scope and Measurement Uncertainty (MU) IATF 16949 - Automotive Quality Systems Standard 1
B Measurement Uncertainty using Minitab Measurement Uncertainty (MU) 2
A Uncertainty Measurement - What this accuracy specification "1% + 10^5": means? General Measurement Device and Calibration Topics 2
Marc Measurement Uncertainty Discussion Sub-Forum General Measurement Device and Calibration Topics 0
C Measurement Uncertainty fluctuates half a tick mark (20 millionths of an inch) Measurement Uncertainty (MU) 1
R MU (Measurement Uncertainty) of a Tensile Tester (e.g. Zwick Roell) Measurement Uncertainty (MU) 4
H Uncertainty Measurement Calculation in Chemical Testing Measurement Uncertainty (MU) 1
T Method Repeatability for MU (Measurement Uncertainty) Measurement Uncertainty (MU) 1
S Combined Uncertainty of an Indirect Measurement Measurement Uncertainty (MU) 8
P Measurement Uncertainty and Equipment Accuracy Measurement Uncertainty (MU) 2
N Uncertainty calculation for a voltage measurement Measurement Uncertainty (MU) 3
T Measurement of uncertainty for standard mechanical properties General Measurement Device and Calibration Topics 1
A Calculating Combined Measurement Uncertainty - VDA 5 Measurement Uncertainty (MU) 1
T Uncertainty of Measurement vs. Tolerance Measurement Uncertainty (MU) 5
Geoff Cotton Measurement Uncertainty (MU) of HR-150A Rockwell Hardness Tester Measurement Uncertainty (MU) 1
Jerry Eldred When Does Resolution of Measurement Standard Apply in Uncertainty Calculations General Measurement Device and Calibration Topics 14
E Uncertainty Measurement for Water COD, TDS, TSS & pH Measurement Uncertainty (MU) 1
D General Procedure for Uncertainty Measurement Measurement Uncertainty (MU) 2
S Measurement Uncertainty for Neutral Salt Spray test to BS EN ISO 9227:2006 General Measurement Device and Calibration Topics 4
S Organizing Uncertainty Budgets for different Measurement Equipment Measurement Uncertainty (MU) 4
R Procedure for Measurement Uncertainty Estimation for Electrical Measurements Measurement Uncertainty (MU) 2
K AATCC 81 pH test - How to calculate MU (Measurement Uncertainty) Measurement Uncertainty (MU) 2
optomist1 Definition Measurement Uncertainty - Standard definitions for Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 8
S Uncertainty Estimation in Temperature Measurement by K type Thermocouples Measurement Uncertainty (MU) 7
DietCokeofEvil ILAC-g8:03/2009 - Reporting Results with Measurement Uncertainty General Measurement Device and Calibration Topics 7
N Rounding the Numerical Value of the Measurement Uncertainty Measurement Uncertainty (MU) 2