Understanding DC Patient Leakage Failure from 0.3VDC

michael Cejnar

Involved In Discussions
#1
Can someone help me understand DC leakage current limits.

We have a computer display in the Patient Area. We changed form Mains AC to 12VDC SELV power to enhance safety. However the long DC power cable drops 0.2V along the ground cable, floating the display's functional earth, exposed at the VESA mount, to 0.2VDC above mains ground.

Now, measured Touch Current of 200uADC (~0.2V/1KOhm) exceeds the limit <100uA. Also by Cl 4.6, we have to comply with Type B Patient Applied part leakage limit (Operator could touch VESA mount and patient), which is 100uA for AC, but a minute 10uA for DC !

Clearly, electrocution by 0.2V is hardly terror inspiring. In fact 0.2V can not stimulate the heart in any way.

The standard explains on P254 that "PATIENT AUXILIARY CURRENT…Lower values are given for d.c. to prevent tissue necrosis with long-term application.". and on P252 "PATIENT LEAKAGE CURRENT…The 50 μA current allowed in SINGLE FAULT CONDITION is not likely to result in a current density sufficient to stimulate neuromuscular tissues nor, if d.c., cause necrosis."

Also, the risk of VF from DC is 1/5 that for AC; on p250 the standard says: "e) The RISK [of VF] is highest and approximately equal for frequencies in the 10 Hz to 200 Hz range. It [risk of VF] is lower, by a factor of nearly 5, at d.c.".

So, for risk of VF from DC current, 5 times the leakage current limits might be more applicable; the imposed 1/10 limit, being for risk of necrosis, surely should be reserved for the relatively rare chronic application of general MEE.

Anyway, my immediate problem is: can I rationalise this 15x over limit DC leakage current from exposed 0.2VDC, by stating it is from brief contact with 0.2VDC? Ho do others handle it - surely this must be a fairly common problem, since an 11mV drop on a DC power cable will fall foul of the standard. Or do I have to spend money trying to insulate the 0.2VDC on the VESA mount into a Type BF part?
 
Elsmar Forum Sponsor

Peter Selvey

Staff member
Super Moderator
#3
First to address the theory side: 0.2Vdc is absolutely safe in this situation. The body approximates 1kohm impedance only for higher voltages (around 100V), for very large, wet, contact surface areas (~100cm2) or if contact is made directly with the blood stream (e.g. catheters). Experiments with 0.1V found that it was possible to get ~100uA but it required two wet hands firmly pressed face down on aluminium foil, hardly a normal situation.

Next to address the standard:
Under Clause 8.4.2 c) is it possible to have exposed areas in medical electrical equipment up to 60Vdc, provided that risk of those parts being brought into contact with the patient (either directly or indirectly) are acceptable. Although the standard refers to the probability of contact, the final judgement should be based on the probability of harm. Noting that for 0.2V you would need large contact areas for significant current to flow, and that for dc you need long duration for harm, it's pretty clear that the risk is acceptable.
 

michael Cejnar

Involved In Discussions
#4
Thanks Peter
I concur with the 0.2V As a clinical electrophysiologist, I know even with a pacing electrode in the heart, 0.2V is well below the chronaxie for any kind of stimulation.

Yes, you have quoted 8.4.2 c) before, but strictly speaking it specifies a list of exempted parts, which do not include a simple exposed part of an enclosure in question. But if this is the closest exemption, then so be it.

While I have found an easy way to insulate the surface just to be sure, the issue is nevertheless intriguing.
 
Last edited:
Thread starter Similar threads Forum Replies Date
Q Understanding "free movement" for CE-marked products EU Medical Device Regulations 2
D High level understanding of EUDAMED EU Medical Device Regulations 3
A 60601-1 understanding of stationary device IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
N Help with basic understanding of Competence requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
D Surface Roughness understanding Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
P Understanding DFMEA and PFMEA - Supplier Related IATF 16949 - Automotive Quality Systems Standard 21
DuncanGibbons Understanding the applicability of Design of Experiments to the IQ OQ PQ qualification approach Qualification and Validation (including 21 CFR Part 11) 5
B Measuring and monitoring equipment - Understanding which procedures to be compliant with ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 6
M Informational Health Canada has launched an e-Learning tool to aid in understanding the premarket regulatory requirements for medical devices in Canada Medical Device and FDA Regulations and Standards News 0
S Understanding UDI requirements - Class 2 medical device (hearing aids) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M Informational Understanding Costs And Risks For HFE Usability Studies — Part 1: Testing In-House Medical Device and FDA Regulations and Standards News 0
P Understanding FDA draft "Management of Cybersecurity in Medical Devices" Medical Information Technology, Medical Software and Health Informatics 3
J Properly understanding SPC - Newbie SPC questions Statistical Analysis Tools, Techniques and SPC 34
S Understanding control chart and measurement capability Statistical Analysis Tools, Techniques and SPC 2
P Minitab Data Analysis - Understanding if a Process is in Control or Not Using Minitab Software 2
R Understanding a few points on ISO 9001's Design and Development Planning ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
Z Understanding Cycle Time - Why the time of the other activities are left out Lean in Manufacturing and Service Industries 11
J Understanding ISO 9001:2015 - 10.3 Continual Improvement ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
J Understanding ISO9001:2015 - 8.3: Design and Development of Products and Services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
E Root Cause Analysis - Is Insufficient Understanding an acceptable Root Cause? General Auditing Discussions 9
E Understanding of TS 16949 Clause 7.6.2 IATF 16949 - Automotive Quality Systems Standard 5
K Understanding IEC 60601-2-68 requirements ISO 13485:2016 - Medical Device Quality Management Systems 1
A Training material for interpretation & understanding Part 11 requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
N Understanding the absolute uncertainty specification for a Fluke 5500A Measurement Uncertainty (MU) 3
N Understanding, Challenging & Approving Supplier Control Plans FMEA and Control Plans 7
M Definition Recommendations - Understanding "recommendations" and "recommended corrective action" Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 8
S Understanding UDI (Unique Device Identification) Other US Medical Device Regulations 10
T Understanding USP <1112> Water Activity as applicable to Medical Devices Other Medical Device and Orthopedic Related Topics 4
K Understanding Risk Management Requirements according to AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
S MIL-HDBK-217 - Understanding the various Environmental Conditions Reliability Analysis - Predictions, Testing and Standards 1
D What is your understanding or interpretation of TS16949 7.4.1.2 IATF 16949 - Automotive Quality Systems Standard 6
C Understanding the relationship between 62304 and the MDD ER IEC 62304 - Medical Device Software Life Cycle Processes 7
S Understanding Subgroup Size - Multi Cavity (Minitab) Statistical Analysis Tools, Techniques and SPC 4
R Understanding clause 15.4.2.1 d) of amendment 1:2012? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M Understanding accreditation, MoUs, certifications Other ISO and International Standards and European Regulations 28
L Mobile Medical App - Understanding 21 CFR Part 820 Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
D Understanding and implementing ISO 17025 ISO 17025 related Discussions 9
M Understanding Versions of Collateral and Particular Standards IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
S Understanding, Analysis and Monitoring Quality Defects on Composite Components Statistical Analysis Tools, Techniques and SPC 3
S Understanding PMS (Post Market Surveillance) and PMCF (Vigilance and PMCF) Quality Manager and Management Related Issues 1
B Understanding why my CpK and PpK are low, and LCL Statistical Analysis Tools, Techniques and SPC 20
S Understanding Quality Objectives, Metrics and KPI ISO 13485:2016 - Medical Device Quality Management Systems 15
Q Beginner's Understanding - The Purpose and Applications of QMS/ISO Standards Philosophy, Gurus, Innovation and Evolution 12
Q Understanding Configuration Management AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
W Understanding PPAP Appearance Approval APQP and PPAP 22
V Understanding Automotive Coating for Seating Mechanism Components Manufacturing and Related Processes 1
M Understanding of Regression and ANOVA in Minitab Statistical Analysis Tools, Techniques and SPC 8
4 Understanding ILAC policy P14:12/2010 6.3 part a) General Measurement Device and Calibration Topics 28
H Understanding 8.2.3 M&M of Processes for our Internal Audit ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
P Understanding ISO 26262 Road Vehicle Functional Safety Other ISO and International Standards and European Regulations 2

Similar threads

Top Bottom