Understanding IEC 60601-1:2005, Clause 8.1 b) - Short Circuiting Double Insulation
P
petruchio
Hi
IEC 60601-1:2005, Clause 8.1 b) requires " ... short circuiting of any constituent part of double insulation ..."
Can anyone explain this to me in plain English? The term "short circuit" applied to insulation just doesn't make sense
I can understand short circuiting creepages and clearances, one literally connects them but that's simply not possible for double insulation.
Standards are like instructions manuals - they only make sense when interpreted by someone else. Kinda makes them useless .... :-(
petruchio
Marcelo
Re: Understanding IEC 60601-1:2005, Clause 8.1 b) - Short Circuiting Double Insulatio
Hi
Double insulation is
insulation comprising both BASIC INSULATION and SUPPLEMENTARY INSULATION
So you need to shot circuit the basic or/and supplementary insulation that are part of the double.
Creepage and clearance are parts of the insulation, so you can short circuit them.
P
petruchio
Thanks but you've missed my point ... a short circuit across a creepage or clearance is obvious since these distances separate conductors by air or basic insulation (PCB usually). However, for double insulation, if one constituent is "short circuited" (ie effectively removed) it still leaves the other constituent which is assumed to still provide insulation since the standard only allows for one fault condition at a time. By definition, the fault condition must never create a safety hazard.
What is my argument missing?
Marcelo
Hum, maybe I´m still note sure what your point is, but...
The other constituent part will only really provide the insulation IF it´s tested to verify that it will. Hence the fault condition of one part being short-circuited - to verify if the other does maintain it´s insulation.
P
petruchio
Hi
Thanks for your feedback ...
I'm going through a 60601-1 checklist, hence this particular question. Any insulation, providing that it is qualified (eg via IEC60227..) must be assumed to be good. So if one has to consider a SFC in which one part of the double insulation fails completely leaving only the other part then this remaining part has to be assumed to be good since the standard allows only for single failures.
My thinking suggests that this test must always be passed by definition. Clause 4.8 allows this 'test' to be performed by inspection unless failure is likely to lead to unacceptable risk, in which case 4.9 has to be applied.
Marcelo
IEC 60227 is a component standard. I´m not sure why you think that anything in conformity with IEC 60227 or any other means that the insulation is good.
Insulation is good for IEC 60601 only if it passes 8.8 Insulation.
4.8 is applicable only to components, not to insulation. Although you can use components a part of the insulation, it does not mean that a "certified" component, for example, pass insulation requirements of IEC 60601.
P
petruchio
Components includes wiring, PCBs ... so if these parts meet 60601-1 and/or applicable normatives then they have to be assumed to hold up in the event of the failure of the second part of the double insulation.
Marcelo
Well, not sure from where you having those conclusions, but let me try again.
Your original question was related to 8.1 b), which says:
SINGLE FAULT CONDITION includes:
1 - short circuit of any one insulation that complies with the requirements for one MEANS OF PROTECTION as specified in 8.8;
NOTE This includes short circuiting of either constituent part of DOUBLE INSULATION that complies with 8.8.
2 – .....
3 – short circuit and open circuit of any component other than a COMPONENT WITH HIGH INTEGRITY CHARACTERISTICS that is connected in parallel with insulation, with an AIR CLEARANCE or with a CREEPAGE DISTANCE unless shorting can be shown not to be a failure mode for the component (see also 4.8 and 4.9);
4.8 and 4.9 apply for the third option.
There´s no link between the first and third option. For the fist option you have to evaluate 8.8 Having a component in conformity with IEC XXXXX means nothing to the first option because it is related to insulation, not components.
P
petruchio
OK, thanks. I'll take this up with our notified body, it still makes no sense to me.
Peter Selvey
Shorting one part of double insulation is a valid fault condition, but it should result in obvious compliance, provided that each part of the insulation meets dielectric strength / creepage / clearance as per 8.8 and 8.9.
In Clause 8.7 (leakage current), shorting of one part of double insulation is excluded (see 8.7.2 3rd dash). Thus, there is no reason to actually short one part of double insulation in practice.
But it is still necessary to mention it in 8.1 (b) because the requirements in 8.8 and 8.9 are derived from this premise. In other words, dielectric strength, creepage and clearance limits for each part of double insulation are derived from the assumption that one insulation is shorted in SFC.
K Understanding IEC 60601-2-68 requirements ISO 13485:2016 - Medical Device Quality Management Systems 1
Q Understanding "free movement" for CE-marked products EU Medical Device Regulations 2
D High level understanding of EUDAMED EU Medical Device Regulations 3
A 60601-1 understanding of stationary device IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
N Help with basic understanding of Competence requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
D Surface Roughness understanding Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
P Understanding DFMEA and PFMEA - Supplier Related IATF 16949 - Automotive Quality Systems Standard 21
DuncanGibbons Understanding the applicability of Design of Experiments to the IQ OQ PQ qualification approach Qualification and Validation (including 21 CFR Part 11) 5
B Measuring and monitoring equipment - Understanding which procedures to be compliant with ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 6
M Informational Health Canada has launched an e-Learning tool to aid in understanding the premarket regulatory requirements for medical devices in Canada Medical Device and FDA Regulations and Standards News 0
S Understanding UDI requirements - Class 2 medical device (hearing aids) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M Informational Understanding Costs And Risks For HFE Usability Studies — Part 1: Testing In-House Medical Device and FDA Regulations and Standards News 0
P Understanding FDA draft "Management of Cybersecurity in Medical Devices" Medical Information Technology, Medical Software and Health Informatics 3
J Properly understanding SPC - Newbie SPC questions Statistical Analysis Tools, Techniques and SPC 34
S Understanding control chart and measurement capability Statistical Analysis Tools, Techniques and SPC 2
P Minitab Data Analysis - Understanding if a Process is in Control or Not Using Minitab Software 2
R Understanding a few points on ISO 9001's Design and Development Planning ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
Z Understanding Cycle Time - Why the time of the other activities are left out Lean in Manufacturing and Service Industries 11
J Understanding ISO 9001:2015 - 10.3 Continual Improvement ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
J Understanding ISO9001:2015 - 8.3: Design and Development of Products and Services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
E Root Cause Analysis - Is Insufficient Understanding an acceptable Root Cause? General Auditing Discussions 9
E Understanding of TS 16949 Clause 7.6.2 IATF 16949 - Automotive Quality Systems Standard 5
A Training material for interpretation & understanding Part 11 requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
N Understanding the absolute uncertainty specification for a Fluke 5500A Measurement Uncertainty (MU) 3
N Understanding, Challenging & Approving Supplier Control Plans FMEA and Control Plans 7
M Definition Recommendations - Understanding "recommendations" and "recommended corrective action" Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 8
S Understanding UDI (Unique Device Identification) Other US Medical Device Regulations 10
T Understanding USP <1112> Water Activity as applicable to Medical Devices Other Medical Device and Orthopedic Related Topics 4
K Understanding Risk Management Requirements according to AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
S MIL-HDBK-217 - Understanding the various Environmental Conditions Reliability Analysis - Predictions, Testing and Standards 1
D What is your understanding or interpretation of TS16949 7.4.1.2 IATF 16949 - Automotive Quality Systems Standard 6
C Understanding the relationship between 62304 and the MDD ER IEC 62304 - Medical Device Software Life Cycle Processes 7
S Understanding Subgroup Size - Multi Cavity (Minitab) Statistical Analysis Tools, Techniques and SPC 4
R Understanding clause 15.4.2.1 d) of amendment 1:2012? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M Understanding accreditation, MoUs, certifications Other ISO and International Standards and European Regulations 28
L Mobile Medical App - Understanding 21 CFR Part 820 Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
D Understanding and implementing ISO 17025 ISO 17025 related Discussions 9
M Understanding Versions of Collateral and Particular Standards IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
S Understanding, Analysis and Monitoring Quality Defects on Composite Components Statistical Analysis Tools, Techniques and SPC 3
S Understanding PMS (Post Market Surveillance) and PMCF (Vigilance and PMCF) Quality Manager and Management Related Issues 1
B Understanding why my CpK and PpK are low, and LCL Statistical Analysis Tools, Techniques and SPC 20
S Understanding Quality Objectives, Metrics and KPI ISO 13485:2016 - Medical Device Quality Management Systems 15
Q Beginner's Understanding - The Purpose and Applications of QMS/ISO Standards Philosophy, Gurus, Innovation and Evolution 12
Q Understanding Configuration Management AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
W Understanding PPAP Appearance Approval APQP and PPAP 22
V Understanding Automotive Coating for Seating Mechanism Components Manufacturing and Related Processes 1
M Understanding of Regression and ANOVA in Minitab Statistical Analysis Tools, Techniques and SPC 8
4 Understanding ILAC policy P14:12/2010 6.3 part a) General Measurement Device and Calibration Topics 28
H Understanding 8.2.3 M&M of Processes for our Internal Audit ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
P Understanding ISO 26262 Road Vehicle Functional Safety Other ISO and International Standards and European Regulations 2