Understanding Requirements for CE Marking and Test Requirements
O
oemdeveloper
We are a manufacturer of computers used in OEM and specialty applications. Most of the systems we build are not certified (UL, FCC, etc.) but we do build several medical devices, which are certified, so we are quite familiar with third party testing.
We have clients that are going to be selling their products (non-medical) in Europe and will clearly require a CE mark.
Every 3rd party testing house we have spoken with has stated that we need to have at least two tests done on the systems; EMC and safety (low voltage directive). Then we will have the necessary documents to put in our technical file, which we will still have to create. The cost for this testing is always right around $15,000.
We came across an alternative plan in the form of Barclay-Phelps. We contacted them and paid a small fee to have them review our application. Their response was that our systems would NOT require 3rd party testing and that for a very reasonable cost they could put the necessary paper work in order for us and we would be all set. (We build these systems from motherboards, drives and the like, not descrete components. These parts have FCC, CE and other certifications.)
We have been unable to validate that this is a legitimate way to go. Barclay won't explain further how their methodology works and contacts at 3rd part testing houses continue to say Barclay-Phelps' method is bunk.
The testing guys all have a stake in this game - $15,000 or more in testing. I don't trust their answers. The Barclay approach (cemarking.com is another in the same vein) seems incomplete. What is the right, and cost effective path?
W
w_grunfeld
In my opinion the testing guys are right (almost) you have to test your equipment, not to the Directives but to harmonised standards listed in the respective directives.
For instance you are required to test against EN 60950 -Safety of Information Technology Equipment Including Electrical Business Equipment, which provides "presumption of conformance" with the LVD. In my opinion, the Directive itself does not mandate third party test house but in practical terms it is not likely that you have in-house all the required equipment, know-how, calibration certificates etc. so your TEST CERTIFICATE and TEST REPORT pass the scrutiny of authorities.(in case they'll want to review). Likewise, EMC tests require special anechoic chambers that you are unlikely to have in-house. So the bottom line is that CE in your case is based on self-declaration but it must be based on valid test results and certificates.
Ronen E
Problem Solver
Moderator
The last sentence I left in the quote above left me a bit puzzled. it sounds as if those clients fall under the OEM domain you described above, i.e. your products would be first integrated into theirs, then *they* would be placing them on the EC market - is that correct? If so, your products do not require CE marking per-se. It would probably be a very handy added value for your clients (and might also become a contractual obligation if you so agree in the contract), but from a regulatory standpoint it is not a requirement. It would be your clients' responsibility to meet all applicable requirements (testing and other) with *their* products, including the components they get from you, in finished device form.
Second, as a general note, I wouldn't trust any non-1000% transparent consultancy firm, e.g. one that doesn't do *everything* in its capacity to have me informed and satisfied with regards to any and all of their processes, rationales and practices. Just to clarify, I have no affiliation whatsoever with any electrical test laboratory.
Cheers,
Ronen.
PS By your phrasing above, I assume you meant those clients are located outside Europe and selling into the EC.
O
oemdeveloper
Thank you for your comments. Ronen, you have touched on something that has occured to me; why not let our customer deal with the CE issue? I've always assumed that we would handle any needed certifications because that is the way historically our customer's have preferred it. If our client is comfortable with Barclay-Phelps approach and puts the necessary documentation into place, we would only be putting CE marks on the system at thier request. Our name is not on the product at all so all responsibity would lie with our customer.
I
imajeenyus
I'm currently going through the process of getting a small electronic product CE marked (I'm in the UK) and, at the moment, I'm pursuing Barclay-Phelps to see what they can offer. Like yourself, I was a bit mystified as to how they can avoid the need for third-party testing, since everything I've read indicates this is an absolute requirement. I'm filling out the initial questionnaire, but I was wondering if you could share any more of your experiences with B-P? Did you go with them eventually for CE marking or did you use a testing house? What sort of job did they do?
It seems impossible to find any information about who they are (you'd think they'd at least have an "About us" page on their site).
Would be very grateful for any comments!
Best regards,
Lindsay Wilson (Mr.)
I
Ima Mauer
I'm not sure if this is the place to ask a new question but here goes:
We are looking at CE marking for a class I medical device -not sterile and does not measure. This is my 1st time working with CE. My understanding is we can self declare by completing the required documents. My questions are what testing is required and how long does that usually take if required. What is the timing from submitting documentation to getting the approval to apply a CE Mark? Do I need to register with a 3rd party. Any help or commnet is appreciated...Thanks
Ronen E
Problem Solver
Moderator
Hello and welcome to the field :bigwave:
Have a read:
Ronen E
Problem Solver
Moderator
In short:
- Required tests and their duration would depend on the nature of your device.
- You don't need to "submit documentation" and "get approval to apply a CE mark" for a non-sterile non-measuring class I device. Once you comply with all the applicable requirements, you're good to go (self-certification).
- You will have to register your class I device(s) with one of the Competent Authorities in the EC. Any manufacturer who doesn't have a registered place of business in one of the EC member states has to nominate an EC authorized representative; this EC rep will handle the registration on your behalf. Once they submit you should get the confirmation in a month or less.
Cheers,
Ronen.
W
w_grunfeld
Ronen,
I agree with the first part of your answer but the second part contradicts the first. The CE mark in this case being based on self declaration, what "Authorities" do you refer to? There are no authorities involved. EU representative is a different matter altogether. It can be the importer or anyone they appoint to play this role
Ronen E
Problem Solver
Moderator
Ronen,
MDD:
Article 14
Registration of persons responsible for placing devices on the market
1. Any manufacturer who, under his own name, places devices on the market in accordance with the procedures referred to in Article 11 (5) and (6) and any other natural or legal person engaged in the activities referred to in Article 12 shall inform the competent authorities of the Member State in which he has his registered place of business of the address of the registered place of business and the description of the devices concerned.
2. ... For devices referred to in the first subparagraph of paragraph 1, the authorised representative shall inform the competent authority of the Member State in which he has his registered place of business of the details referred to in paragraph 1.
(Emphasis added)
Article 11
Conformity assessment procedures
5. In the case of devices falling within Class I, other than devices which are custom-made or intended for clinical investigations, the manufacturer shall, in order to affix the CE marking...
N Help with basic understanding of Competence requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
M Informational Health Canada has launched an e-Learning tool to aid in understanding the premarket regulatory requirements for medical devices in Canada Medical Device and FDA Regulations and Standards News 0
S Understanding UDI requirements - Class 2 medical device (hearing aids) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
K Understanding IEC 60601-2-68 requirements ISO 13485:2016 - Medical Device Quality Management Systems 1
A Training material for interpretation & understanding Part 11 requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
K Understanding Risk Management Requirements according to AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
L Mobile Medical App - Understanding 21 CFR Part 820 Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M Learning ISO 13485 - Getting a better understanding of the requirements ISO 13485:2016 - Medical Device Quality Management Systems 6
M Understanding AS9100 Clause 7.5.1.4 Post-Delivery Support Requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
B Correct my understanding about Design Requirements Design and Development of Products and Processes 6
J Assessing the understanding of occupational health and safety requirements Occupational Health & Safety Management Standards 3
L Understanding FDA requirements for Software Validation Qualification and Validation (including 21 CFR Part 11) 7
C Costs of not understanding customer requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
E Contract manufacturer seeking help understanding ISO 13485 Validation requirements ISO 13485:2016 - Medical Device Quality Management Systems 9
Q Understanding "free movement" for CE-marked products EU Medical Device Regulations 2
D High level understanding of EUDAMED EU Medical Device Regulations 3
A 60601-1 understanding of stationary device IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D Surface Roughness understanding Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
P Understanding DFMEA and PFMEA - Supplier Related IATF 16949 - Automotive Quality Systems Standard 21
DuncanGibbons Understanding the applicability of Design of Experiments to the IQ OQ PQ qualification approach Qualification and Validation (including 21 CFR Part 11) 5
B Measuring and monitoring equipment - Understanding which procedures to be compliant with ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 6
M Informational Understanding Costs And Risks For HFE Usability Studies — Part 1: Testing In-House Medical Device and FDA Regulations and Standards News 0
P Understanding FDA draft "Management of Cybersecurity in Medical Devices" Medical Information Technology, Medical Software and Health Informatics 3
J Properly understanding SPC - Newbie SPC questions Statistical Analysis Tools, Techniques and SPC 34
S Understanding control chart and measurement capability Statistical Analysis Tools, Techniques and SPC 2
P Minitab Data Analysis - Understanding if a Process is in Control or Not Using Minitab Software 2
R Understanding a few points on ISO 9001's Design and Development Planning ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
Z Understanding Cycle Time - Why the time of the other activities are left out Lean in Manufacturing and Service Industries 11
J Understanding ISO 9001:2015 - 10.3 Continual Improvement ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
J Understanding ISO9001:2015 - 8.3: Design and Development of Products and Services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
E Root Cause Analysis - Is Insufficient Understanding an acceptable Root Cause? General Auditing Discussions 9
E Understanding of TS 16949 Clause 7.6.2 IATF 16949 - Automotive Quality Systems Standard 5
N Understanding the absolute uncertainty specification for a Fluke 5500A Measurement Uncertainty (MU) 3
N Understanding, Challenging & Approving Supplier Control Plans FMEA and Control Plans 7
M Definition Recommendations - Understanding "recommendations" and "recommended corrective action" Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 8
S Understanding UDI (Unique Device Identification) Other US Medical Device Regulations 10
T Understanding USP <1112> Water Activity as applicable to Medical Devices Other Medical Device and Orthopedic Related Topics 4
S MIL-HDBK-217 - Understanding the various Environmental Conditions Reliability Analysis - Predictions, Testing and Standards 1
D What is your understanding or interpretation of TS16949 7.4.1.2 IATF 16949 - Automotive Quality Systems Standard 6
C Understanding the relationship between 62304 and the MDD ER IEC 62304 - Medical Device Software Life Cycle Processes 7
S Understanding Subgroup Size - Multi Cavity (Minitab) Statistical Analysis Tools, Techniques and SPC 4
R Understanding clause 15.4.2.1 d) of amendment 1:2012? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M Understanding accreditation, MoUs, certifications Other ISO and International Standards and European Regulations 28
D Understanding and implementing ISO 17025 ISO 17025 related Discussions 9
M Understanding Versions of Collateral and Particular Standards IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
S Understanding, Analysis and Monitoring Quality Defects on Composite Components Statistical Analysis Tools, Techniques and SPC 3
S Understanding PMS (Post Market Surveillance) and PMCF (Vigilance and PMCF) Quality Manager and Management Related Issues 1
B Understanding why my CpK and PpK are low, and LCL Statistical Analysis Tools, Techniques and SPC 20
S Understanding Quality Objectives, Metrics and KPI ISO 13485:2016 - Medical Device Quality Management Systems 15
Q Beginner's Understanding - The Purpose and Applications of QMS/ISO Standards Philosophy, Gurus, Innovation and Evolution 12