Understanding the Test Method of Defibrillation

S

sankui

#1
Hi all,

In 3rd edition, the standard requires that except for PERMANENTLY INSTALLED ME EQUIPMENT, the ME EQUIPMENT is to be tested with and without the PROTECTIVE EARTH CONDUCTOR connected.

However, the rationale of 8.5.5.1 states that NORMAL USE includes the situation that a PATIENT is defibrillated while connected to the EQUIPMENT and, at the same time, the OPERATOR or another person is in contact with the ENCLOSURE. The possibility of this occurring at the same time as the SINGLE FAULT CONDITION of a defective PROTECTIVE EARTH CONNECTION is very unlikely and is therefore disregarded.

From mentioned above (especially the phrase underlined), it seems that the rationale conflict with the requirement. I am confused that whether the PROTECTIVE EARTH CONDUCTOR interrupted when conducting the defibrillation test for class I equipment (NON-PERMANENTLY INSTALLED).

Please kindly advise it. Thanks in advance.
 
Elsmar Forum Sponsor

Peter Selvey

Staff member
Super Moderator
#2
Hi Sankui,

Looks like you found a technical mistake in the standard!

But, in practice it is not so much of an issue.

The most critical equipment for this test are patient monitors, which are normally battery operated and hence need to be tested without an earth anyway.

Secondly, most equipment has a Type BF/CF isolation barrier which can withstand a 5kV defib pulse without breakdown nor letting any significant energy across.

Some manufacturers bridge this isolation barrier with a spark gap, which can breakdown from a 5kV pulse. However, the spark gap is normally rated for at least 3kV to avoid breakdown in the 1.5kVrms test (technically, a spark gap should be removed, but for convenience e.g. in production, it normally remains in circuit).

Modern biphasic defibrillators typically use much lower voltages less than 3kV, which means breakdown is unlikely in practice.

So, for a potential hazardous situation to occur we would need (a) earth open, (b) a spark gap to be fitted, and (c) the spark gap is rated less than the defib pulse.

Basically, it won't happen.

So, it's safe, but not because of the reason in the standard. It could be be dangerous situation for a Type B defib proof applied part, but I've never seen a such a rating in practice.
 
S

sankui

#3
Hello peter,

Thanks for your technical explanation. I totally agree with your point. However, the standard allows DEFIBRILLATION-PROOF TYPE B APPLIED PART, though we have not seen such device so far.

Therefore, I believe that a mains operated device without internal power source (e.g. battery) shall be tested with earth (the rationale clearly shows that interruption of earth is unlikely while conducting the defib test).
 

Peter Selvey

Staff member
Super Moderator
#4
Yes, but the standard rationale does not fit with basic risk management.

Statistics taken in Australian hospitals show that around 1-2% of earths are open at any time, and that is in a fairly controlled environment with regular service tests. The moderately high result is due to the heavy mechanical abuse cords and connectors get in hospitals from equipment being moved around all the time.

Defibrillation is considered normal condition. We can expect equipment in critical care environment to be exposed to several defib events / year.

So, we have an overall probability (frequency) of ~0.1 events/year/device that a defib pulse will occur when the earth is open.

In the world of safety, a probability of 0.1 is quite high, even for low to moderate severity of harm.

I would suggest that it is worth testing, but we can predict most equipment will pass due to BF/CF isolation barrier.
 
S

sankui

#5
Hi Peter,

How do you calculate the overall probability? The value of 0.1 is really high.
 

Peter Selvey

Staff member
Super Moderator
#6
Actually there is a bit of a mistake in ISO 14971 to use probability as the basis of risk.

The correct parameter is frequency of (adverse) events.

When the frequency is << 1 (e.g. 0.1 or less), probability and frequency are usually the same. But, when frequency is high relative to the period of analysis (e.g. several times a year), it is no longer appropriate to use probability for estimates of risk.

Compare the following calculations in probability (P) and frequency (F):

Defib events, per year: Pd = 1 (100%); Fd = 5
Broken earth, per device: Pe = 0.02; Fe = 0.02

Combined event (defib + open earth):
Pc = Pd x Pe = 0.02 events / year
Fc = Fd x Fe = 0.1 events / year

These are quite different results. But I can assure you the frequency based result is the correct one.

The alternate would be to calculate on a shorter time span (e.g. per day or per use), in which case both probabilty and frequency would yeild the same results.

But from a manufacturer's and the operator's point of view, number of events per year is resonable basis on which to make decisions.
 

Roland chung

Trusted Information Resource
#7
I would think it is not suitable to regard the frequency as the basis of risk.

The frequency is a random variable. Unless the frequency is small enough, the results shall be very different if one uses frequency to calculate the occurrence of events. Based on your example, I would assume that:

1) Defib events/ year: Fd = 5; Broken earth/ device: Fe = 0.02 --> Fc = Fd x Fe = 0.1 events/ year

2) Defib events/ year: Fd = 50; Broken earth/ device: Fe = 0.02 --> Fc = Fd x Fe = 1 events/ year

For case 2), defib event assumed to be 50 times per year may be reasonable in large hospital (large hospitals have more patients). The result (1 events/ year) means that the defib test with defective earth is an inevitable event.

This is clearly not correct. It shall be overestimated. The probability of the real world is 0.02 events / year.
 

Peter Selvey

Staff member
Super Moderator
#8
By frequency I mean the average rate of occurrence, which is not random.

Probabilities are also based on the average rate of occurrence, but by definition, a probability can never exceed 1, whereas frequency can and often does exceed 1, especially over longer periods like 1 year.

Of course, it is not expected that the final probability (or frequency) of harm will ever be close to 1 in risk management.

However, harm is often the result of a sequence of events, and individual events in that sequence may have a frequency exceeding 1. In that case, calculations should be based on frequency not probability.

Calculations are normally on a per device basis which will be the same regardless of the size of the hospital. And in a large hospital we can expect these events to be inevitable regardless the method of calculation.

If we have a hospital with 100 devices, each device gets a defib pulse 5 times a year on average, and on average 2% of devices have open earths, using probability you would get 2 events per year (defib + open earth), using frequency would be 10 events per year. Either way, it's high and testing (defib + open earth) is warranted. But it should be clear the value of 10 events / year is the correct one on which risk decisions should be based.

In this particular case, even if we have a defib pulse + broken earth, harm still requires a number of other unlikely events, including most significantly, breakdown of the patient isolation barrier.

If the equipment is properly designed for Type BF/CF (4mm / 1.5kVrms), the probability of breakdown is extremely low. So although defib + broken earth is an inevitable event, it's not something that is expected to cause harm.
 
Thread starter Similar threads Forum Replies Date
O Understanding Requirements for CE Marking and Test Requirements CE Marking (Conformité Européene) / CB Scheme 12
Q Understanding "free movement" for CE-marked products EU Medical Device Regulations 2
D High level understanding of EUDAMED EU Medical Device Regulations 3
A 60601-1 understanding of stationary device IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
N Help with basic understanding of Competence requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
D Surface Roughness understanding Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
P Understanding DFMEA and PFMEA - Supplier Related IATF 16949 - Automotive Quality Systems Standard 21
DuncanGibbons Understanding the applicability of Design of Experiments to the IQ OQ PQ qualification approach Qualification and Validation (including 21 CFR Part 11) 5
B Measuring and monitoring equipment - Understanding which procedures to be compliant with ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 6
M Informational Health Canada has launched an e-Learning tool to aid in understanding the premarket regulatory requirements for medical devices in Canada Medical Device and FDA Regulations and Standards News 0
S Understanding UDI requirements - Class 2 medical device (hearing aids) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M Informational Understanding Costs And Risks For HFE Usability Studies — Part 1: Testing In-House Medical Device and FDA Regulations and Standards News 0
P Understanding FDA draft "Management of Cybersecurity in Medical Devices" Medical Information Technology, Medical Software and Health Informatics 3
J Properly understanding SPC - Newbie SPC questions Statistical Analysis Tools, Techniques and SPC 34
S Understanding control chart and measurement capability Statistical Analysis Tools, Techniques and SPC 2
P Minitab Data Analysis - Understanding if a Process is in Control or Not Using Minitab Software 2
R Understanding a few points on ISO 9001's Design and Development Planning ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
Z Understanding Cycle Time - Why the time of the other activities are left out Lean in Manufacturing and Service Industries 11
J Understanding ISO 9001:2015 - 10.3 Continual Improvement ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
J Understanding ISO9001:2015 - 8.3: Design and Development of Products and Services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
E Root Cause Analysis - Is Insufficient Understanding an acceptable Root Cause? General Auditing Discussions 9
E Understanding of TS 16949 Clause 7.6.2 IATF 16949 - Automotive Quality Systems Standard 5
K Understanding IEC 60601-2-68 requirements ISO 13485:2016 - Medical Device Quality Management Systems 1
A Training material for interpretation & understanding Part 11 requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
N Understanding the absolute uncertainty specification for a Fluke 5500A Measurement Uncertainty (MU) 3
N Understanding, Challenging & Approving Supplier Control Plans FMEA and Control Plans 7
M Definition Recommendations - Understanding "recommendations" and "recommended corrective action" Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 8
S Understanding UDI (Unique Device Identification) Other US Medical Device Regulations 10
T Understanding USP <1112> Water Activity as applicable to Medical Devices Other Medical Device and Orthopedic Related Topics 4
K Understanding Risk Management Requirements according to AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
S MIL-HDBK-217 - Understanding the various Environmental Conditions Reliability Analysis - Predictions, Testing and Standards 1
D What is your understanding or interpretation of TS16949 7.4.1.2 IATF 16949 - Automotive Quality Systems Standard 6
C Understanding the relationship between 62304 and the MDD ER IEC 62304 - Medical Device Software Life Cycle Processes 7
S Understanding Subgroup Size - Multi Cavity (Minitab) Statistical Analysis Tools, Techniques and SPC 4
R Understanding clause 15.4.2.1 d) of amendment 1:2012? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M Understanding accreditation, MoUs, certifications Other ISO and International Standards and European Regulations 28
L Mobile Medical App - Understanding 21 CFR Part 820 Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
D Understanding and implementing ISO 17025 ISO 17025 related Discussions 9
M Understanding Versions of Collateral and Particular Standards IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
S Understanding, Analysis and Monitoring Quality Defects on Composite Components Statistical Analysis Tools, Techniques and SPC 3
S Understanding PMS (Post Market Surveillance) and PMCF (Vigilance and PMCF) Quality Manager and Management Related Issues 1
B Understanding why my CpK and PpK are low, and LCL Statistical Analysis Tools, Techniques and SPC 20
S Understanding Quality Objectives, Metrics and KPI ISO 13485:2016 - Medical Device Quality Management Systems 15
Q Beginner's Understanding - The Purpose and Applications of QMS/ISO Standards Philosophy, Gurus, Innovation and Evolution 12
Q Understanding Configuration Management AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
W Understanding PPAP Appearance Approval APQP and PPAP 22
V Understanding Automotive Coating for Seating Mechanism Components Manufacturing and Related Processes 1
M Understanding of Regression and ANOVA in Minitab Statistical Analysis Tools, Techniques and SPC 8
4 Understanding ILAC policy P14:12/2010 6.3 part a) General Measurement Device and Calibration Topics 28
H Understanding 8.2.3 M&M of Processes for our Internal Audit ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6