Unit of Expanded Uncertainty in Percentage

A

akmal122g

Hello. Can someone teach me how to convert the unit of expanded uncertainty for current from percentage to ampere? Example in this certificate is

Nominal value = 1 A
Measured value = 1.0001 A
Error = 0.0001 A
Expanded uncertainty = 0.04 %

Thank you.
 
A

akmal122g

So the expanded uncertainty is 1.0001 * 0.04%=0.0004 A. Am I right?
 

Hershal

Metrologist-Auditor
Trusted Information Resource
The math seems right but the application is the key. If reporting a single measurement, then the same unit is preferred. If across a range, then often a percentage is a better and more understandable approach. Of course, for a range, one can also use a floor and range concept.
 
Thread starter Similar threads Forum Replies Date
E Heater-Cooler Unit for Heart-Lung Machine IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E Please help! ! I AM FRASTRATED SO MUCH! Should inactive ingredient be included in the unit formula if it is removed during the manufacturing process? US Food and Drug Administration (FDA) 0
T Software Unit definition - IEC 62304 - Medical Device Software Life Cycle Processes 3
D 62304 code review, unit test, integration test example IEC 62304 - Medical Device Software Life Cycle Processes 2
F Configurator for a power unit - Software or other solution? Manufacturing and Related Processes 0
I Calibrating the monitoring device but not the unit General Measurement Device and Calibration Topics 7
A MDR requirement where unit of use packaging is too small for UDI carrier EU Medical Device Regulations 1
R HV Test of HF Surgical Unit IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
F UDI - Unit of Use and Class I, single-use devices EU Medical Device Regulations 5
B Unit of Use DI (Device Identifier) - Products using the same device US Food and Drug Administration (FDA) 0
M What to do in SOP Phase of Head unit testing? IATF 16949 - Automotive Quality Systems Standard 0
D Torque fractional unit of measure; cN-m or N-cm? General Measurement Device and Calibration Topics 2
E Unit Dose Packaging Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
M Measurement Unit Rounding Nonconformity - Notified Body Audit Registrars and Notified Bodies 6
P HVAC System vs Air Handling Unit (AHU) - Differences Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
P Fan filter unit (un-ducted to air supply) vs fan filter unit (ducted to air supply) Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
K Production Unit Labeling - 21 CFR 820.184.e 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
P Qualification of a Temperature Controlled Unit have to be empty? Qualification and Validation (including 21 CFR Part 11) 5
M For a Type 1 Gage R&R I want to use 1 unit and measure it 30 times Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
C Software Unit Acceptance Criteria (5.5.4) IEC 62304 - Medical Device Software Life Cycle Processes 3
T ISO 14001:2015 cover more than 1 company or business unit? ISO 14001:2015 Specific Discussions 13
F Setting up a new sterilization unit in a new facility 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
R VSM - Value Stream Mapping (catheter manufacturing unit) Customer Complaints 10
A 5.5.3 - Software Unit Acceptance Criteria (Risk Control Measures) IEC 62304 - Medical Device Software Life Cycle Processes 3
S Is 0.1 micron monitoring required for ISO Class 5 Laminar Air Flow unit ? Other ISO and International Standards and European Regulations 4
O Can unit be categorized as EMI class A when it has EMI class B power supply in it? CE Marking (Conformité Européene) / CB Scheme 2
Z Final Validation Unit Requirements - Annex D of IEC 62366 IEC 62366 - Medical Device Usability Engineering 6
T Is IEC 61010 needed for a populated PCB within a unit? CE Marking (Conformité Européene) / CB Scheme 2
G Help with an old Spectrum 1000BX FTIR unit - User Manual wanted Manufacturing and Related Processes 2
D Calibration Cert - Measurement Unit for Nominal Value of a Calibration Resistor General Measurement Device and Calibration Topics 6
R Does dental unit need to meet the requirements of clause 14 (PEMS)? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
somashekar Use of Set or Pair as Unit of Measure in the Bill of Material (BOM) Manufacturing and Related Processes 5
B Blood Bag Classification - with Disposable Centrifugal Unit for processing blood EU Medical Device Regulations 4
R Does applied part of battery powered TENS unit need to be designed as type BF? IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
F Does the SI acronym stand for Syst?me International d'Unit?s? General Measurement Device and Calibration Topics 1
W What is the applicable standard for Chair Mounted Dental Unit for CE IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Calculating MTBF (Mean Time Between Failures) for the whole Production Unit Reliability Analysis - Predictions, Testing and Standards 2
P How to set up HSE Unit? What is the difference btw HSE Plan and HSE Manual? Occupational Health & Safety Management Standards 32
N Dental Treatment Unit (aka Dentist's Chair) requirements EU Medical Device Regulations 7
I IEC 62304:2006 Definitions - Software System, a Software Element and Software Unit IEC 62304 - Medical Device Software Life Cycle Processes 13
S How to Design the Means of Protection for Home Use Unit IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
V SPC for Continuous Steam Sterilization Unit Statistical Analysis Tools, Techniques and SPC 2
S What is the remedy if Unit level testing is not documented? IEC 62304 - Medical Device Software Life Cycle Processes 13
M Decontamination Label on Unit Packages per UL (Underwriters Laboratories) Manufacturing and Related Processes 1
E Using Customer Unit as Reference Standard General Measurement Device and Calibration Topics 14
S About Ozone Concentration's Unit "pphm" and "ppcm" Reliability Analysis - Predictions, Testing and Standards 4
B GMP Compliant Flooring for Food Manufacturing unit Food Safety - ISO 22000, HACCP (21 CFR 120) 8
W EN 61000-4-4:2004 Compliance - Electrosurgical Unit Transient Voltage Pulse EU Medical Device Regulations 3
B Minitab Gage R&R using Defect Count per Unit Area Using Minitab Software 1
M Customer Complaint vs. Repair - Unit goes out of specification after the warrenty Customer Complaints 8

Similar threads

Top Bottom