Until when can we CE-mark and sell MDD class 1 devices?

#11
Note that if you have class I devices that have been placed on the market prior to May 26th, 2020 and are correctly placed on your declaration of conformity, you may qualify to meet the newly published corrigendum that is to be voted on in the next couple days. However, you must ensure the two requirements are met:
1. The devices are class I devices without EC certificates that are on the declaration of conformity prior to May 26, 2020.
2. The class I devices under MDR conformity assessment rules are required to be up-classified and undergo NB assessment.

If both of those conditions are true, then you can continue to market the class I devices under MDD until may 26, 2024.

This is designed to give a buffer for manufacturers and NBs to undergo necessary up-classification of legacy devices.
 
Elsmar Forum Sponsor
#12
Note that if you have class I devices that have been placed on the market prior to May 26th, 2020 and are correctly placed on your declaration of conformity, you may qualify to meet the newly published corrigendum that is to be voted on in the next couple days. However, you must ensure the two requirements are met:
1. The devices are class I devices without EC certificates that are on the declaration of conformity prior to May 26, 2020.
2. The class I devices under MDR conformity assessment rules are required to be up-classified and undergo NB assessment.

If both of those conditions are true, then you can continue to market the class I devices under MDD until may 26, 2024.

Thank you for that succinct description.
However not clear to me is:
  1. "up-classification":
    Does that mean that devices must have a higher class under MDR than they under MDD? Our MDD class 1 products remain class 1 MDR
  2. "undergo NB assessment":
    Several of our class 1 devices will not require NB authorisation under MDR. Does that mean that our devices that remain class 1 and the ones that remain both class 1 and don't need a NB, do not meet condition 2 and thus do not qualify for the corrigendum?



Additional question to all:
From time to time I read articles and postings about Class 1 medical devices using the following terms as if they are official elements in regulatory documentation:
  • 1s: Devices that are placed on the market in sterile condition
  • 1r: Reusable surgical instruments (r stands for “reusable”)
  • 1m: Devices with a measuring function
Where does this terminology come from and where in official regulatory documentation is it defined? Is it an official subclassification?

Thank you in advance.
 

Ronen E

Problem Solver
Staff member
Moderator
#13
Does that mean that our devices that remain class 1 and the ones that remain both class 1 and don't need a NB, do not meet condition 2 and thus do not qualify for the corrigendum?
First, there is no "qualify for the corrigendum". The corrigendum applies to all, unconditional. You probably mean "do they qualify for the extension/concession promulgated by the corrigendum?"
The answer is that the only devices that remain class 1 and benefit from that new extension are those that are classified 1r under the MDR.
Where does this terminology come from and where in official regulatory documentation is it defined?
This terminology is unofficial and used as a convenience shorthand. Devices classified as "1m", "1s", and "1r" are still all class 1 devices, officially. Nonetheless, they have some special provisions that apply to them.
 
Last edited:
#14
The first part of what Ronen says is correct, in terms of the corrigendum applying to all. I meant in terms of qualification for extension. However, to qualify for the extension, Class I devices via MDD conformity assessment must have an EC cert or have to be any previously self-declared class I devices under MDD that now fall under Class I m/r/s or Class IIa and above (anything needing NB scrutiny) via the MDR conformity assessment.

If your product was class I self-declared under MDD, and continues to be class I self-declared under MDR, the extension does not apply.
 

Ronen E

Problem Solver
Staff member
Moderator
#15
The first part of what Ronen says is correct... However, to qualify for the extension, Class I devices via MDD conformity assessment must have an EC cert or have to be any previously self-declared class I devices under MDD that now fall under Class I m/r/s or Class IIa and above (anything needing NB scrutiny) via the MDR conformity assessment.
(Emphasis added)
The word However above suggests that we are in some sort of disagreement, but I think we're not. Your statement is accurate, but it's possibly difficult to decipher by someone less used to regulatory language, and while it's complete, it exceeds @Son of Paperman's question. My statement was intended to answer his question directly, not necessarily to cover the entire scope of the (corrected) article.

Specifically, when I related to "the extension" it related to the latest, added through the corrigendum (that's why I wrote "that new extension"). The grace period allowed for devices already certified under the MDD (Is, Im or higher classes) was there from the start so I don't see it as "an extension".

I don't see how a self-certified device under the MDD can become a 1m/1s device under the MDR - have any related definitions changed? So the scenario of MDD-self-certified to MDR-class-1-NB-scrutinised boils down to 1r (because there was no "1r" under the MDD).

If you think my statement is not accurate I'd be grateful if you highlight in what way and why.
 
Last edited:

Biomedical_engeener

Starting to get Involved
#16
Does the custom made implants (class I under MDD --> class III under MDR) qualify for the extension promulgated by the corrigendum? The corrigendum does not exclude nor include custom made devices.
Many thanks
 
Thread starter Similar threads Forum Replies Date
D Is it okay to CE mark the product until the RoHS2 directive goes into effect? CE Marking (Conformité Européene) / CB Scheme 7
T EXCEL - How do I make the cell stay blank until a date is entered in N7? Excel .xls Spreadsheet Templates and Tools 6
S ISO 9001 Certification Renewal until 2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
T Weibull analysis of life time test of 8 electric motors until failure Reliability Analysis - Predictions, Testing and Standards 9
A CE Certificate is valid until 3.12.2013 - Can I still sell the products in Europe? EU Medical Device Regulations 10
Randy It's only 97 days until the 2013 DAYTONA 500 Coffee Break and Water Cooler Discussions 4
H ISO 9001:2000 Certificate - Original approval date 18 July 1995 and valid until 2012 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 27
A Special 510k Update - Cease Sales and Distribution until a Decision is Rendered? Other US Medical Device Regulations 11
X Cannot Verify until Customer uses Product - Clause 7.5.2 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
P CBs are not permitted to certify to AS9100 Rev C until 2011? True or False? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 18
D Gap Analysis - Three months until our ISO/TS 16949 3rd party surveillance audit IATF 16949 - Automotive Quality Systems Standard 4
G Until where goes ISO TS 16949 for Suppliers ? IATF 16949 - Automotive Quality Systems Standard 1
D Assessment Time - Time Taken Until In-Patient is Assessed Statistical Analysis Tools, Techniques and SPC 1
ScottK XP supported until 2012, Vista until 2011? After Work and Weekend Discussion Topics 1
A Positive Control of in-process nonconforming parts until rendered physically unusable Nonconformance and Corrective Action 3
ScottK Crud! Less than 3 weeks until the next CMQ/OE exam Professional Certifications and Degrees 11
Crusader ISO 9001 Update Delayed until the Middle of 2009 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
M Record Retention - Obsolete revisions shall be maintained until the next revision Records and Data - Quality, Legal and Other Evidence 24
R Holy liftin' - 11 days left until I become a Mrs.! Coffee Break and Water Cooler Discussions 32
V ISO Review of TS 16949 Extended until approximately 2008 IATF 16949 - Automotive Quality Systems Standard 12
Wes Bucey To sue, or not to sue: That is the question until you are the person bringing suit... Coffee Break and Water Cooler Discussions 73
T How long until we get 'Official Paper' Certificate? Registrars and Notified Bodies 6
S Purchasing Procedure - Purchasing process is not complete until product is received? IATF 16949 - Automotive Quality Systems Standard 11
S Audited against ISO 9002/3/1:1994 until December 2003? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
D PPAP parts held until new PPAP APQP and PPAP 1
V New Product Model CE Mark Certification under MDR EU Medical Device Regulations 0
P Violation of CE mark - Re-use of single-use Products CE Marking (Conformité Européene) / CB Scheme 2
K One CE Mark holder - multiple manufacturing sites EU Medical Device Regulations 3
A First CE-mark class III implantable device under MDD EU Medical Device Regulations 8
F List of countries that 'prefer' a CE Mark CE Marking (Conformité Européene) / CB Scheme 1
C Surgical mask stability testing (CE mark) EU Medical Device Regulations 2
M MDR EU Distributor for our CE Mark product in the for Netherlands EU Medical Device Regulations 6
D Does Manufacture can submit CE mark application under MDD with NB for his New product after May 2020? EU Medical Device Regulations 3
Ed Panek CE Mark Requirements on Kits (off the shelf product requirements met) CE Marking (Conformité Européene) / CB Scheme 2
R IVDR CE Mark for Class A products - Declaration of Conformity CE Marking (Conformité Européene) / CB Scheme 2
G UVC sterilizer CE Mark CE Marking (Conformité Européene) / CB Scheme 5
C Does a CE mark infer meeting all applicable standards? CE Marking (Conformité Européene) / CB Scheme 4
M CE Mark and 510(k) -Global registration requirements for every country Other Medical Device Regulations World-Wide 3
A CE Mark and PMCF study EU Medical Device Regulations 3
T CE mark start date for a device EU Medical Device Regulations 7
P API monogram mark requirements Oil and Gas Industry Standards and Regulations 1
S CE Mark - Classification Confusion EU Medical Device Regulations 12
N CE Mark for Class 1 medical device CE Marking (Conformité Européene) / CB Scheme 11
Y CE Mark - Placing the CE mark on medical device labels EU Medical Device Regulations 2
B Need For BIS Standard Mark? Imported OEM's Power supply,Li-Ion Battery Other Medical Device Regulations World-Wide 0
D Vietnam requirement of CE mark for medical devices CE Marking (Conformité Européene) / CB Scheme 4
A CE mark - Measuring Instruments Directive confusion! CE Marking (Conformité Européene) / CB Scheme 0
S CE mark on a medical device - Textile product CE Marking (Conformité Européene) / CB Scheme 2
V CE Mark under MDD or IVDR - IVD company doing Virtual Manufacturing Other Medical Device Regulations World-Wide 3
S REF and CE mark symbols on multi-component device EU Medical Device Regulations 3

Similar threads

Top Bottom