Note that if you have class I devices that have been placed on the market prior to May 26th, 2020 and are correctly placed on your declaration of conformity, you may qualify to meet the newly published corrigendum that is to be voted on in the next couple days. However, you must ensure the two requirements are met:
1. The devices are class I devices without EC certificates that are on the declaration of conformity prior to May 26, 2020.
2. The class I devices under MDR conformity assessment rules are required to be up-classified and undergo NB assessment.
If both of those conditions are true, then you can continue to market the class I devices under MDD until may 26, 2024.
This is designed to give a buffer for manufacturers and NBs to undergo necessary up-classification of legacy devices.
1. The devices are class I devices without EC certificates that are on the declaration of conformity prior to May 26, 2020.
2. The class I devices under MDR conformity assessment rules are required to be up-classified and undergo NB assessment.
If both of those conditions are true, then you can continue to market the class I devices under MDD until may 26, 2024.
This is designed to give a buffer for manufacturers and NBs to undergo necessary up-classification of legacy devices.