FDA has stated that the although the DMR is a subset of the DHF, the DHF does not necessarily need to contain the actual DMR (whew!). However, if I choose to include the actual DMR in the DHF, do those documents (by default) then need to be constantly updated within the DHF (a maintenance nightmare)? Or - can I indicate that current revisions of those documents are controlled by 820.40 (Document control) and as such are maintained elsewhere? Seems to make sense to me...!
 
Elsmar Forum Sponsor

Al Rosen

Staff member
Super Moderator
#2
Re: Update DHF when DMR changes?

FDA has stated that the although the DMR is a subset of the DHF, the DHF does not necessarily need to contain the actual DMR (whew!). However, if I choose to include the actual DMR in the DHF, do those documents (by default) then need to be constantly updated within the DHF (a maintenance nightmare)? Or - can I indicate that current revisions of those documents are controlled by 820.40 (Document control) and as such are maintained elsewhere? Seems to make sense to me...!
Yes you can. The DHF can reference the documents in the DMR that is located elsewhere. Otherwise you will constantly have to have update it.
 

yodon

Staff member
Super Moderator
#3
Re: Update DHF when DMR changes?

It's my understanding that the DMR is a 'snapshot' of the configuration for a particular configuration. For example, revision A of the device is built according to drawing 1 rev A, drawing 2 rev B, etc. If you transition to revision B of the device and change, say, drawing 1 to revision B, you would update the DMR to reflect the new configuration.

So from my perspective, it would not suffice to say "current revision" of referenced documents in the DMR. You have no basis, then on how to completely identify the configuration for a particular device revision.
 
W

Watchwait

#4
Re: Update DHF when DMR changes?

Thanks Al & Yodon: so here's my "take-away" on both of your inputs..

If we elect to include the DMR in the DHF (either as a list of documents with revision dates, or the actual documents themselves), then the list, or the documents, within the DHF would need to be constantly updated.

Sounds like a great reason for NOT including the DMR in the DHF. Or - if doing so - do so as a list of documents without revision dates. Then the only time the DHF would need to be updated is when we added or removed a document from the DMR (a situation that occurs far less frequently than the revision of existing documents).

If requested, we could always provide the current revision of any document on the DMR though our normal document control process.
 

yodon

Staff member
Super Moderator
#5
Re: Update DHF when DMR changes?

The Design *History* Folder (DHF) is just that - the history of the design of the device. Indeed, it's not revision controlled, but you do need to know what the current release of any particular document (referenced) in the DHF is. The DHF is updated constantly - with any design change.

The DMR is, in essence, a recipe for building the device. If you list the ingredients (documents, etc.) without listing the amounts (revisions), you would not be able to "bake your cake" correctly. The DMR is updated whenever you change the recipe.

There is considerable overlap between the DMR and DHR but they serve different purposes. The DHF is focused on design whereas the DMR is focused on manufacturing.
 
W

Watchwait

#6
Yodon, I've asked this question before but I'll ask it again.

Do you revision control the DMR itself? I know all about DMR Indices, but if I maintain a seperate document listing all the components of a particular device, does this document itself need to be revision controlled? I've never gotten a clear answer back from anyone on this...so hence my continued curiosity! FYI, we do NOT maintain a seperate, physical DMR. Instead, we only reference a DMR Index in an SOP that "points" to the location of all controlled documents.
 

yodon

Staff member
Super Moderator
#7
I say (a definitive) yes. The DMR identifies a particular configuration of a device. If you change that configuration (via drawing change, manufacturing process change, etc.), you need to change the DMR. This provides the necessary linkage for the device pedigree.

Take a look at:

http://www.fda.gov/cdrh/qsr/08dmr.html

To me, the implications in this guidance are quite clear that revision control is needed and expected.

Consider this example. You have a device with 5 different configurations in the field. All of a sudden you start getting complaints / returns. What can you use to identify the EXACT configuration of each device? The DHR might help but may not take you all the way back to the complete configuration. In my world, you would be able to identify the device by serial number, go to the DHR to find the DMR revision, and then have the complete and exact configuration of that device.

I believe that's the intent for the DMR. If you can achieve those results otherwise, then maybe you don't need to revision control the DMR.
 

Al Rosen

Staff member
Super Moderator
#8
I believe that's the intent for the DMR. If you can achieve those results otherwise, then maybe you don't need to revision control the DMR.
:agree1: If you can achieve it in another way you don't have to. Basically it's a means of configuration management.
 
J

javielo

#9
The FDA clearly gives you two options;

1- Create a DMR document for each finished good you have in your facility that includes all the different components of the finished good and you know this includes but it is not limited to, prints, labels, routings, material specs, quality specs, packaging specs, inspection procedures, fixtures, bill of materials, maintenance documents, etc... If your company has many finished goods this is a nightmare and I would use option 2.

2. Create one document and include a table that shows each specific document or component of the DMR in the first column and shows ITS LOCATION in the second column. i.e. Drawings > location could be a master file with hardcopies or it could be an electronic directory or attached to the Document Control Software Database (Oracle, Documentum, Agile, etc...) If you have over 100 finished goods go with this option is the easiest one.

The DHR and DMR are related since the documents in the DHR used to manufacture an item at a specific time must match the revision of the DMR at that specific time. i.e. If I have a DHR record of 1999 the documents inside the DHR record must match the revision listed for all the components at that same specific date in 1999. Its simple if you have any more questions contact me victorjlebron at hot
 

Ronen E

Problem Solver
Staff member
Moderator
#10
I think there's a bit of potential confusion in this thread and a clarification of terms would be in place.

DHF = Design History File = A compilation of the Design Control process(es) outputs, for a certain device / device family.
DMR = Device Master Record = The device's "recipe" (configuration and process specifications) for a specific device revision = The device's Design Transfer's output.
DHR = Device History Record = The evidence that a specific manufactured batch or lot conforms (or not) to a specific DMR revision + its final disposition record.

Each and all of the above can be an actual compilation of documentation / evidence, or just a collection of references to the identity and location of the relevant documentation / evidence, or any combination of these two (as long as clarity & unambiguity is maintained).
 
Thread starter Similar threads Forum Replies Date
P Update on NBOG 2014-3 to address MDR/IVDR, or any plan to do that? EU Medical Device Regulations 0
M Post Market Safety Update Report for devices that will be up-classified under MDR but are currently under MDD EU Medical Device Regulations 3
A When will ISO13485 Update be Released? EU Medical Device Regulations 1
S ISO9001:2015 Implementation / update plan template ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
Sidney Vianna IATF 16949 News Update on the IATF CARA Project (“Common Audit Report Application”) - 12/2020 IATF 16949 - Automotive Quality Systems Standard 1
M Trigger to update of TLL VDA Standards - Germany's Automotive Standards 0
O IATF 16949 News Ford Motors Customer Specific Requirements Update - Nov 2020 IATF 16949 - Automotive Quality Systems Standard 5
D GMDN code update - Canada - Do you know if there is guidance? Canada Medical Device Regulations 0
Ninja Windows 10 100 % disk use after update Coffee Break and Water Cooler Discussions 33
mike10 QMS Update - Rev. up or start over? Document Control Systems, Procedures, Forms and Templates 3
M Do we need to create a new CER or can we just update the existing CER EU Medical Device Regulations 3
dgrainger Informational MHRA guidance update on withdrawal of Notified Bodies Medical Device and FDA Regulations and Standards News 0
F Software development plan for SW update IEC 62304 - Medical Device Software Life Cycle Processes 2
M Informational Update on ISO TC 210 JWG 1 activities Medical Device and FDA Regulations and Standards News 0
M Informational EU – October 2019 update – State-of-play of joint assessments of Notified Bodies in the medical device sector Medical Device and FDA Regulations and Standards News 2
M Informational US FDA Final and update guidances – Software Medical Device and FDA Regulations and Standards News 2
M Informational From Medtech Insight – QSR/ISO 13485 Harmonization Update: FDA Enforcement Discretion Likely When New Rule Stands Up; Draft Reg Coming By Year’s End; Medical Device and FDA Regulations and Standards News 0
M Informational Update from GOV.UK – Regulating medical devices in the event of a no-deal Brexit – UK Responsible Person Medical Device and FDA Regulations and Standards News 0
M Informational Device Shortages Update – The US FDA Announces Two New Innovation Challenges on Device Sterilization Medical Device and FDA Regulations and Standards News 0
J Where can I find the latest update to Reach and ROHS? REACH and RoHS Conversations 9
M Informational Update – MDR and IVDR implementing measures rolling plan – 2 more NBs designated under the new regulations Medical Device and FDA Regulations and Standards News 0
M Informational US FDA – Digital Health Update: Mid-Year Update on Software Precertification Program Medical Device and FDA Regulations and Standards News 0
M Informational US FDA – Device Shortages Update: Challenges to encourage the development of new approaches to device sterilization Medical Device and FDA Regulations and Standards News 0
M Informational MHRA update – Medical devices: guidance for manufacturers on vigilance - June 2019 Medical Device and FDA Regulations and Standards News 0
M Is it compulsory to update the obsolete GMDN codes in ARTG? Other Medical Device Regulations World-Wide 4
M Informational USFDA – Digital Health Update: The FDA Announces an Opportunity for Test Case Volunteers for the Test Plan for the Software Precertification Program Medical Device and FDA Regulations and Standards News 0
M Informational US FDA Guidance update – Requests for Feedback and Meetings for Medical Device Submissions: The Q-Submission Program Medical Device and FDA Regulations and Standards News 0
M Informational TGA – Webinar: An update on the consultation for the Regulation of Software, including Software as a Medical Device (SaMD) Medical Device and FDA Regulations and Standards News 0
M Informational EU – Update on state-of-play of joint assessments of Notified Bodies in the medical device sector Medical Device and FDA Regulations and Standards News 1
E Gamma Sterilization Product Update Manufacturing and Related Processes 2
M Informational EU – April 2019 update of the MDR and IVDR implementing measures rolling plan Medical Device and FDA Regulations and Standards News 0
B IEC 62304 - Update Checklist IEC 62304 - Medical Device Software Life Cycle Processes 2
M Informational ISO 14971 / ISO TR 24971 revision update – atualizações sobre a revisão Medical Device and FDA Regulations and Standards News 1
M Informational IMDRF – Proposed update to Clinical Evaluation documents Medical Device and FDA Regulations and Standards News 0
M Informational USFDA Draft Guidance – Review and Update of Device Establishment Inspection Processes and Standards Medical Device and FDA Regulations and Standards News 0
C Do we need to make a new OFI (Opportunity for Improvement) for each document/form we update? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
M Medical Device News Last update of the MDR and IVDR implementing measures rolling plan – December 2018 Medical Device and FDA Regulations and Standards News 0
M FDA News USFDA Digital Health Update – New actions and documents Medical Device and FDA Regulations and Standards News 0
M FDA News Digital Health Update: USFDA Report on Non-Device Software Functions Now Available Medical Device and FDA Regulations and Standards News 0
M Medical Device News Letter to the health and care sector: update on preparations for a potential no-deal Brexit Medical Device and FDA Regulations and Standards News 0
Marc Software Update 23 November 2018 Forum News and General Information 1
M Medical Device News Joint Action On Market Surveillance Of Medical Devices (JAMS) Releases Progress Update EU Medical Device Regulations 0
Marc Attachment List Update - 2018-11-16 Forum News and General Information 0
K Medical Device Software Update in Field of AIMD IEC 62304 - Medical Device Software Life Cycle Processes 1
M Medical Device News Health Canada update - Applications for Medical Device Investigational Testing Authorizations Canada Medical Device Regulations 0
M Medical Device News IMDRF update - 26-09-18 - Cybersecurity, Premarket Reviews, Personalized Devices Other Medical Device Regulations World-Wide 0
M Medical Device News TGA update 26-09-18 - Medical Devices Safety Update, Volume 6 September 2018 Other Medical Device Regulations World-Wide 0
M Medical Device News Health Canada Update - 05-09-18 - Pilot Device Advice Canada Medical Device Regulations 0
S How to enforce my colleagues to update SOPs? Document Control Systems, Procedures, Forms and Templates 4
M Medical Device News Health Canada update 28-08-18 - Licensing Requirements for 3D-Printed Devices Canada Medical Device Regulations 0

Similar threads

Top Bottom