USB charging and NIBP device
My question may have been at least partly answered in other posts already, but I would like to be 100% sure.
We are designing a NIBP device. The only applicable part (BF) is the cuff itself.
The device will have an accessible USB connector used both for data transmission and charging.
It will be possible to charge it when in use, using a dedicated mains / 5V adapter connected to the USB connector. As part of the risks management it is however anticipated that the user may use another adapter, even it the documentation will forbid it.
As the applicable part has no PATIENT CONNECTION as per 60601-1 definition, and considering chapter 8 of the standard do we still need a galvanic isolation between the USB connector and the rest of the device?
If so, can the isolation be provided by the mains / 5V adapter or is it mandatory to place the isolation in the device itself, after the connector?
Thanks for your views
Peter Selvey
If it's home use you might have to delineate between the user being both "operator" and "patient" at the same time, for example pressing the start button is operator (hence no isolation from USB/internal voltages required), while the cuff measurement is patient (isolation required, but inherent, but obviously OK, air filled tube).
I would be comfortable with this but if you are getting the device tested by a lab they may not see it that way. You should discuss this up front before deciding the test lab.
E SIP/SOP requirements for USB port used for charging IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
F USB-IF standard EU Medical Device Regulations 0
F Is USB-IF standard compliant with ISO13485? Medical Device and FDA Regulations and Standards News 3
S Medical watch Class II (AP Type CF) with USB connection IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
F USB powered handheld medical device - Isolation requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R Classification of USB DC operated medical equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K USB VID/PID certification Various Other Specifications, Standards, and related Requirements 2
E Isolation requirements for USB/Ethernet IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
O Mitutoyo Digital Caliper to PC USB question General Measurement Device and Calibration Topics 2
K USB DC operated medical equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
B USB-Connection - Creepages and Clearances IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Medical Device Software Distribution for a CE-marked medical device via a USB memory stick EU Medical Device Regulations 8
S Development of 2xMOPP galvanic isolator for USB-connected device IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
L PC in patient environment - BF applied part that uses USB to connect to PC IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
N USB Powered EEG - USB provides 1 MOPP want 2 MOPP IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M Labelling Requirements for USB-Powered Device Other Medical Device Related Standards 1
K UL 6601 - Would an off-the-shelf Android tablet, powered by USB, pass UL 6601? Other Medical Device Related Standards 3
R USB power between Medical Device Data Collection Device Component IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
P HIPAA Privacy - Login password or USB Access key? Other US Medical Device Regulations 3
U USB Connector Operator Protection (MOPP) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
E Safety Requirements for Bioimpedance Measuring Device Power via USB Cable IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K USB Charger Adapter/USB Cable - A Medical Device Accessory? EU Medical Device Regulations 2
C Medical device worn on body with USB, isolation, 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S How can I justify adequacy of USB connector soldered on the product Reliability Analysis - Predictions, Testing and Standards 4
J Cost Effective Solution for Isolating USB 5V 500mA Power IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Off the Shelf Parts - Serial to USB adapter for a Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
Q Mitutoyo USB Cables for Variables Data Collection using a Dial Snap Gage Statistical Analysis Tools, Techniques and SPC 2
T USB Drive Encryption and Protection After Work and Weekend Discussion Topics 5
C USB, Clearances and Earth Leakage - Medical Devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S Off the shelf Serial to USB Adapter Impact on 510k 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
S USB Cable (an accessory to our data management software) Changes and 510k IEC 62304 - Medical Device Software Life Cycle Processes 2
I Minitab 14 USB Hardware Key Availbility Using Minitab Software 7
C USB dongle licensing system for Medical Imaging Software Other Medical Device and Orthopedic Related Topics 1
A Medical device with NON medical PC - USB galvanic insulation ISO 13485:2016 - Medical Device Quality Management Systems 2
A USB Port Isolation as per IEC60601-1 - Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R USB Port (Signal I/O) Isolation as per IEC60601-1 - Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 40
T Calibration of a digital usb temprature probe. General Measurement Device and Calibration Topics 3
Jen Kirley Have you tried recording vinyl to MP3s and or CDs? Turntable with USB port? After Work and Weekend Discussion Topics 22
Marc Credit Union Audit - Social Engineering, the USB Way After Work and Weekend Discussion Topics 2
J Free USB Drive from Microsoft! After Work and Weekend Discussion Topics 47
L Class 2b USB enabled diagnostic devices...Clinical trials ISO 13485:2016 - Medical Device Quality Management Systems 4
A RoHS & Electric vehicle charging cables RoHS, REACH, ELV, IMDS and Restricted Substances 1
L EN 62368 (In an Annex - M.4.2.1) Secondary Lithium Battery Charging Safeguards CE Marking (Conformité Européene) / CB Scheme 2
M Charging consulting fees for SOP development and guidance Document Control Systems, Procedures, Forms and Templates 14
C Touch Current for Exposed Charging Contacts on Bodyworn Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Q Charging Customers for REACH Documentation RoHS, REACH, ELV, IMDS and Restricted Substances 8
S Charging Consulting Fees prior to 510(k) Approval Other US Medical Device Regulations 3
C Charging a customer who wants to audit our QMS? ISO 13485:2016 - Medical Device Quality Management Systems 16
T Charging Cradle to charge our Lithium Ion Batteries - IEC 60601 Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R Is reversed charging protection necessary for Rechargeable Batteries IEC 60601 - Medical Electrical Equipment Safety Standards Series 13