Hi All,
We have received a Minor NC by Notified Body as follows:
"The process validation performed by the manufacturer does not follow the requirement of the ISO 13485:2016 standard to use statistical techniques and rationale for sample sizes. The sample size selected for the above mentioned validation report is not justified. The procedure (SOP) related to process validation is also not addressing the requirement for the usage of "statistical techniques with rationales for sample sizes"."
While I do not argue we should have a statement in Validation SOP on statistical techniques use if apropiate as one of the solutions, it seems for me a ISO 13485:2016 standard does not require that as obligatory to be used (see a standard quote below: section 7.5.6) and our "mistake" was rather we have not applied justification for sampling plan we chose during validation rather than not using actual statistical techniques.
Section 7.5.6 of ISO13485:2016 quote:
The organization shall document procedures for validation of processes, including:
d) as appropriate, statistical techniques with rationale for sample sizes(...)"
Having said that:
Is my way of thinking correct that ISO13485:2016 does not make obligatory to use statistical techniques for sampling plans but is leaving it as option?
If so what conditions sampling plans need to meet in order to comply with standard yet not using statistical techniques actually?
Having said that: