Use of Y1 capacitors for MOPP (means of patient protection)

B

BMR96

Hello,


I would like to see if anyone can clarify a point on Y1 capacitors. Section 8.5.1.2 says that a Y1 cap can be used as 1 MOPP if it passes the dielectric strength test of 2 MOPP. I have heard it said, and it stands to reason, that to achieve 2 MOPP with Y1 caps, you simply put 2 caps in series. The standard gives guidelines for using Y1 caps in series, but it does not explicitly say that 2 caps in series will actually be equivalent to 2 MOPP. So, any thoughts on this? Is it as simple as 2 caps = 2 MOPP?


Thank you.
 

Peter Selvey

Leader
Super Moderator
Yes, it is fairly common practice, and obviously the intended purpose of the statement about keeping the values the same capacitance.

Clause 8.5.1.1 is the basic clause that says 2 MOP are required, I guess that stays in the background.
 