Our company plans to include a sterile saline flush syringe from an external company in the package for our class II device, just for the convenience of the user. The saline syringe is a 510(k) cleared class II device, and the whole device will also be a class II device. My question is: Will the FDA require us to include the instructions for use / user manual (from the other company) for the syringe in our package?
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#2
First you need to clarify whether "your package" is a convenience kit (specific FDA expectations apply) or just 2 devices - yours, and the syringe in its original package - packaged together for shipping / wholesale purposes.
A third scenario is that "your device" is the syringe + other objects that you make (or source), packaged together is a primary/retail packaging. In this case the syringe is a component and its 510(k) is less relevant. If that's the case you'll need to look at all that as one device, and provide appropriate labelling.
 
#3
Thanks for your reply, Ronen E. I think we have the third scenario. In EU there is a rule saying that you don't necessarily need to provide IFUs for device classes IIa and under, but there is no similar rule in the US, am I right?
 

Ronen E

Problem Solver
Staff member
Moderator
#4
In EU there is a rule saying that you don't necessarily need to provide IFUs for device classes IIa and under
This is an inaccurate statement (as phrased).

You can find all the applicable labeling rules (USA), including any exemptions, in 21 CFR 801 and here.
 
Last edited:
Thread starter Similar threads Forum Replies Date
S Procedures and/or Work Instructions for New User Manual? Document Control Systems, Procedures, Forms and Templates 2
K EU MDR Art. 22 - Device + insertion pack - User manual and Labeling EU Medical Device Regulations 4
shrutisancheti EU User manual / operator manual / service manual guidance document(s) CE Marking (Conformité Européene) / CB Scheme 2
J TOPCON PP-70 Optical Comparator - User Manual Needed General Measurement Device and Calibration Topics 0
JoCam User Manual Release before CE Mark EU Medical Device Regulations 5
M China Medical Device Requirrements - User Manual Information China Medical Device Regulations 2
P Modification to Medical Device User Manual Other US Medical Device Regulations 3
G Help with an old Spectrum 1000BX FTIR unit - User Manual wanted Manufacturing and Related Processes 2
E Micro-Vu Inspec Software User Manual Needed Quality Assurance and Compliance Software Tools and Solutions 1
J Packaging and User Manual for Medical Device and Accessory CE Marking (Conformité Européene) / CB Scheme 3
N User Manual Content Requirements and EMC information ISO 13485:2016 - Medical Device Quality Management Systems 1
W Do I have to review the User Manual? Other US Medical Device Regulations 1
shimonv Requirement for a Printed Copy of the User Manual with a Medical Device US Food and Drug Administration (FDA) 5
H Spelling Errors in User Manual (IFU) for Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 13
K How to Write a User's Manual for Peachtree - Seeking help Document Control Systems, Procedures, Forms and Templates 4
C Indications for Use vs. Intended Use in User Manual 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
K Starrett RGC4040-24 CMM User and Installation Manual General Measurement Device and Calibration Topics 2
N The 'user-friendly' Quality Manual ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
N Is it compulsory to print the CE mark on our software & user manual CD-ROM EU Medical Device Regulations 3
K User Manual and Label Language Requirements ISO 13485:2016 - Medical Device Quality Management Systems 10
D Looking for a Gigatronics 610 User's Manual or Service Manual or both General Measurement Device and Calibration Topics 3
K Apogee V2.5 User Manual Quality Assurance and Compliance Software Tools and Solutions 0
Z Choice of PTC heater voltage to maximize user safety and ease of CE compliance to LVD CE Marking (Conformité Européene) / CB Scheme 0
Brizilla User Profile - How do I edit my profile info? Elsmar Xenforo Forum Software Instructions and Help 3
T Documenting hazardous situations associated with user/patient population ISO 14971 - Medical Device Risk Management 3
A The refund of 510(k) user fee Medical Device and FDA Regulations and Standards News 0
A User interface Language Requirements - EU MDR EU Medical Device Regulations 1
S User evaluation for self monitoring blood glucose test systems US Food and Drug Administration (FDA) 4
P Equipment 21 CFR 820.70(g) - User Requirements Document for Off the shelf equipment 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
S Forced ServiceNow validation - No change in our current user and functional requirements IT (Information Technology) Service Management 4
D Software User Interface Languages for LVD and IVD CE Marking (Conformité Européene) / CB Scheme 2
Q User Requirement Specification for HR (Human Resource Management System) Manufacturing and Related Processes 1
P Electrosurgical Device User Need: Cord Flexibility -> Requirement Other Medical Device and Orthopedic Related Topics 4
S Recommendation for user friendly Gaga R&R and Cpk software Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
B ISO 11607-1 for surgical instruments sterilized by end user Other Medical Device Related Standards 1
D Do User Manuals (Not Device Labeling) HAVE to be in their national languages for these countries? Will English Hardcopy not do? EU Medical Device Regulations 12
A Cassette Representation - GUI (Graphical User Interface) on my medical device - A mammography Device Other Medical Device Related Standards 0
R Medical device software without user interface Other Medical Device and Orthopedic Related Topics 3
F EUDAMED UDI Medical Devices User's Guide Medical Device and FDA Regulations and Standards News 7
M Informational US FDA Medical Device User Fee Rates for Fiscal Year 2020 Medical Device and FDA Regulations and Standards News 0
G Posting Measuring Equipment Accuracy for User Information General Measurement Device and Calibration Topics 4
M Does ISO 13485 or MDR require you to state the origins of customer requirements or user needs? Design and Development of Products and Processes 2
V Sequence of performing risk assessment: User_FMEA (User Errors) vs Design Inputs FMEA and Control Plans 1
Ed Panek Splitting UI (User Interface) into two development paths 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
R Publication of user information in manufacturer website for medical devices Other Medical Device Regulations World-Wide 2
Ed Panek User Feedback both negative and positive and acting upon those metrics 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
P Detection rating for a user (surgeon) related failure mode in DFMEA APQP and PPAP 3
Marc Search - Search for Discussion Threads started by a specific User Elsmar Xenforo Forum Software Instructions and Help 0
Marc Search for Posts - By specific User Name(s) Elsmar Xenforo Forum Software Instructions and Help 1
M Digital user manuals CE Marking (Conformité Européene) / CB Scheme 0

Similar threads

Top Bottom