Using a "broken" measurement device in a TS 16949 company - Optical comparator

D

danno101

In a TS 16949 calibration system, is it allowable to continue to use broken measuring equipment? We have an optical comparator that is used to measure parts and tooling. It is calibrated yearly by a 17025 certified cal house.

What actions are needed to take for this condition? What is the unit is found to be out-of-calibration after repairs?

The floor supervisors do not see any risk involved with using it. How can I present an argument against their position?
 

howste

Thaumaturge
Trusted Information Resource
It would be allowed only if the functions being used are still working, are calibrated, and are reliable.

Can you describe what's broken?
 
D

danno101

Hi. The y-axis is jumping all around. The repair technician said (over the phone) that someone probably damaged it.
 

howste

Thaumaturge
Trusted Information Resource
Then I wouldn't use the y axis without repair. If you need to convince someone that it's unreliable perform a gage R&R and it will show how poorly suited it is for the tolerances being measured.
 

BradM

Leader
Admin
I cannot speak towards the TS16949 aspect, but as a general rule...

But if you have a piece of calibrated (or qualified) piece of equipment and there is a known issue with it, that equipment should be caution-tagged at the very least. What if someone comes up and uses it, records data, and that axis measurement gives them incorrect data?

Or, depending on your local procedures, you can assess and tag the equipment for a limited calibration. So they could you it for X and Z axis measurements, but not Y axis measurements.

However, there should still be a risk assessment made. There is no guarantee that 'whatever' knocked the Y axis out of alignment, didn't happen to the X and Z axis also.
 
Thread starter Similar threads Forum Replies Date
Stoic Which medical device companies are using Recombinant Factor C (rFC) instead of Limulus Amebocyte Lysate (LAL) for endotoxin testing? Sustainability, Green Initiatives and Ecology 0
J Using MSA instead of calibration Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
J Using current gauge to verify masters General Measurement Device and Calibration Topics 1
Q Using the N/A for blank fields ISO 13485:2016 - Medical Device Quality Management Systems 8
S Anyone using a GOM ATOS Scanbox 3D Vision System? General Measurement Device and Calibration Topics 0
M Any alternative method instead of using GR&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 26
R Using RPN to Confirm Risk Reduced to an Acceptable Level Risk Management Principles and Generic Guidelines 12
P Comparing Two Test Variables Using Attribute Data Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
J Using real time shelf life study as accelerated ISO 13485:2016 - Medical Device Quality Management Systems 1
C Clinical Trial using Prototype devices (EU MDR & ISO 14155) EU Medical Device Regulations 4
J Using an online software to maintain your QMS Quality Assurance and Compliance Software Tools and Solutions 7
S Process Monitoring using SPC software Quality Assurance and Compliance Software Tools and Solutions 6
R Using Scrapped NC items for training Medical Device and FDA Regulations and Standards News 15
M What is the Risk of Using Obsolete Versions of C=0 & ANSI/ ASQ Z1.4 Sampling Plans? ISO 13485:2016 - Medical Device Quality Management Systems 8
F Using non-randomized clinical study for change in existing 510(k) device? US Food and Drug Administration (FDA) 1
J Using MD on a flight EU Medical Device Regulations 9
K Using Google Drive/Docs ISO 13485:2016 - Medical Device Quality Management Systems 32
J What risk to cover when NOT using ISO 17025 accredited/certified labs for calibration ISO 17025 related Discussions 3
T Poisson regression using Minitab Statistical Analysis Tools, Techniques and SPC 3
T Using NICE for guidance during CER CE Marking (Conformité Européene) / CB Scheme 0
T Using Review articles from Pubmed for CER CE Marking (Conformité Européene) / CB Scheme 8
Tagin Using a meme generator to poke fun at quality Funny Stuff - Jokes and Humour 2
D Finding Optimum Design Parameters using Taguchi method? Using Minitab Software 2
V MSA - Using a R&R study for a Tape Measure Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
D Using electronic lab notebooks in the design control process ISO 13485:2016 - Medical Device Quality Management Systems 3
M Using a Non Certified vendor Supplier Quality Assurance and other Supplier Issues 8
H Using/Selling power banks with your product as a "mobile version" CE Marking (Conformité Européene) / CB Scheme 3
cgaro62 Using internal audits as part of document reviews Document Control Systems, Procedures, Forms and Templates 21
R Select the 1 Supplier based on the Parts Durability from 6 Supplier Samples using Minitab Using Minitab Software 11
M MSA angle issue - Nylon parts using Aberlink 3D Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
D Question on using audit checklist ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 20
J Using of CE marking - Two CE markings on the product CE Marking (Conformité Européene) / CB Scheme 5
R Using R package to implement Bayesian phase I/II dose-finding design for three outcomes ISO 13485:2016 - Medical Device Quality Management Systems 6
S Defibrillator Protection applicability When using Multiple applied parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
P Using Zinc Plated Steel in a Cleanroom Other Medical Device Related Standards 4
J Incoming Inspection Records using Excel File ISO 13485:2016 - Medical Device Quality Management Systems 2
J Using ring gauges General Measurement Device and Calibration Topics 2
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
P Can Neoprene be Cleaned Using Isopropyl Alcohol (IPA) Manufacturing and Related Processes 4
GreatNate Anyone using the Intellect QMS software? Quality Assurance and Compliance Software Tools and Solutions 1
chris1price Sterilization using beta radiation Other Medical Device Related Standards 2
M Using the phrase "herein referred to" Document Control Systems, Procedures, Forms and Templates 8
B Struggling with using the 5.6 version Ford Capacity Analysis Report APQP and PPAP 5
cnbrosa Study Type 1 on a CMM using a measuring support Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
rerusk1 MRB (Material Review Board) Process using MS Sharepoint or MS Teams Manufacturing and Related Processes 2
K 510k FDA review, will they accept Biocompatibility result generated using feasibility product lots? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
U Medical Device CE Marking - Using a disposable bearing CE Marking (Conformité Européene) / CB Scheme 3
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 3
D Risk Analysis using Monte Carlo Simulation instead of Scoring and Heat Map Risk Management Principles and Generic Guidelines 2

Similar threads

Top Bottom