Hello,

I am new to quality systems. I am working with a start up and my task is to build the quality system.

I am using a software we bought that includes a template of the ISO 13485 and modifying to best fit our company. I have lots of questions, but the one that is at the top of my list is about using a vendors quality system as part of our quality system.

The main component of our device is software. We do not build software, and none of us are computer programmers. We have contracted a company that specializes in creating software for medical device companies. Their quality system appears to be rock solid.

I did some research and asked someone at the FDA if it was ok to rely on a vendor's QS. They said it was ok as long is it was part of our quality system and procedures.

My questions is: Where in the manual or system would I put this? would it be part of the Quality manual itself? or do I put it under a SOP?

I am not sure if this is the correct forum for this thread, so please move if needed.

Thank you,

GR

PS. We are building a mobile telemetry monitor from components that are already approved. We are outsourcing the creation of the software that is going to link everything.
 
Elsmar Forum Sponsor

somashekar

Staff member
Super Moderator
#2
Re: Using a vendor's quality system...

Welcome to the Cove Noblelobo .... (GR)
You would simply put the vendor's quality system as one of the controls of your out sourced process as required in the clause 4.1 of ISO13485.
No SOP required here. This can be stated in the manual if you wish to OR detailed within your procedure for purchasing (7.4)
 
Last edited:

somashekar

Staff member
Super Moderator
#3
Re: Using a vendor's quality system...

I have moved this thread to "ISO 13485 and ISO 14969 - Medical Devices - Quality Management Systems"
 

yodon

Staff member
Super Moderator
#4
Re: Using a vendor's quality system...

While what Somashekar notes would work, I would question why you would want to do it. The Quality System is supposed to be reflective of how YOU manage your business.

If the question is about adopting your vendor's procedures for developing software, that should NOT be in your QS since you don't develop software. You are contracting out for that service so your QS would describe how you manage software vendors (as part of purchasing controls, for example).

If you're wanting to adopt more generic procedures (e.g., CAPA) then even if it works for you, you still need to perform it internally so make it yours. If there was an error in their system then you would be adopting a system with an error! If they correct the error, how would you know to update? If they call out specific roles that you don't have, will you create roles just to adapt to the QS? If their QS assumes some skills for some roles, will you have those skills? Will you even know what the skills are?

I would not recommend adopting at all. I think it's just too big a risk. Use the good ideas but make them fit your company.
 

sagai

Quite Involved in Discussions
#5
Re: Using a vendor's quality system...

I am 100% with the principles pointed by Yodon.
I am just looking out of my head and I do not understand how the supplier qms could be an option for a manufacturer. Doing the job you have described can be as complicated as a surgery operation in this way and at the end likely you will spend more than you would have done alone.
Regards, Szabolcs
 
N

Noblelobo

#6
Our problem is that we are creating a system that uses already FDA approved devices, minus the software that links them. This software is being developed by a software consulting firm. We have no experience with software and do not plan on making our software anytime soon. We do not really manufacture anything. We will be contracting everything out. The devices that are already approved will shipped to our facility, we will hook them up to a mobile phone via blutooth, and the software installed in the phone and our servers link up everything.
 

somashekar

Staff member
Super Moderator
#7
Our problem is that we are creating a system that uses already FDA approved devices, minus the software that links them. This software is being developed by a software consulting firm. We have no experience with software and do not plan on making our software anytime soon. We do not really manufacture anything. We will be contracting everything out. The devices that are already approved will shipped to our facility, we will hook them up to a mobile phone via blutooth, and the software installed in the phone and our servers link up everything.
They said it was ok as long is it was part of our quality system and procedures.
As the s/w is a major component and is outsourced, a good quality agreement that is established independent of other prevailing agreements between you and the outsourced s/w consulting firm, in itself can be said as being part of your quality system and procedures.
 

sagai

Quite Involved in Discussions
#8
Hi,
I would raise some considerations.
1.,
Regardless you are "hook them up to a mobile phone via blutooth" I think the "system" is subject to market clearance, due to obvious reason, it has it is own intended use.
So I think you will be the manufacturer with all the consequences and it is not limited to ISO13845 only.
2.,
I would spend an investigation on whether or not the SW component manufacturer subject to establishment registration.
3.,
the totally outsourced SW development
Well, if you can not claim control on the r&d QMS, I think you should validate all released sw version prior building into your system (possible you do not want to do that).
I think, the only way not to take that lane, if you can ensure the conformance of your SW vendor QMS to cGMP and to ISO13485. To do so, you should ensure that in the quality agreement, you should state that, and you should check them in the course of customer audit and follow up and control the whole process.
If you do not have such knowledge, involvement of third party expertise could help, well, lot of option could be, it depends on your intent.

Regards
Szabolcs
 
Thread starter Similar threads Forum Replies Date
M Evaluating Vendor (Supplier) Products using SQC Statistical Analysis Tools, Techniques and SPC 6
J Incoming Inspection Records using Excel File ISO 13485:2016 - Medical Device Quality Management Systems 2
J Using ring gauges General Measurement Device and Calibration Topics 2
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
P Can Neoprene be Cleaned Using Isopropyl Alcohol (IPA) Manufacturing and Related Processes 4
GreatNate Anyone using the Intellect QMS software? Quality Assurance and Compliance Software Tools and Solutions 1
chris1price Sterilization using beta radiation Other Medical Device Related Standards 2
M Using the phrase "herein referred to" Document Control Systems, Procedures, Forms and Templates 8
B Struggling with using the 5.6 version Ford Capacity Analysis Report APQP and PPAP 5
cnbrosa Study Type 1 on a CMM using a measuring support Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
R MRB (Material Review Board) Process using MS Sharepoint or MS Teams Manufacturing and Related Processes 2
K 510k FDA review, will they accept Biocompatibility result generated using feasibility product lots? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
U Medical Device CE Marking - Using a disposable bearing CE Marking (Conformité Européene) / CB Scheme 3
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 1
D Risk Analysis using Monte Carlo Simulation instead of Scoring and Heat Map Risk Management Principles and Generic Guidelines 2
W Using tailoring guidelines to tailor a QMS procedure ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
Y We found out we have been using a equipment without validation for past 4 years Quality Manager and Management Related Issues 6
D Using Laboratory Notebooks in R&D and Design and Development ISO 13485:2016 - Medical Device Quality Management Systems 3
C Importer shell game - Using a third party logistics provider (3PL) in the EU EU Medical Device Regulations 5
S Work performed in Canada on US patients using US device Canada Medical Device Regulations 1
S Is using ANSI/ASQ Z1.4-2008 the correct sampling plan to determine Pass/Fail of Apparel measurements? AQL - Acceptable Quality Level 4
A What are the pros and cons of using an audit software for internal auditing? General Auditing Discussions 7
Tagin Evaluating nonconformances for escalation using Bayesian methods? Statistical Analysis Tools, Techniques and SPC 2
D Using non-conforming components even though the final assembly is conforming? Manufacturing and Related Processes 5
B Using Unreleased Documents & Process Maps for Internal Audit purposes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
R Clause 7.7 Replicate, Recalibration and Intermediate checks using Artifact ISO 17025 related Discussions 1
Stoic Are any medical device companies using the 2011 FDA process validation guidance instead of GHTF/SG3/N99-10:2004? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
B Unit of Use DI (Device Identifier) - Products using the same device US Food and Drug Administration (FDA) 0
A Using Arduino based sensors for Poke-Yoke Manufacturing and Related Processes 6
M Using your Manufacturer's ISO certification ISO 13485:2016 - Medical Device Quality Management Systems 5
Ed Panek Adverse Event Clinical Trial using a 510K approved Device Other US Medical Device Regulations 6
B Using non CE parts in a machine CE Marking (Conformité Européene) / CB Scheme 1
G Gauge R&R on multiple dimensions using 3D measurement system Capability, Accuracy and Stability - Processes, Machines, etc. 6
A Hospital IT expectations for connected medical device using WIFI Medical Information Technology, Medical Software and Health Informatics 0
Proud Liberal Cp / Cpk on position using multiple MMC bonuses Capability, Accuracy and Stability - Processes, Machines, etc. 2
B Using external FDA and ISO 13485 audit as internal audit Internal Auditing 6
D Using "Particle Size Standard" templates as gauges - How to avoid giving a gauge # while using for process control? General Measurement Device and Calibration Topics 2
J Class 1 Medical Device - Using a UPC over the UDI? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
I Reducing CE marking cost using manufacturer test reports CE Marking (Conformité Européene) / CB Scheme 5
CPhelan Using clinical trial safety data for evidence for CE marking EU Medical Device Regulations 8
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate Qualification and Validation (including 21 CFR Part 11) 3
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate General Measurement Device and Calibration Topics 2
M Raw stock material testing discrepancy using an XRF (x-ray fluorescence) analyzer Manufacturing and Related Processes 7
Z Using FMEA for Knowledge Management FMEA and Control Plans 6
Sidney Vianna NASA to Develop a Novel Approach for All-Electric Aircraft Using Cryogenic Liquid Hydrogen as Energy Storage World News 2
S How many of you are using Robotic process automation for calibration lab management? ISO 17025 related Discussions 0
J Including Repeats in DoE using Minitab Using Minitab Software 5
S Legal Manufacturer FDA Reporting Obligations for Using New Contract Sterilization Site Other Medical Device Regulations World-Wide 0
John Predmore What size pinhole can be reliably detected using visual inspection? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7

Similar threads

Top Bottom