Using Calibrated OD mic over non-Calibrated ID mic for final dimension

Michael_M

Trusted Information Resource
#11
[Michael M] - I don't understand your statement. No matter how you approach it, you ARE using the inside micrometer as the actual measurement device. You're just choosing to obtain your readout from another separate device.
You are correct. I never trusted an inside micrometer for the reading itself. The more I thought about this (after posting), the more I think I used the OD micrometer because we did not have anything to qualify an inside micrometer properly. I think I would still use the OD micrometer to check the actual diameter but fortunately, I don't check parts that are that big anymore.

However the (accurate?) measurement is obtained, unless there is a reason to restrict the measurement method (PO clause, quality issue, etc.) I don't see any reason to force a change.
 
Elsmar Forum Sponsor

BradM

Staff member
Admin
#12
So I understand....

They are manufacturing a product for you, where they are creating a hole in something. They are using device A to measure the inside diameter of that hole. Let's say they measure 1.0 inch

They then take device B to measure device A in the same position where it measured device A. Device B then yields 1.0 inch. They then accept the values given by device B for product acceptance.

If that is accurate,

1. That seems like an odd way to go from point A to point B. I don't understand why they have such a convoluted process.
2. More importantly, have they done any kind of gauge analysis/ uncertainty estimation of how good this process is? It sounds to me like there is quite a bit of error in that. Have they had a lot of failures?
3. Umm... last time I checked, you're the customer. I find it sad customer service on their part that they don't at least listen to the customer better and possibly see if there is a better way of doing their measurements.

They have the responsibility to not only assure their processes are "compliant" but more importantly, that they are satisfying the customer. I know that sounds hokey; but to me, is important. They need to win your confidence that they are producing quality product for you. If they want your long term business, they need to go the extra mile for that. Otherwise, it's worth your time and effort to search for other potential vendors, IMO.
 

Eredhel

Quality Manager
#13
It also sounds like they are sending all their tools out to be calibrated. I wonder if you could suggest they send out standards and then calibrate in house.
 

Scott.Hilton

Starting to get Involved
#14
Using an inside micrometer and then transferring it to an OD micrometer is the typical practice I followed for years (when I made parts big enough to use an ID micrometer). I never trusted the inside micrometer enough to use it as the actual measurement device. On a hole size with a tolerance of .003-.005 (total tolerance I assume not +/- .003-.005) I would be willing to accept the transfer. I might tighten up the tolerance a bit (and if I was performing the measurement I would tighten up).

Are you having quality issues with the supplier? Do you have any specific purchase order clauses that restrict this type of measurement? If no to these questions, I would not pursue it. If yes to these question then you have someplace to start.
-we are having some issues with the supplier and some of them even involve ID's. However, I am "politically" stuck now on it. We don't have clauses that restrict the usage.
 

Scott.Hilton

Starting to get Involved
#15
So I understand....

They are manufacturing a product for you, where they are creating a hole in something. They are using device A to measure the inside diameter of that hole. Let's say they measure 1.0 inch

They then take device B to measure device A in the same position where it measured device A. Device B then yields 1.0 inch. They then accept the values given by device B for product acceptance.

If that is accurate,

1. That seems like an odd way to go from point A to point B. I don't understand why they have such a convoluted process.
2. More importantly, have they done any kind of gauge analysis/ uncertainty estimation of how good this process is? It sounds to me like there is quite a bit of error in that. Have they had a lot of failures?
3. Umm... last time I checked, you're the customer. I find it sad customer service on their part that they don't at least listen to the customer better and possibly see if there is a better way of doing their measurements.

They have the responsibility to not only assure their processes are "compliant" but more importantly, that they are satisfying the customer. I know that sounds hokey; but to me, is important. They need to win your confidence that they are producing quality product for you. If they want your long term business, they need to go the extra mile for that. Otherwise, it's worth your time and effort to search for other potential vendors, IMO.
You pretty much stated exactly all the thoughts that went through my head. I would love to request a Gage R&R.............but it is now out of my hands.
 
I

iamtroll

#16
We use transfer gaging for all dimensions over 4" on a routine basis since our Mitutoyo borematics only go up to 4" max. So we are very comfortable using a micrometer to set up our bore gages for over 4". The borematics are three point contact so you can't use them to transfer to a micrometer. But if we had two point gages that we were not confident in the indicators, we would probably consider using them as transfer gages rather than buy new ones at least in the short to medium term (maybe even long term). I'm sure that some of our customers might object, but making measuring decisions are an internal business decision unless you are contractually obligated to a methodology. In cases where we have been contractually obliged or found that we could not measure parts accurately to our own or our customers satisfaction, we have even leased some higher end measuring devices. Another alternative we have pursued is sharing gaging with the customer. It all depends on the confidence that you have in your measuring capabilities and the relationship you have with your customer. Customers have to understand that for suppliers, particularly in a job shop environment, the best measuring devices are not always available and management is probably not going to allow the quality department the spend the money unless there is obvious payback and justifiably so.
 

normzone

Trusted Information Resource
#17
Yeah, I remember the first time I saw a gage travel with a product from department to department, then to a customer and back again.

Now here's a fun idea - somebody buy a pi tape.
 

Michael_M

Trusted Information Resource
#18
Now here's a fun idea - somebody buy a pi tape.

I have not used (or even seen) one of those in 20+ years. I used one to measure a 61 inch diameter. It took 3 of us to keep the tape 'straight' and 'remove' gravity from the equation (Vertical lathe). I remember those days, and not fondly.
 
Thread starter Similar threads Forum Replies Date
G Filling out inspection report using calibrated pins Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
D Calibration of Pin Gages using a Calibrated Caliper that measures to 4 decimal places General Measurement Device and Calibration Topics 15
H "Internal Calibration" using other Calibrated Equipment Calibration Frequency (Interval) 13
M Inhouse Calibration of Regulated Power Supplies (RPS) using calibrated Multimeter General Measurement Device and Calibration Topics 7
D ASTM E-18 - We test our rockwell tester using calibrated test blocks General Measurement Device and Calibration Topics 3
D Interesting Discussion Using "Not A Calibrated Device" or "No Calibration Required" stickers General Measurement Device and Calibration Topics 63
J Incoming Inspection Records using Excel File ISO 13485:2016 - Medical Device Quality Management Systems 2
J Using ring gauges General Measurement Device and Calibration Topics 2
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
P Can Neoprene be Cleaned Using Isopropyl Alcohol (IPA) Manufacturing and Related Processes 4
GreatNate Anyone using the Intellect QMS software? Quality Assurance and Compliance Software Tools and Solutions 1
chris1price Sterilization using beta radiation Other Medical Device Related Standards 2
M Using the phrase "herein referred to" Document Control Systems, Procedures, Forms and Templates 8
B Struggling with using the 5.6 version Ford Capacity Analysis Report APQP and PPAP 5
cnbrosa Study Type 1 on a CMM using a measuring support Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
rerusk1 MRB (Material Review Board) Process using MS Sharepoint or MS Teams Manufacturing and Related Processes 2
K 510k FDA review, will they accept Biocompatibility result generated using feasibility product lots? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
U Medical Device CE Marking - Using a disposable bearing CE Marking (Conformité Européene) / CB Scheme 3
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 1
D Risk Analysis using Monte Carlo Simulation instead of Scoring and Heat Map Risk Management Principles and Generic Guidelines 2
W Using tailoring guidelines to tailor a QMS procedure ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
Y We found out we have been using a equipment without validation for past 4 years Quality Manager and Management Related Issues 6
D Using Laboratory Notebooks in R&D and Design and Development ISO 13485:2016 - Medical Device Quality Management Systems 3
C Importer shell game - Using a third party logistics provider (3PL) in the EU EU Medical Device Regulations 5
S Work performed in Canada on US patients using US device Canada Medical Device Regulations 1
S Is using ANSI/ASQ Z1.4-2008 the correct sampling plan to determine Pass/Fail of Apparel measurements? AQL - Acceptable Quality Level 4
A What are the pros and cons of using an audit software for internal auditing? General Auditing Discussions 7
Tagin Evaluating nonconformances for escalation using Bayesian methods? Statistical Analysis Tools, Techniques and SPC 2
D Using non-conforming components even though the final assembly is conforming? Manufacturing and Related Processes 5
B Using Unreleased Documents & Process Maps for Internal Audit purposes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
R Clause 7.7 Replicate, Recalibration and Intermediate checks using Artifact ISO 17025 related Discussions 1
Stoic Are any medical device companies using the 2011 FDA process validation guidance instead of GHTF/SG3/N99-10:2004? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
B Unit of Use DI (Device Identifier) - Products using the same device US Food and Drug Administration (FDA) 0
A Using Arduino based sensors for Poke-Yoke Manufacturing and Related Processes 6
M Using your Manufacturer's ISO certification ISO 13485:2016 - Medical Device Quality Management Systems 5
Ed Panek Adverse Event Clinical Trial using a 510K approved Device Other US Medical Device Regulations 6
B Using non CE parts in a machine CE Marking (Conformité Européene) / CB Scheme 1
G Gauge R&R on multiple dimensions using 3D measurement system Capability, Accuracy and Stability - Processes, Machines, etc. 6
A Hospital IT expectations for connected medical device using WIFI Medical Information Technology, Medical Software and Health Informatics 0
Proud Liberal Cp / Cpk on position using multiple MMC bonuses Capability, Accuracy and Stability - Processes, Machines, etc. 2
B Using external FDA and ISO 13485 audit as internal audit Internal Auditing 6
D Using "Particle Size Standard" templates as gauges - How to avoid giving a gauge # while using for process control? General Measurement Device and Calibration Topics 2
J Class 1 Medical Device - Using a UPC over the UDI? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
I Reducing CE marking cost using manufacturer test reports CE Marking (Conformité Européene) / CB Scheme 5
CPhelan Using clinical trial safety data for evidence for CE marking EU Medical Device Regulations 8
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate Qualification and Validation (including 21 CFR Part 11) 3
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate General Measurement Device and Calibration Topics 2
M Raw stock material testing discrepancy using an XRF (x-ray fluorescence) analyzer Manufacturing and Related Processes 7
Z Using FMEA for Knowledge Management FMEA and Control Plans 6

Similar threads

Top Bottom