Using K-Factor(Tolerance Interval) Analysis for Design Verification

V

vbhatta

#1
Hi,
I am working on a Design Verification project for a customer where the acceptance criteria used for a diameter feature is K-Factor Analysis at 90%/90% Confidence/Reliability (C/R). The idea is that at stated C/R values, once you compute your Tolerance Intervals, they should be within your specification limits (Upper-one sided limit in our case)

Out of 30 datapoint collected recently, 1 point is outside specification. However, the K-factor (or Tolerance Interval) is still within specification limit (as the process average is well within the USL and std. Deviation is very small).

Our team is suggesting that we can chose to ignore the fact that one point was outside specification since the K-factor is still within specification limits! Doesn't this violate the very first requirement to perform any statistical analysis, that is, the data has to be "in control" prior to analysis?

Has anyone else run into this situation? If so, how did you handle this predicament?

Thanks in advance for your help!
 
Elsmar Forum Sponsor

Bev D

Heretical Statistician
Leader
Super Moderator
#2
can you describe what the 'k-factor' analysis is? I think I know, but I've been surprised by Customer requirements before. It might also be helpful for you to post your data if you can.

The requirement for statistical analysis is not that the data (or process) be in control. but rather that the choice statistical analysis matches the data's true state of distribution.

If you are referring to the traditional process capability study, then yes the process needs to be in a state of statistical control prior to assessing capability; however one point outside of the spec limits is not an indication of instability. again helpful to see your data
 
V

vbhatta

#3
Hello Bev,

Thanks for a quick response!

K-factor analysis is used in our facility in lieu of a plain +/- 3-sigma tolerance interval. Here are a couple of good primer on Tolerance Interval-

7.2.6.3. Tolerance intervals for a normal distribution
Statistical Tolerance Intervals | Quality Digest


The idea is that, instead of computing average and std. dev. and then using +/- 3 Std. Deviations to figure out your process spread, the K-factor (or K-multiplier) depends on your sample size. Higher the sample size, lower the K-multiplier is. For the same sample size, the K-factor is different for a one-sided Vs. a two-sided specification limit.

For this analysis, once we have computed Average and Std. Deviation from collected data, we have to figure out what K-factor to use. To get this K-factor value, we have to know what Confidence/Reliability (C/R) value we need to use (which in turn is derived based on the Severity/Occurrence values one has on their FMEA- most companies have such a matrix prepared in their site) and we also need to know the sample size. Then it's just a matter of going to this table and looking for a K-factor value corresponding to the C/R value and sample size. Once K-factor is know, you simply multiply it by the standard deviation and then use it for your "Average +/- K x Sigma" equation to compute the process spread and compare it to your specification limit to ensure that this computed Tolerance Interval is smaller than your specification limit.
 

Attachments

Last edited by a moderator:
Thread starter Similar threads Forum Replies Date
T 4 Factor Optimization using Minitab Using Minitab Software 7
C Design of Experiments or Factor analysis using existing data Quality Tools, Improvement and Analysis 7
S Process Monitoring using SPC software Quality Assurance and Compliance Software Tools and Solutions 6
R Using Scrapped NC items for training Medical Device and FDA Regulations and Standards News 14
M What is the Risk of Using Obsolete Versions of C=0 & ANSI/ ASQ Z1.4 Sampling Plans? ISO 13485:2016 - Medical Device Quality Management Systems 8
F Using non-randomized clinical study for change in existing 510(k) device? US Food and Drug Administration (FDA) 1
J Using MD on a flight EU Medical Device Regulations 9
K Using Google Drive/Docs ISO 13485:2016 - Medical Device Quality Management Systems 16
J What risk to cover when NOT using ISO 17025 accredited/certified labs for calibration ISO 17025 related Discussions 3
T Poisson regression using Minitab Statistical Analysis Tools, Techniques and SPC 3
T Using NICE for guidance during CER CE Marking (Conformité Européene) / CB Scheme 0
T Using Review articles from Pubmed for CER CE Marking (Conformité Européene) / CB Scheme 8
Tagin Using a meme generator to poke fun at quality Funny Stuff - Jokes and Humour 2
D Finding Optimum Design Parameters using Taguchi method? Using Minitab Software 2
V MSA - Using a R&R study for a Tape Measure Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
D Using electronic lab notebooks in the design control process ISO 13485:2016 - Medical Device Quality Management Systems 3
M Using a Non Certified vendor Supplier Quality Assurance and other Supplier Issues 5
H Using/Selling power banks with your product as a "mobile version" CE Marking (Conformité Européene) / CB Scheme 3
cgaro62 Using internal audits as part of document reviews Document Control Systems, Procedures, Forms and Templates 21
R Select the 1 Supplier based on the Parts Durability from 6 Supplier Samples using Minitab Using Minitab Software 11
M MSA angle issue - Nylon parts using Aberlink 3D Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
D Question on using audit checklist ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 20
J Using of CE marking - Two CE markings on the product CE Marking (Conformité Européene) / CB Scheme 5
R Using R package to implement Bayesian phase I/II dose-finding design for three outcomes ISO 13485:2016 - Medical Device Quality Management Systems 6
S Defibrillator Protection applicability When using Multiple applied parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
P Using Zinc Plated Steel in a Cleanroom Other Medical Device Related Standards 3
J Incoming Inspection Records using Excel File ISO 13485:2016 - Medical Device Quality Management Systems 2
J Using ring gauges General Measurement Device and Calibration Topics 2
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
P Can Neoprene be Cleaned Using Isopropyl Alcohol (IPA) Manufacturing and Related Processes 4
GreatNate Anyone using the Intellect QMS software? Quality Assurance and Compliance Software Tools and Solutions 1
chris1price Sterilization using beta radiation Other Medical Device Related Standards 2
M Using the phrase "herein referred to" Document Control Systems, Procedures, Forms and Templates 8
B Struggling with using the 5.6 version Ford Capacity Analysis Report APQP and PPAP 5
cnbrosa Study Type 1 on a CMM using a measuring support Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
rerusk1 MRB (Material Review Board) Process using MS Sharepoint or MS Teams Manufacturing and Related Processes 2
K 510k FDA review, will they accept Biocompatibility result generated using feasibility product lots? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
U Medical Device CE Marking - Using a disposable bearing CE Marking (Conformité Européene) / CB Scheme 3
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 3
D Risk Analysis using Monte Carlo Simulation instead of Scoring and Heat Map Risk Management Principles and Generic Guidelines 2
W Using tailoring guidelines to tailor a QMS procedure ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
Y We found out we have been using a equipment without validation for past 4 years Quality Manager and Management Related Issues 6
D Using Laboratory Notebooks in R&D and Design and Development ISO 13485:2016 - Medical Device Quality Management Systems 3
C Importer shell game - Using a third party logistics provider (3PL) in the EU EU Medical Device Regulations 5
S Work performed in Canada on US patients using US device Canada Medical Device Regulations 1
S Is using ANSI/ASQ Z1.4-2008 the correct sampling plan to determine Pass/Fail of Apparel measurements? AQL - Acceptable Quality Level 8
A What are the pros and cons of using an audit software for internal auditing? General Auditing Discussions 7
Tagin Evaluating nonconformances for escalation using Bayesian methods? Statistical Analysis Tools, Techniques and SPC 2
D Using non-conforming components even though the final assembly is conforming? Manufacturing and Related Processes 5

Similar threads

Top Bottom