Using Laboratory Notebooks in R&D and Design and Development

duff999

Involved In Discussions
#1
Question regarding the use of laboratory notebooks in design and development - does anyone have information on using Lab notebooks in design and development and will the use of the lab notebook support the verification process or do I need to create a separate verification plan, protocol and test summary report?

Any feedback is appreciated.
 
Elsmar Forum Sponsor

Tidge

Trusted Information Resource
#2
My preference would be that you have an approved verification plan which allows (among other methodologies) the acceptance of lab notebook contents as evidence of verification provided that the contents are subject to an independent (and critical) review. If the lab notebooks are subject to appropriate controls and the documentation within them is not suspect in anyway (records are cleanly made by an identified person on a specified date, records aren't 'back-corrected', etc.) then it is possible they might serve as evidence of verification.

The lab notebooks would become part of the Design History File and would be subject to all the necessary documentation controls for such records.

Generally, I would prefer that protocols be created and approved prior to verification testing. Verification testing is also intended to be done on established/released/controlled designs with validated test methods, specified (and if necessary calibrated) equipment, and trained testers.

I've kept lab notebooks myself (for a variety of medical device compliance-adjacent reasons) and I would prefer that people not poke around in my notebooks as objective evidence. I've written up analyses based on my notebooks for consumption by folks looking for something specific, and let those write-ups be my official account. To my knowledge such reports have never been used as design verification.
 

duff999

Involved In Discussions
#3
Thank you for your response, you've made some very good points that I did not think about. The lab notebooks are subject to document control but currently we don't approve the protocols prior to running the experiments, these are documented, and can not be back dated, but as far as getting the verification plan and protocols approved this is not happening as it would under similar design controls. Perhaps I could use the findings from the lab notebooks to feed into the verification plan/protocols
 

yodon

Staff member
Super Moderator
#4
Completely agree with @Tidge but if you do want to use lab notebooks, you have to really kick it up a few notches to ensure data integrity. I wouldn't do it for design V&V.

Perhaps I could use the findings from the lab notebooks to feed into the verification plan/protocols
Without the pre-approved protocols with the pre-determined acceptance criteria, you run the risk of giving the impression that you tested into compliance. You'd probably need to take extraordinary measures to avoid that impression. Again, I wouldn't do it.

We've used lab notebooks before but only when we wanted to capture early design notes in the event we needed to support patent application claims (and had to jump through the data integrity hoops when doing so).
 
Thread starter Similar threads Forum Replies Date
rob73 Requirement for independent certification of device using an ISO 17025 laboratory EU Medical Device Regulations 2
I Laboratory for Checking Tensile Strength using UTM (Universal Tensile Machine) General Measurement Device and Calibration Topics 13
R Using R package to implement Bayesian phase I/II dose-finding design for three outcomes ISO 13485:2016 - Medical Device Quality Management Systems 6
S Defibrillator Protection applicability When using Multiple applied parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
P Using Zinc Plated Steel in a Cleanroom Other Medical Device Related Standards 3
J Incoming Inspection Records using Excel File ISO 13485:2016 - Medical Device Quality Management Systems 2
J Using ring gauges General Measurement Device and Calibration Topics 2
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
P Can Neoprene be Cleaned Using Isopropyl Alcohol (IPA) Manufacturing and Related Processes 4
GreatNate Anyone using the Intellect QMS software? Quality Assurance and Compliance Software Tools and Solutions 1
chris1price Sterilization using beta radiation Other Medical Device Related Standards 2
M Using the phrase "herein referred to" Document Control Systems, Procedures, Forms and Templates 8
B Struggling with using the 5.6 version Ford Capacity Analysis Report APQP and PPAP 5
cnbrosa Study Type 1 on a CMM using a measuring support Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
rerusk1 MRB (Material Review Board) Process using MS Sharepoint or MS Teams Manufacturing and Related Processes 2
K 510k FDA review, will they accept Biocompatibility result generated using feasibility product lots? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
U Medical Device CE Marking - Using a disposable bearing CE Marking (Conformité Européene) / CB Scheme 3
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 3
D Risk Analysis using Monte Carlo Simulation instead of Scoring and Heat Map Risk Management Principles and Generic Guidelines 2
W Using tailoring guidelines to tailor a QMS procedure ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
Y We found out we have been using a equipment without validation for past 4 years Quality Manager and Management Related Issues 6
C Importer shell game - Using a third party logistics provider (3PL) in the EU EU Medical Device Regulations 5
S Work performed in Canada on US patients using US device Canada Medical Device Regulations 1
S Is using ANSI/ASQ Z1.4-2008 the correct sampling plan to determine Pass/Fail of Apparel measurements? AQL - Acceptable Quality Level 4
A What are the pros and cons of using an audit software for internal auditing? General Auditing Discussions 7
Tagin Evaluating nonconformances for escalation using Bayesian methods? Statistical Analysis Tools, Techniques and SPC 2
D Using non-conforming components even though the final assembly is conforming? Manufacturing and Related Processes 5
B Using Unreleased Documents & Process Maps for Internal Audit purposes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
R Clause 7.7 Replicate, Recalibration and Intermediate checks using Artifact ISO 17025 related Discussions 1
Stoic Are any medical device companies using the 2011 FDA process validation guidance instead of GHTF/SG3/N99-10:2004? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
B Unit of Use DI (Device Identifier) - Products using the same device US Food and Drug Administration (FDA) 0
A Using Arduino based sensors for Poke-Yoke Manufacturing and Related Processes 6
M Using your Manufacturer's ISO certification ISO 13485:2016 - Medical Device Quality Management Systems 5
Ed Panek Adverse Event Clinical Trial using a 510K approved Device Other US Medical Device Regulations 6
B Using non CE parts in a machine CE Marking (Conformité Européene) / CB Scheme 1
G Gauge R&R on multiple dimensions using 3D measurement system Capability, Accuracy and Stability - Processes, Machines, etc. 6
A Hospital IT expectations for connected medical device using WIFI Medical Information Technology, Medical Software and Health Informatics 0
Proud Liberal Cp / Cpk on position using multiple MMC bonuses Capability, Accuracy and Stability - Processes, Machines, etc. 2
B Using external FDA and ISO 13485 audit as internal audit Internal Auditing 6
D Using "Particle Size Standard" templates as gauges - How to avoid giving a gauge # while using for process control? General Measurement Device and Calibration Topics 2
J Class 1 Medical Device - Using a UPC over the UDI? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
I Reducing CE marking cost using manufacturer test reports CE Marking (Conformité Européene) / CB Scheme 5
CPhelan Using clinical trial safety data for evidence for CE marking EU Medical Device Regulations 8
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate Qualification and Validation (including 21 CFR Part 11) 3
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate General Measurement Device and Calibration Topics 2
M Raw stock material testing discrepancy using an XRF (x-ray fluorescence) analyzer Manufacturing and Related Processes 7
Z Using FMEA for Knowledge Management FMEA and Control Plans 6
Sidney Vianna NASA to Develop a Novel Approach for All-Electric Aircraft Using Cryogenic Liquid Hydrogen as Energy Storage World News 2
S How many of you are using Robotic process automation for calibration lab management? ISO 17025 related Discussions 0

Similar threads

Top Bottom