UUT Accuracy as part of Uncertainty Budget

itsbiodiversity

Involved In Discussions
Can someone please explain to me why I would use the UUT tolerances in the uncertainty budget? Specifically for a 5.5digit DMM.

I am currently including:
  • fluke 1281 uncertainty
  • resolution of uut
  • repeatability

I'm not too concerned with smaller effects, but any guidance would be great.

I know I come across some who put tolerances in the budget and others that do not. My issue is this...

If the tolerance is say 1V, and 1V enters my equation divided by the sqroot3, then multiplied by k=2, am I not 100% of the time going to be above my tolerance limits for MU vs Tolerance of the UUT?

Thank you so much, as I have picked up a lot through this forum through the years.
 

BradM

Leader
Admin
I haven't seen the tolerance of the UUT included in the budget. Typically the budget you are estimating is to determine the error contributing to the calibration of the UUT. Thus, I would see the UUT error as independent of the standard uncertainty.
 

dgriffith

Quite Involved in Discussions
UUT tolerance specification is not normally part of the calibration uncertainty budget. The tolerance is the limit(s) against which the uncertainty of the calibration system is measured against to make a decision, such as pass/fail. etc.
The tolerance specification would be used when the meter becomes part of a calibration/test system used on other equipment.
Check with the mfr. of the meter--you may find that their spec is already at k=2 or 3, in which case you would not divide by sqr3.
 

itsbiodiversity

Involved In Discussions
Do you include both the spec and uncertainty of the standard or just the uncertainty? I have seen it stated I believe that if you have the uncertainty of the standard then you do not include the tolerance of the standard in the budget.
 

dgriffith

Quite Involved in Discussions
Do you include both the spec and uncertainty of the standard or just the uncertainty? I have seen it stated I believe that if you have the uncertainty of the standard then you do not include the tolerance of the standard in the budget.
Confused. Are you now talking about after cal when you are using it, or still during calibration? If after, then do one or the other. Sometimes you don't know what the uncertainty of the calibration for your MTE was, so you use the specification for the measurement, e.g. ppm range, ppm reading, floor, if any.
Many meter manuals give you an example of the unc/error/tol for a given measurement. That result would be divided by the k factor as mentioned previously to get the standard uncertainty, which is then added to the other budget contributors in the usual way (GUM, et al).

If you know what the uncertainty for the calibration is/was, you can use it instead, but be careful. What kind of MTE are you talking about? If a DMM, that might be more difficult. If a resistor standard with a single assigned value, then use the uncertainty value stated on the cert. My opinion.
 

itsbiodiversity

Involved In Discussions
It was a DMM. I believe I need both the accuracy of my standard and the uncertainty of the standard along with the resolution of the UUT/Std, and repeatability.
 

dgriffith

Quite Involved in Discussions
It was a DMM. I believe I need both the accuracy of my standard and the uncertainty of the standard along with the resolution of the UUT/Std, and repeatability.
You do not need both the accuracy and the uncertainty, nor do I think you can have both, of a standard when performing a calibration--much too conservative, leading to increased workload due to false reject. It's either a tolerance spec or an uncertainty.

The manufacturer will usually state a tolerance (acceptable performance limits (accuracy)) or an uncertainty (estimated error for a parameter) at some coverage factor.

Example: A 100psia pressure controller tolerance might be ?0.01% Full Scale at k=2. This becomes 0.0001*100=0.01 psi/k=?0.005psi standard uncertainty for every reading on the scale. As you can see, the lower the reading, the greater the potential error--0.005 is a much larger part of 1 than it is of 100.

Example: Fluke 8508A ref. dmm; uncertainties are listed for k=2 and k=3 for range and reading of each parameter; 10Vdc on 20V range might be [(50ppm rdg+35ppm rng)/k]=?[(0.0005+0.0007)/k]Vdc standard uncertainty for the specific parameter.

Only errors reduced to standard uncertainties can be summed together before expanding with coverage factor, whether performance specifications, resolution, repeatability, et al.
 
Thread starter Similar threads Forum Replies Date
E How does Non-Linear UUT affect TAR (Test Accuracy Ratio) Calculation? Measurement Uncertainty (MU) 3
C Accuracy on calibration certificates - How to calculate the accuracy of UUT General Measurement Device and Calibration Topics 3
I Uncalibrated source to calibrated meter, then compare to UUT General Measurement Device and Calibration Topics 4
A Do you include UUT Specification in Uncertainty of Calibration? Measurement Uncertainty (MU) 6
C UUT resolution and reporting uncertainty for each reading Measurement Uncertainty (MU) 2
J Test Uncertainty Ratio TUR (UUT Tolerance / STD Uncertainty) Question General Measurement Device and Calibration Topics 8
Z How to interpret and apply IEC60601-2-54, clause 203.6.4.3.104.3 Accuracy of X-RAY TUBE VOLTAGE, second para. IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
R Accuracy tests for enteral nutrition pump IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
F IEC 60601-2-54:2009 Accuracy of CURRENT TIME PRODUCT IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
CaliperJim Uncertainty and Accuracy Requirement Deviation Provision General Measurement Device and Calibration Topics 2
D IVD Accuracy and Precision Acceptance Criteria ISO 13485:2016 - Medical Device Quality Management Systems 1
N Accuracy & Precision ISO 17025 related Discussions 2
C ISO 80601-2-56 Laboratory accuracy Other Medical Device Related Standards 7
J Setting accuracy limit for a read out General Measurement Device and Calibration Topics 4
R Clinical accuracy and repeatability of IR(infrared) thermometer, no maximum error criteria is recommended in ISO Other Medical Device Related Standards 11
R SAT (System Accuracy Test) temperature according to AMS2750E AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
H Analog Pressure Gauge with no EN837 accuracy class information General Measurement Device and Calibration Topics 5
shimonv IEC 60601-1 Essential Performance - Is the signal accuracy specification an essential requirement? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
G Warning about Deltronic Gage Pins - The "Certification of Accuracy" document General Measurement Device and Calibration Topics 13
G Posting Measuring Equipment Accuracy for User Information General Measurement Device and Calibration Topics 4
S Gagemaker MT-3000 accuracy for calibration Calibration and Metrology Software and Hardware 2
J CMM Accuracy Formula General Measurement Device and Calibration Topics 3
L Calibration Accuracy In Relation To Reporting General Measurement Device and Calibration Topics 1
Q Acceptable calibration accuracy of a 60" linear measuring device General Measurement Device and Calibration Topics 16
M Calibration certificates avowing accuracy beyond the uncertainty limits? Measurement Uncertainty (MU) 7
S Jewelry vs. Normal Laboratory Balances - Accuracy and calibration General Measurement Device and Calibration Topics 2
T Tape Measure, the right device for 2 or 3 decimals accuracy on prints General Measurement Device and Calibration Topics 8
W Equipment for Oven Calibration, Uniformity and System Accuracy Tests AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
S Opinions on the "Best" CMM for High Accuracy Machining General Measurement Device and Calibration Topics 9
Q AMS 2750E System Accuracy Test (SAT) Frequency AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
A Process to Validate the Accuracy of Test Reports (AS9100) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 38
D AMS 2750E - Digital Instrument Calibration Accuracy Requirements General Measurement Device and Calibration Topics 1
shimonv Soldering Station Accuracy out of specification 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M Help trying to explain accuracy procedure General Measurement Device and Calibration Topics 1
D Long length measurement of 18" up to 100" with an accuracy of .005 min General Measurement Device and Calibration Topics 8
N Cry for help - FDA calibration requirement for accuracy & precision General Measurement Device and Calibration Topics 2
C Accuracy Equation - Linear gage with a stated accuracy of (1.5+L/50) ?m L= mm General Measurement Device and Calibration Topics 5
W Dimensional Measurement of 10% accuracy and 25% uncertainty standard General Measurement Device and Calibration Topics 6
A Uncertainty Measurement - What this accuracy specification "1% + 10^5": means? General Measurement Device and Calibration Topics 2
L Inspection Equipment Accuracy - Measuring the diameters of small holes Measurement Uncertainty (MU) 5
G Digital Autoclave Accuracy Requirements ISO 13485:2016 - Medical Device Quality Management Systems 2
S AMS 2750E Heat Treat Furnace System Accuracy Test SAT Report question AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
A Accuracy for electrosurgical analyzer General Measurement Device and Calibration Topics 2
A Calculating Accuracy for a Multimeter General Measurement Device and Calibration Topics 1
D Verification of weighing scale 150kg capacity 0.05g accuracy General Measurement Device and Calibration Topics 9
A Accuracy Calculation of Leakage Current for an Electrical Safety Analyzer General Measurement Device and Calibration Topics 5
G Laser Precision Linear Measurement - Inherrent Accuracy of 10 -6 - Nice PDF Paper General Measurement Device and Calibration Topics 1
S Stated Accuracy of Portable CMM Arm and Real World Use Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
A Intrinsic Uncertainty and Accuracy of a Medical Electrical Safety Analyzer General Measurement Device and Calibration Topics 1

Similar threads

Top Bottom