Validating a Process Based on Lot Release Criteria

G

GTJunghoon

#1
I'm validating a liquid filling process. The risks associated with the product failure have been assessed and preventative measures have been put in place. So, I am confident that the product will rarely fail the QC inspections.

It appears that the Jefes' want to eventually release a product lot of up to 1400 items after QC inspects and passes 2 items selected from that same product lot. I suppose this means I have to prove that the process is capable of this "lot release criteria".


How do I statistically prove that the defect rate of a process is so low that a lot of 1400 items is released after inspecting 2? :mad:
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#3
I'm validating a liquid filling process. The risks associated with the product failure have been assessed and preventative measures have been put in place. So, I am confident that the product will rarely fail the QC inspections.

It appears that the Jefes' want to eventually release a product lot of up to 1400 items after QC inspects and passes 2 items selected from that same product lot. I suppose this means I have to prove that the process is capable of this "lot release criteria".


How do I statistically prove that the defect rate of a process is so low that a lot of 1400 items is released after inspecting 2? :mad:
Hi,

It seems that according to the AQL-based methodology you would need to sample at least 5 units for a lot size of 1400. If all 5 are pass, the lot passes. If one fails, the lot fails. This regime correlates with an AQL not worse than 2.5 (Normal inspection).

Cheers,
Ronen.
 

somashekar

Staff member
Super Moderator
#4
I'm validating a liquid filling process. The risks associated with the product failure have been assessed and preventative measures have been put in place. So, I am confident that the product will rarely fail the QC inspections.

It appears that the Jefes' want to eventually release a product lot of up to 1400 items after QC inspects and passes 2 items selected from that same product lot. I suppose this means I have to prove that the process is capable of this "lot release criteria".


How do I statistically prove that the defect rate of a process is so low that a lot of 1400 items is released after inspecting 2? :mad:
What is the container made of ... Is it transparent ... Do you have some sort of a scale marked on the container as it gets processed (comes from your design)
How does QC inspect ?
If I had it as said above, I would not be much bothered, as QC can inspect more samples based on your AQL selection easily.
Validation is a exercise done prior to putting the liquid filling process into use. Lot release criteria is more a check of your process effectiveness / Validation effectiveness.
 
Thread starter Similar threads Forum Replies Date
S Validating a Bonding Process Qualification and Validation (including 21 CFR Part 11) 3
G Validating a Sandblasting Process - Any pointers? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
Q Validating the AOI process (Automated Optical Inspection) Reliability Analysis - Predictions, Testing and Standards 9
N Validating Software before getting approved as Class 2 device US Food and Drug Administration (FDA) 5
Ed Panek Validating Excel (XLS) formulas 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
A Validating my own application developed with Agile Qualification and Validation (including 21 CFR Part 11) 5
T Risk analysis of QMS software - Validating software we use for QMS ISO 13485:2016 - Medical Device Quality Management Systems 5
H Validating Nondestructive leak testing machine Qualification and Validation (including 21 CFR Part 11) 3
M FDA's expectation for validating OTS software updates Other US Medical Device Regulations 8
R Re-validating revised Medical Device and Aerospace Product COTS software ISO 13485:2016 - Medical Device Quality Management Systems 2
J Validating a stopped test due to never finding any defects Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
J Validating multiple injection molding machines efficiently ISO 13485:2016 - Medical Device Quality Management Systems 5
J Grouping Products and Testing Sterile and Biocompatibility - Validating Products Other Medical Device Regulations World-Wide 3
Q Meeting the Requirement of Clause 7.5.2 on Validating Special Processes AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
M Validating a Product through Tool Control and Tool Inspections Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
G Testing and Validating a New Vertical Milling Machine AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 10
Q Validating Tape Measures used to Measure Product Conformity to Requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Protocol for Validating Pull Tester and Force Gage Qualification and Validation (including 21 CFR Part 11) 9
T Validating Supplier Statement - Rusty shield that supplier denied responsibility Manufacturing and Related Processes 7
M Validating 3rd Party Software which Tracks Archived Samples Qualification and Validation (including 21 CFR Part 11) 2
Q Validating ESD (Electrostatic Discharge) Flooring ISO 13485:2016 - Medical Device Quality Management Systems 2
M Procedure for Validating Laser Marking & Electropolishing of Orthopedic Implants Other Medical Device and Orthopedic Related Topics 13
K Validating a gage/fixture in-house under TS 16949 - Need information General Measurement Device and Calibration Topics 2
C Validating Custom adaptations of Medical Devices ISO 13485:2016 - Medical Device Quality Management Systems 11
C Validating a Gage R&R spreadsheet - Defining Formulas in spreadsheet cells Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
G Software Validating Adobe Professional according to Part 11? Quality Assurance and Compliance Software Tools and Solutions 10
D Validating software in materials testing - Integrated system of individual components General Measurement Device and Calibration Topics 3
D Validating Multimeters (how to) General Measurement Device and Calibration Topics 8
B Toyota PPAP Process - Three Questions APQP and PPAP 3
R Changes vs CMO - How can we simplify this process? Supplier Quality Assurance and other Supplier Issues 3
A Ethics Committee Review Process for IVD Products EU Medical Device Regulations 2
V Laser Welding Process - Impact on Electrical Properties Reliability Analysis - Predictions, Testing and Standards 4
Q Process: Knowledge Section 7.1.6 of ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
L Documented Information in Internal Audits Process (9.2) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
A Sampling plan for in-process QC (medical devices) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
R MRB (Material Review Board) Process using MS Sharepoint or MS Teams Manufacturing and Related Processes 2
M Clinical Benefit of device that only aids in a process for managing or treating disease EU Medical Device Regulations 2
C In-process inspection - Tooling and assembly lines for automotive companies AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
M Efficacy of an IT process after a cyber attack ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
N Riveting - special process Manufacturing and Related Processes 11
M Material incoming to the production process reflected in PFMEA FMEA and Control Plans 9
A API Spec Q1 Purchasing Process - Supplier Reevaluation based on Supplier Risks 5.6.1.4 Oil and Gas Industry Standards and Regulations 15
B Handling lower detection limits for SPC and process performance Statistical Analysis Tools, Techniques and SPC 1
D Measurables for Plastic Injection molding process Manufacturing and Related Processes 1
S Cleaning process center change ISO 13485:2016 - Medical Device Quality Management Systems 4
Z Rapid audit template for plastic parts manufacturing process Manufacturing and Related Processes 12
R Inspection and Work order process Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
T ISO 13485:2016 Clauses related to process matrix ISO 13485:2016 - Medical Device Quality Management Systems 3
A How to reduce the process SPC monitoring Capability, Accuracy and Stability - Processes, Machines, etc. 3

Similar threads

Top Bottom