Validation and Verification - Catheter into a tyvek pouch
T
theGrod
#1
My company puts a catheter into a tyvek pouch, seals it and then sends it off for sterilization. We have run out of 9"X9" pouches and are going to use 8"x10" pouches all made from the same material by the same supplier and sterilized using the same process. Regulatory affairs says a validation and verification study must be done in order to use a different size pouch. According to the development engineer a larger pouch would actually have better sterilization performance due to the larger surface area to allow the EO gas to penetrate the pouch. We're being quoted ISO11607 as the rationale for the V&V study. Any thoughts. As a straight quality guy I fail to see the problem with this as it is the same product, same material, same specs, just a different size.
Elsmar Forum Sponsor
Ronen E
Problem Solver
Moderator
#2
...however not validated in the context of your specific process. For instance, the bigger pouches might be manufactured (by your supplier) using different equipment and different settings. Are you going to use the very same equipment for sealing? How would you determine the settings?
Start off with a risk assessment. Perhaps you could get away with a limited scope validation.
Cheers,
Ronen.
AgnieszkaSz
Involved In Discussions
#3
I know from experience that the customer requires validation for every size of the pouch, particularly because the seal is different size. We use sealing machine with different settings for each seal size. The seal is tested for breaking and for tightness.
Thread starter Similar threads Forum Replies Date
K ISO 17025 Method Validation and Verification for Test Lab ISO 17025 related Discussions 4
G Injection Molded Parts in Verification & Validation Other Medical Device and Orthopedic Related Topics 3
M Root Cause and Corrective Action for CAPA's lacking validation/verification ISO 13485:2016 - Medical Device Quality Management Systems 19
G Pad Printing Validation OR Verification ISO 13485:2016 - Medical Device Quality Management Systems 4
G How to Record Informal Testing (Not Verification/Validation) Other Medical Device and Orthopedic Related Topics 15
R Validation of Software used in Verification Testing ISO 13485:2016 - Medical Device Quality Management Systems 2
R SaMD Verification & Validation IEC 62304 - Medical Device Software Life Cycle Processes 6
M Validation of Data verification tool per 21 CFR 820 Quality Assurance and Compliance Software Tools and Solutions 1
M Software verification and validation AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
C 8.3.4 Verification and Validation ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
P Software verification and validation procedure IEC 62304 - Medical Device Software Life Cycle Processes 6
Y SaMD Verification and Validation SOP and Plan IEC 62304 - Medical Device Software Life Cycle Processes 8
N Example for design and development planning,input,output,review,verification,validation and transfer Misc. Quality Assurance and Business Systems Related Topics 4
T Laboratory Verification after validation ISO 17025 related Discussions 3
P Design verification driven by new equipment. How is this different than process validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
G Devices from IQ, OQ or PQ process to be used for verification, validation and summative? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
Bev D Verification and Validation of Measurement Systems Misc. Quality Assurance and Business Systems Related Topics 0
A Our auditor told if we didn't have a patent we would have to do a validation or verification ISO 13485:2016 - Medical Device Quality Management Systems 6
N Design Verification & Process Validation - Statistical sample sizes Design and Development of Products and Processes 2
D 510K and Changes to Verification and Validation US Food and Drug Administration (FDA) 2
Ed Panek Label verification and validation US Food and Drug Administration (FDA) 5
V Which batches should or could be considered for design validation and design verification? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
I Medical device Validation/Verification template ISO 13485:2016 - Medical Device Quality Management Systems 1
M Does anyone have a good verification and validation plan template? ISO 13485:2016 - Medical Device Quality Management Systems 8
S Source of QMS templates including templates for design verification/design validation Document Control Systems, Procedures, Forms and Templates 2
Ed Panek Verification & Validation requirements - Patch that adheres to skin ISO 13485:2016 - Medical Device Quality Management Systems 2
M Bringing blending in-house - Verification or full validation? (Disinfectants) ISO 13485:2016 - Medical Device Quality Management Systems 3
A Level of details required for Class IIb Product Verification and Validation CE Marking (Conformité Européene) / CB Scheme 1
S CE Mark Approval Design Verification and Validation Questions CE Marking (Conformité Européene) / CB Scheme 8
S What is the difference between Verification and Validation of Test Reports? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
M IEC 62366-1:2015 Usability Verification / Validation Plan example wanted IEC 62366 - Medical Device Usability Engineering 20
shimonv System Verification & Validation in HC Class III Submission Canada Medical Device Regulations 2
G Verification or Validation for Glue Dispenser Units ? Qualification and Validation (including 21 CFR Part 11) 4
D User Requirements Tracing - Verification and Validation Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Validation before Verification. Why not? ISO 13485:2016 - Medical Device Quality Management Systems 3
H Is design verification and validation required for samples accompanying the quotes ? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
E How to write requirements for Medical App Validation and Verification IEC 62366 - Medical Device Usability Engineering 7
Chennaiite Cause Analysis for Verification/Validation Failures - Seeking Opinions Problem Solving, Root Cause Fault and Failure Analysis 3
F Usability Objectives and Verification & Validation Plan IEC 62366 - Medical Device Usability Engineering 7
S Verification and Validation of Post Market Design Change Design and Development of Products and Processes 2
M Validation/Verification Trace matrix - examples or recipe? Document Control Systems, Procedures, Forms and Templates 1
T Class II Software Device 510k V&V (Verification and Validation) Criteria and Results 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
V Utility (power production) Company - Mitigating risks in Verification and Validation. ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
A Should the FSMS have a Documented Generic Procedure for Verification and Validation Food Safety - ISO 22000, HACCP (21 CFR 120) 2
P What is Validation, Verification and Improvement of FSMS in ISO 22000 Clause 8? Food Safety - ISO 22000, HACCP (21 CFR 120) 2
P How to separate Product Design Review, Verification, and Validation Activities ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 26
T Medical Device Software Verification & Validation for Client Specific Configurations Design and Development of Products and Processes 2
Q Verification & Validation Project Time Line Qualification and Validation (including 21 CFR Part 11) 5
M Design and Development Outputs Approval, Verification and Validation IATF 16949 - Automotive Quality Systems Standard 4
G Legacy Product - Do we need follow QSR 820? Design Verification and Validation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 12
Similar threads
Top Bottom