Validation of a change to a CE Marked Medical Device

JoCam

Trusted Information Resource
#1
Hello,

A quick question about validation requirements for a change to a CE marked electrical device.

The electrical device in question is a Class 11b, and at present is battery powered. The mains powered battery recharging station supplied, together with the battery powered device have both been electrically tested.

We are looking to replace the battery system and move across to mains powered adaptor. Whilst we can see numerous requirements for verification of this change, are there any requirements for validation? We fully validated the battery powered device (with mains adaptor charging station) to obtain the CE mark, would we be expected to repeat all this validation as a result of this change?

Your help with this will be very much appreciated.

Regards,

Jo
 
Last edited:
Elsmar Forum Sponsor

mihzago

Trusted Information Resource
#2
Re: Validation of a change to a medical device

what was the reason for this change? e.g. to prevent use errors, to decrease risk, lower cost, etc? did you meet that goal?

did you check whether this change will meet the needs of your users?

The validation is also any testing by a test lab (safety/electrical, mechanical) because the test is performed on a production equivalent unit and simulates actual conditions.
 

JoCam

Trusted Information Resource
#3
Re: Validation of a change to a medical device

Thank you mihzago,

The reason for the change is in response to end-users requesting a non-battery version. We will still be manufacturing the battery powered device, until hopefully the demand falls away due to replacement with the mains powered.

We are aware that we will need to perform safety/electrical, mechanical testing but aren't sure if there are any validation requirements. I believe that providing evidence that (following satisfactory safety testing) it performs exactly as the battery powered device should satisfy the regulatory requirements.
 

yodon

Staff member
Super Moderator
#4
Re: Validation of a change to a medical device

You said you fully validated the battery-powered version - so what did you do for that? Do an analysis to determine if the change impacts those results. (Document your work, of course). Then you should only need to do what's impacted and you have documented rationale for not doing the other. (Of course, you may find, in updating your risk analysis, etc. that additional efforts are required anyway but you may be able to limit what gets re-done).
 
B

Burgmeister

#5
Re: Validation of a change to a medical device

One further point to consider - your Notified Body will need to review the changes and, if they are anything like ours, there will be a looooong waiting list to get the changes reviewed. Contact them as soon as possible to arrange a date for the review otherwise you will complete the project and then have to wait 8 months or so for them to look at it and issue a CE mark.
 
Last edited by a moderator:
Thread starter Similar threads Forum Replies Date
S Forced ServiceNow validation - No change in our current user and functional requirements IT (Information Technology) Service Management 6
K Re-validation of EtO sterilization after change of tape Other Medical Device and Orthopedic Related Topics 1
S Verification and Validation of Post Market Design Change Design and Development of Products and Processes 2
C Change Control Forms Post Software Validation Medical Information Technology, Medical Software and Health Informatics 2
E Help with ECR/DCR Document Change Control Software Validation Qualification and Validation (including 21 CFR Part 11) 6
G Sterilization Re-Validation after Manufacturer Change Qualification and Validation (including 21 CFR Part 11) 17
R Equipment Change or Process Change? Validation protocol for a planar freezer Misc. Quality Assurance and Business Systems Related Topics 3
S Tool validation / Tool change over program Quality Tools, Improvement and Analysis 3
A Engineering change validation - Companies that have no design responsibility Design and Development of Products and Processes 5
D Engineering Change Validation - 4.2.4.3 QS-9000 - American Automotive Manufacturers Standard 7
S Rationalising the level of effort and depth of software validation based on risk ISO 13485:2016 - Medical Device Quality Management Systems 6
D Questions regarding process validation ISO 13485:2016 - Medical Device Quality Management Systems 5
Y We found out we have been using a equipment without validation for past 4 years Quality Manager and Management Related Issues 6
Z Is IQ necessary for laser marking validation? EU Medical Device Regulations 3
E 13485:2016, Sections 4.1.6, 7.5.6 and 7.6 - Validation of Software - Need some Advice please ISO 13485:2016 - Medical Device Quality Management Systems 2
A Validation of Forced Aeration Process ISO 13485:2016 - Medical Device Quality Management Systems 3
E Mentor for Test Method Validation (TMV) Design and Development of Products and Processes 2
M API 4F/7K/8C Design Package Validation Oil and Gas Industry Standards and Regulations 1
I ISO 2233:2000 Question - Medical Device Shipping/Transportation Validation Other ISO and International Standards and European Regulations 1
T Annual Validation as a detection mode on a PFMEA? FMEA and Control Plans 5
B TMV - Selection of TM's for Validation ISO 13485:2016 - Medical Device Quality Management Systems 5
P Human Factors / Usability validation in the time of COVID Human Factors and Ergonomics in Engineering 9
C Template for Excel Validation Reliability Analysis - Predictions, Testing and Standards 5
M IT validation for a paper based MD repair company QMS ISO 13485:2016 - Medical Device Quality Management Systems 6
P Unrealistic Packaging Validation Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 13
D Test summary report example for design validation wanted - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
M Is Validation of Plating Processes required and who is responsible? Qualification and Validation (including 21 CFR Part 11) 11
T ISO 13485 - Process validation at critical suppliers ISO 13485:2016 - Medical Device Quality Management Systems 7
K Software Validation for Measurement Tools used in Process Validation ISO 13485:2016 - Medical Device Quality Management Systems 2
Stoic Manual soldering processes - 100% verifiable, or always requiring validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
P Design verification driven by new equipment. How is this different than process validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Rees System Validation Qualification and Validation (including 21 CFR Part 11) 1
K PQ validation qualification - Asked to write a PQ protocol ISO 13485:2016 - Medical Device Quality Management Systems 6
Stoic Are any medical device companies using the 2011 FDA process validation guidance instead of GHTF/SG3/N99-10:2004? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
K Old medical devices -> 7.3.7. Design and development validation ISO 13485:2016 - Medical Device Quality Management Systems 1
S SOP for ISO 13485:2016 Quality related Software validation ISO 13485:2016 - Medical Device Quality Management Systems 9
Y Retrospective Validation - Class I device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3
M Production approval testing - Alternative ideas for Validation Reliability Analysis - Predictions, Testing and Standards 4
M Validation of two nearly identical products Other Medical Device Regulations World-Wide 5
J Requested Validation plan and reports Manufacturing and Related Processes 4
S Validation Records - Very young QMS Qualification and Validation (including 21 CFR Part 11) 2
M Test method validation - Is MSA (MSA1, MSA2, MSA3 and linearity) a good solution? Medical Device and FDA Regulations and Standards News 1
G Devices from IQ, OQ or PQ process to be used for verification, validation and summative? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
K ERP System Software Validation - ISO13485 2016 4.1.6 Design and Development of Products and Processes 8
W ASTM F1929 dye penetration test - Validation for in-house testing ISO 13485:2016 - Medical Device Quality Management Systems 13
Bev D Verification and Validation of Measurement Systems Misc. Quality Assurance and Business Systems Related Topics 0
Y Does Solidworks (2D/3D drafting modules) need validation? Other Medical Device and Orthopedic Related Topics 5
D Software validation in Medical Equipment Other Medical Device and Orthopedic Related Topics 20
K Validation of new machine (second machine of the same type) Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2

Similar threads

Top Bottom