Validation of A Procedure for Calibrating Dial Calipers using Traceable Gauge Blocks

Rustedamy

Starting to get Involved
#1
I have a procedure for calibrating dial calipers using traceable gauge blocks. Pretty straightforward yet assessor is requesting documents that indicate the procedure is validated. Traceability isn't sufficient? I have deviations & uncertainty. What am I not understanding about validation?
 
Elsmar Forum Sponsor

Jerry Eldred

Forum Moderator
Super Moderator
#2
Re: Validation of A Procedure

I'm not the ISO expert (more broad technical issues). So ISO folks, please correct me if needed.

My understanding that a procedure used in an ISO17025 context must be adequate. That is, there must be some basis that it correctly checks an instrument's specifications. If you use one developed by an OEM, military (GIDEP/MIDAS), ANSI, ASTM or other nationally or internationally recognized organization with their approval process, it is assumed that it was validated to assure it correctly and completely checked the instrument.

When you develop one yourself, you must somehow verify that it fully and correctly checks specifications.

For example, if verifying a high accuracy digital indicator with a 12 inch gage block, temperature is a significant source of error. If you wrote a procedure that did not include a proper method to deal with temperature implications, although the tolerance limits and other instructions may be correct, that missing detail (temperature correction) could invalidate the procedure. I don't imply that is a problem with yours, or that your procedure has any deficiencies that would invalidate it. Your procedure may be fully correct and complete. What I believe the assessor wants to know is what method you used to assure that it was complete and correct.

What constitutes adequate validation? I have numerous thoughts, but I'll leave that level of detail to more qualified folks.

One alternative may be to locate a published procedure for your mfr/model or it's equivalent. There are definitely GIDEP procedures to cover it (if you are eligible to participate). There may also be other acceptable published procedures.
 

DietCokeofEvil

Trusted Information Resource
#4
Re: Validation of A Procedure for Calibrating Dial Calipers using Traceable Gauge Blo

True enough. We cite the standard we are following in our written procedure. For calipers- I believe we cite a Navair standard, but I will have to double check that when I get to work.

Edited to add: Are you sure you can get away with only using gage blocks? How are you checking the ID jaws and the depth of stem etc?
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#5
Re: Validation of A Procedure for Calibrating Dial Calipers using Traceable Gauge Blo

I have a procedure for calibrating dial calipers using traceable gauge blocks. Pretty straightforward yet assessor is requesting documents that indicate the procedure is validated. Traceability isn't sufficient? I have deviations & uncertainty. What am I not understanding about validation?
First question is, did your procedure come from an accepted procedure (e.g., NA 17-20 series)? If so, reference that in your procedure, and depending on your AB you should be OK.

If this was developed in-house without referencing an accepted procedure then you must validate it under ANS/ISO/IEC 17025. The validation includes a plan to describe what the procedure is supposed to deliver, some number of actual calibration to obtain the information, and a report at the end matching results to expectation, and a statement that the procedure is fit for intended use.....now.....you must also calculate your measurement uncertainty as a part of the validation, and you should include basic checks such as the jaw and dirt checks (Light between closed jaws; run the calipers all the way out/in slowly and a sticky point indicate dirt.

Traceable gage blocks are a necessity, but remember the so-called NIST numbers by themselves are not valid for traceability.

Hope this helps.
 
J

jfgunn

#6
Re: Validation of A Procedure for Calibrating Dial Calipers using Traceable Gauge Blo

I think you have received some good responses thus far.

I recently had an L-A-B auditor tell me that NAVAIR and T.O. procedures still had to be validated as non-standard methods for ISO 17025. He was OK with ASME procedures being taken as is without the validation. To be honest, I did not really agree, but the solution was more simple than the arguement.

Since we had been using these procedures for several years the validation was pretty simple. I documented 10 examples of calibrations we had performed using the procedure and made a statement that it was suitable for use. Certainly the fact theat it was a procedure acceptable to the military also gives you some confidence in its suitability. I calucalted an uncertainty budget. I stapled it all together, put a nice "verfied" stamp on the procedure and put it in a file.

One more thing about calipers. There are actually 4 measurement for a caliper: OD, ID, Depth and the less know Step. Most procedures (including the one I use) ignore this function on the caliper.
 

Rustedamy

Starting to get Involved
#7
Re: Validation of A Procedure for Calibrating Dial Calipers using Traceable Gauge Blo

Thanks SO much, Hershal and jfgunn for your helpful posts! What a treasure to have found such a great website! :D
 
Thread starter Similar threads Forum Replies Date
N Validation procedure, Major NC CAP - Equipment Validation process is not effective ISO 13485:2016 - Medical Device Quality Management Systems 5
R ISO 13485 Software validation procedure and Quality Objectives Monitoring wanted Document Control Systems, Procedures, Forms and Templates 1
L ISO 17025 Para. 5.4.5 - Method (Calibration Procedure) Validation Requirements ISO 17025 related Discussions 2
M Chosing a procedure title - "Validation of equipment" or "Qualification of equipment" Document Control Systems, Procedures, Forms and Templates 8
J Sterilization Validation Procedure Needed? ISO 13485:2016 - Medical Device Quality Management Systems 3
M Looking for a simple procedure on section 7.5.2 for Validation of Processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
A Writing a Procedure for Validation of an Analytical Method General Measurement Device and Calibration Topics 5
A Should the FSMS have a Documented Generic Procedure for Verification and Validation Food Safety - ISO 22000, HACCP (21 CFR 120) 2
S Writing a Test Method Validation Procedure as well as Validation of Test Methods General Measurement Device and Calibration Topics 1
A Installation and Validation Procedure for IT Laboratory ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
C Software Validation Procedure per FDA requirement Qualification and Validation (including 21 CFR Part 11) 3
C Process Validation procedure for ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 5
C Do I have to establish a software validation procedure? Qualification and Validation (including 21 CFR Part 11) 11
T Oven Validation Protocol (procedure) needed General Measurement Device and Calibration Topics 19
V Validation of Processes for Production and Service Provision Procedure Document Control Systems, Procedures, Forms and Templates 8
B Software Validation Procedure Question - Complaint Handling Process software Software Quality Assurance 9
T ISO 17025 Validation Procedure - Need Example ISO 17025 related Discussions 11
Ajit Basrur Molding Validation procedure - New medical component ISO 13485:2016 - Medical Device Quality Management Systems 1
V Explain the process validation procedure in the manufacture of IVD products ISO 13485:2016 - Medical Device Quality Management Systems 1
T Searching for: Procedure for validation of dry measure Document Control Systems, Procedures, Forms and Templates 2
F Procedure for Software Validation and Control Quality Assurance and Compliance Software Tools and Solutions 4
T Anyone have an example of a calibration method validation procedure? General Measurement Device and Calibration Topics 4
Y SaMD Verification and Validation SOP and Plan IEC 62304 - Medical Device Software Life Cycle Processes 2
C Spreadsheet Validation and Release Qualification and Validation (including 21 CFR Part 11) 2
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
M Sterile packing validation tests to be performed and protocol Other Medical Device and Orthopedic Related Topics 1
S OEM full range calibrations vs Validation special test points Medical Device and FDA Regulations and Standards News 0
H Software Validation for FFS Packaging Machine Qualification and Validation (including 21 CFR Part 11) 1
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
M ERP / QMS related software standards for Validation IEC 62304 - Medical Device Software Life Cycle Processes 6
C ISO 19227 Validation Cost Other Medical Device Related Standards 2
A CNC Mill and Lathe - Validation Manufacturing and Related Processes 2
N Example for design and development planning,input,output,review,verification,validation and transfer Misc. Quality Assurance and Business Systems Related Topics 4
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
B Oracle Cloud ERP Validation during Quarterly Patch ISO 13485:2016 - Medical Device Quality Management Systems 1
D Software validation team Misc. Quality Assurance and Business Systems Related Topics 3
W LTPD, AQL, Ppk and Cpk validation sampling plan table Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
J Validation Sample Size for Tray Seal Qualification and Validation (including 21 CFR Part 11) 2
F AS9100 - Validation, FAIR's, ITAR and Sub-Contracting AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
R PCBA process validation Qualification and Validation (including 21 CFR Part 11) 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Sterilization validation after changing sterilization process provider Qualification and Validation (including 21 CFR Part 11) 3
B Sterilization Validation Plan Other Medical Device Related Standards 3
D Difference between Test Method Validation and Gage R&R Qualification and Validation (including 21 CFR Part 11) 18
T Laboratory Verification after validation ISO 17025 related Discussions 3
silentmonkey Rationalising the level of effort and depth of software validation based on risk ISO 13485:2016 - Medical Device Quality Management Systems 10
D Questions regarding process validation ISO 13485:2016 - Medical Device Quality Management Systems 6
Y We found out we have been using a equipment without validation for past 4 years Quality Manager and Management Related Issues 6
Z Is IQ necessary for laser marking validation? EU Medical Device Regulations 3
E 13485:2016, Sections 4.1.6, 7.5.6 and 7.6 - Validation of Software - Need some Advice please ISO 13485:2016 - Medical Device Quality Management Systems 3

Similar threads

Top Bottom