I'm a contract manufacturer that specializes in swiss machining orthopaedic implants/instruments. We are working on our first bone screw project that requires validation.
These screws will be swiss turned complete, ultrasonically cleaned, finished & laser marked, then, bagged with an IFU and sealed with a label (They don't require sterilization, they just need to be cleaned).
My customer wants us to validate our swiss machining process, as it relates to this type of screw and also validate our ultrasonic cleaning process. They have supplied me with the GHTF Process Validation Guidance document and have been working with us to develop our validation system, but, the guidelines and information are so generic.
I was wondering if anyone out there has devised validation processes that are specific to swiss machining and ultrasonic cleaning? If so, would you mind supplying some sample documentation?
I've got a handle on the IQ stuff, but I'm having trouble figuring out how to apply the OQ and PQ aspects. I'm thinking for the swiss machining of the bone screw, I'll have to come up with a validation plan that's specific to the manufacture of that specific type of part. But for ultrasonic cleaning, it seems like I could validate on a broad general basis and apply the results to a variety of different parts that we might manufacture and clean.
Any help would be greatly appreciated!
U
Umang Vidyarthi
Hi Alongtain,
:bigwave: Welcome to the cove. :bigwave:
I am not familiarised in this field, but experts on the cove will be able to help you out. In the meantime you may browse the 'Validation Times' issued by High Beam here: http://www.highbeam.com/Search.aspx?q=Validation+of+Bone+screw+manufacturing+process
Have a good day.
Umang :D
Ajit Basrur
Hi alongtain,
Very true - the GHTF document is a genric one but using the concepts mentioned in that document, you could draft something meaningful for you.
You could start with identifying all the equipment that are involved in the process and perform Equipment Qualification (IQ, OQ and PQ).
Once the equipment qualification is successful, draft a Process Validation protocol and validate all your processes. The GHTF document gives very good guidance on both qualifications and validations.
Pl come back if you have specific queries. :)
jscholen
The primary focus in an OQ is testing the limits of your manufacturing process. In the case of the screw machine, you want to establish the range of variable parameters that can be set on the machine, ie, screw feed, spindle speed, feed rate, etc. You will want to establish which combination of variable parameter work well together and what the range is for each parameter to produce product to specification. Your in essence trying to find the right and wrong combination of parameters to ensure you are always confident you will produce good screws. Once you establich a range of values that produce good screws, then you will execute a PQ.
In PQ, you want to establish that your process can maintain the middle of the range of each parameter you tested over the long haul...current Good Manufacturing Processes dictate 3 separate manufacturing orders....This is where you want to ensure that you can maintain good product between separate operations, ie, shift changes, weekend shutdowns, Operators, etc.
Primary focus of a Validation is ensure that there is a plan, procedures are in place for the equipment and people are trained AND everyone involved in the validation is aware and trained in the validation plan. Keep training records.
Also, have a provision for deviations in your initial validation plan, because stuff always happens that you didn't expect, so you want to be sure you provide for a way to address deviations...but you address them in a logical manner and the corrective actions makes sense or you will need to do it all over again.
Hope this helps, and I'm sure I have missed some aspect of validation because there is also the factor of risk management and how to ensure you are addressing potential risk in manufacturing that could affect the device at the point of use(human body)...this is best done with risk analysis tools such as FTA and FMEA.
Jeff
alongtain
Thanks alot, I'll check it out today and should be able to apply it to my situation.
alongtain
Thanks to all of you! I've gotten some great feedback and the cleaning validation document is a real help. :thanx:
rebecca.hess
Great info here! I am also working on a similar validation project (actually several!) and having issues with the sampling plan to use for this type of process.
I have the general Process PQ essentially complete except my customer who owns the product design is waffling on the sampling justification they would like to see used.
I have seen several references to a Bayes formula but am having trouble interpreting it as I am definetly not a statistician. I can easily plug the 90%/90% into a calculator and come up with the 22 piece sample but need to explain my sample size rational in the PQ.
Any help anyone can give on this would be greatly appreciated.
Y SaMD Verification and Validation SOP and Plan IEC 62304 - Medical Device Software Life Cycle Processes 2
C Spreadsheet Validation and Release Qualification and Validation (including 21 CFR Part 11) 2
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
M Sterile packing validation tests to be performed and protocol Other Medical Device and Orthopedic Related Topics 1
S OEM full range calibrations vs Validation special test points Medical Device and FDA Regulations and Standards News 0
H Software Validation for FFS Packaging Machine Qualification and Validation (including 21 CFR Part 11) 1
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
M ERP / QMS related software standards for Validation IEC 62304 - Medical Device Software Life Cycle Processes 6
C ISO 19227 Validation Cost Other Medical Device Related Standards 2
A CNC Mill and Lathe - Validation Manufacturing and Related Processes 2
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
B Oracle Cloud ERP Validation during Quarterly Patch ISO 13485:2016 - Medical Device Quality Management Systems 1
D Software validation team Misc. Quality Assurance and Business Systems Related Topics 3
J Validation Sample Size for Tray Seal Qualification and Validation (including 21 CFR Part 11) 3
F AS9100 - Validation, FAIR's, ITAR and Sub-Contracting AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
R PCBA process validation Qualification and Validation (including 21 CFR Part 11) 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Sterilization validation after changing sterilization process provider Qualification and Validation (including 21 CFR Part 11) 3
B Sterilization Validation Plan Other Medical Device Related Standards 3
T Laboratory Verification after validation ISO 17025 related Discussions 3
silentmonkey Rationalising the level of effort and depth of software validation based on risk ISO 13485:2016 - Medical Device Quality Management Systems 10
D Questions regarding process validation ISO 13485:2016 - Medical Device Quality Management Systems 6
Y We found out we have been using a equipment without validation for past 4 years Quality Manager and Management Related Issues 6
Z Is IQ necessary for laser marking validation? EU Medical Device Regulations 3
E 13485:2016, Sections 4.1.6, 7.5.6 and 7.6 - Validation of Software - Need some Advice please ISO 13485:2016 - Medical Device Quality Management Systems 3
A Validation of Forced Aeration Process ISO 13485:2016 - Medical Device Quality Management Systems 3
M API 4F/7K/8C Design Package Validation Oil and Gas Industry Standards and Regulations 2
I ISO 2233:2000 Question - Medical Device Shipping/Transportation Validation Other ISO and International Standards and European Regulations 1
B TMV - Selection of TM's for Validation ISO 13485:2016 - Medical Device Quality Management Systems 5
S Forced ServiceNow validation - No change in our current user and functional requirements IT (Information Technology) Service Management 4
P Human Factors / Usability validation in the time of COVID Human Factors and Ergonomics in Engineering 17
C Template for Excel Validation Reliability Analysis - Predictions, Testing and Standards 6
M IT validation for a paper based MD repair company QMS ISO 13485:2016 - Medical Device Quality Management Systems 6
P Unrealistic Packaging Validation Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 39
D Test summary report example for design validation wanted - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
M Is Validation of Plating Processes required and who is responsible? Qualification and Validation (including 21 CFR Part 11) 11
T ISO 13485 - Process validation at critical suppliers ISO 13485:2016 - Medical Device Quality Management Systems 7
K Software Validation for Measurement Tools used in Process Validation ISO 13485:2016 - Medical Device Quality Management Systems 2
Stoic Manual soldering processes - 100% verifiable, or always requiring validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
P Design verification driven by new equipment. How is this different than process validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Rees System Validation Qualification and Validation (including 21 CFR Part 11) 1
K PQ validation qualification - Asked to write a PQ protocol ISO 13485:2016 - Medical Device Quality Management Systems 6
Stoic Are any medical device companies using the 2011 FDA process validation guidance instead of GHTF/SG3/N99-10:2004? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
K Old medical devices -> 7.3.7. Design and development validation ISO 13485:2016 - Medical Device Quality Management Systems 1
S SOP for ISO 13485:2016 Quality related Software validation ISO 13485:2016 - Medical Device Quality Management Systems 9