Validation of Forced Aeration Process

akyt

Registered
#1
Hello people,

My company has a forced aeration process after the sterilization cycle, which considerably reduces the time to remove EO residue from products. The process is carried out in a dedicated sterilization chamber, in which vacuum pulses and inert gas injection are responsible for aerating the product.

We are needing to validate this process and some doubts have arisen. The ISO 11135: 2014 standard talks about evaluating the product and packaging performance in the worst process conditions, and I understand that this aeration is stressful for the product. So, how can I proceed with the validation?

I need to ensure that forced aeration is effective in removing waste, and at the same time does not affect the product performance and the packaging integrity. Any ideas on what tests, conditions and sampling should I use?

Thank you very much in advance.
 
Elsmar Forum Sponsor

chris1price

Trusted Information Resource
#2
Hi, without additional knowledge of your products, identifying the tests, etc is difficult. However, what you are describing is standard practice.

Sterilisation and aeration can both cause damage to products. Significant vacuum is applied and the resultant outgassing can inflate and damage, even burst, packaging. Remember the product may not be at room temperature, possible 60C or higher. PVC balloons are easily damaged at these conditions. You should consider performing all design verification tests after sterilisation (and aeration), and after your shelf-life testing. You should clearly document the rationale for not performing any tests after exposure.
 

akyt

Registered
#3
Thank you very much for your contribution.

Our sterilization and forced aeration process is carried out at 55°C. I have some specific tests to evaluate the product's performance and I can apply them. Likewise, I will also apply tests to the packaging. Do you think that only the sterile barrier system seal integrity test (ASTM F1929) is sufficient?
Should I do these same tests after exposure to aging?

Another doubt, in an environmental aeration the ISO 10993-7 standard requires 3 cycles. In this case, how many forced aeration cycles should I consider for evaluation? Is one cycle sufficient?
 

chris1price

Trusted Information Resource
#4
Its up to you to decide the best test method depending on the packaging configuration. I would recommend using a quantitative test (peel test, or burst test) as well as the dye penetration test. Whichever, you need to validate the method and establish failure criteria before you start the testing. Take a look at ISO11607 for a list of test methods.

Yes, you should repeat the seal tests and product tests after aging. Doing it immediately after sterilisation will give you the confidence to progress to ageing, after which it should be repeated in full.

Per ISO10993-7, you are doing the three cycles for testing of residuals. It doesn't matter whether the aeration is forced or ambient, you still need to establish the aeration time.
 
Thread starter Similar threads Forum Replies Date
S Forced ServiceNow validation - No change in our current user and functional requirements IT (Information Technology) Service Management 4
C Spreadsheet Validation and Release Qualification and Validation (including 21 CFR Part 11) 2
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
M Sterile packing validation tests to be performed and protocol Other Medical Device and Orthopedic Related Topics 1
S OEM full range calibrations vs Validation special test points Medical Device and FDA Regulations and Standards News 0
H Software Validation for FFS Packaging Machine Qualification and Validation (including 21 CFR Part 11) 1
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
M ERP / QMS related software standards for Validation IEC 62304 - Medical Device Software Life Cycle Processes 6
C ISO 19227 Validation Cost Other Medical Device Related Standards 2
A CNC Mill and Lathe - Validation Manufacturing and Related Processes 2
N Example for design and development planning,input,output,review,verification,validation and transfer Misc. Quality Assurance and Business Systems Related Topics 4
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
B Oracle Cloud ERP Validation during Quarterly Patch ISO 13485:2016 - Medical Device Quality Management Systems 1
D Software validation team Misc. Quality Assurance and Business Systems Related Topics 3
W LTPD, AQL, Ppk and Cpk validation sampling plan table Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
J Validation Sample Size for Tray Seal Qualification and Validation (including 21 CFR Part 11) 1
F AS9100 - Validation, FAIR's, ITAR and Sub-Contracting AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
R PCBA process validation Qualification and Validation (including 21 CFR Part 11) 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Sterilization validation after changing sterilization process provider Qualification and Validation (including 21 CFR Part 11) 3
B Sterilization Validation Plan Other Medical Device Related Standards 3
D Difference between Test Method Validation and Gage R&R Qualification and Validation (including 21 CFR Part 11) 18
T Laboratory Verification after validation ISO 17025 related Discussions 3
silentmonkey Rationalising the level of effort and depth of software validation based on risk ISO 13485:2016 - Medical Device Quality Management Systems 10
D Questions regarding process validation ISO 13485:2016 - Medical Device Quality Management Systems 6
Y We found out we have been using a equipment without validation for past 4 years Quality Manager and Management Related Issues 6
Z Is IQ necessary for laser marking validation? EU Medical Device Regulations 3
E 13485:2016, Sections 4.1.6, 7.5.6 and 7.6 - Validation of Software - Need some Advice please ISO 13485:2016 - Medical Device Quality Management Systems 3
E Mentor for Test Method Validation (TMV) Design and Development of Products and Processes 2
M API 4F/7K/8C Design Package Validation Oil and Gas Industry Standards and Regulations 2
I ISO 2233:2000 Question - Medical Device Shipping/Transportation Validation Other ISO and International Standards and European Regulations 1
T Annual Validation as a detection mode on a PFMEA? FMEA and Control Plans 5
B TMV - Selection of TM's for Validation ISO 13485:2016 - Medical Device Quality Management Systems 5
P Human Factors / Usability validation in the time of COVID Human Factors and Ergonomics in Engineering 17
C Template for Excel Validation Reliability Analysis - Predictions, Testing and Standards 6
M IT validation for a paper based MD repair company QMS ISO 13485:2016 - Medical Device Quality Management Systems 6
P Unrealistic Packaging Validation Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 26
D Test summary report example for design validation wanted - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
M Is Validation of Plating Processes required and who is responsible? Qualification and Validation (including 21 CFR Part 11) 11
T ISO 13485 - Process validation at critical suppliers ISO 13485:2016 - Medical Device Quality Management Systems 7
K Software Validation for Measurement Tools used in Process Validation ISO 13485:2016 - Medical Device Quality Management Systems 2
Stoic Manual soldering processes - 100% verifiable, or always requiring validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
P Design verification driven by new equipment. How is this different than process validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Rees System Validation Qualification and Validation (including 21 CFR Part 11) 1
K PQ validation qualification - Asked to write a PQ protocol ISO 13485:2016 - Medical Device Quality Management Systems 6
Stoic Are any medical device companies using the 2011 FDA process validation guidance instead of GHTF/SG3/N99-10:2004? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
K Old medical devices -> 7.3.7. Design and development validation ISO 13485:2016 - Medical Device Quality Management Systems 1
S SOP for ISO 13485:2016 Quality related Software validation ISO 13485:2016 - Medical Device Quality Management Systems 9
Y Retrospective Validation - Class I device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3

Similar threads

Top Bottom