Validation of Medical Device Hardware containing Software - How many to Validate

racglobal

Involved In Discussions
#1
Hi everyone,

I just want to clarify the concept of validation. If a company manufactures an electromedical device (hardware containing software), how many units of this are we required to validate to show the auditor we can manufacture this device? In the past, a company I worked at arrived at an arbitrary # (e.g. 5 units) and I wasn't sure why. Because of the large scale of the medical equipment, we cannot afford to validate more than a couple. Of all the QMS processes, validation is still the part I am a bit shaky on. Any advice would be appreciated.
 
Elsmar Forum Sponsor

Bev D

Heretical Statistician
Leader
Super Moderator
#2
The sample size isn’t so much about how many units of the device, but how many uses of the device. While software is deterministic, it is dependent on input conditions and these will change from use to use as will the performance of the other electromechanical components. You may be able to validate with only one device if you vary the use conditions. The number of uses can be determined by:
- knowing the acceptable failure rate(s) and using one of the formulas for estimating the failure rate. You would need to know the amount of precision you want in the estimate and the alpha risk and potentially the bet risk. This would typically entail random variation of the use conditions. It does lead to rather large sample sizes tho as low failure rates require larger sample sizes.
- a second approach is to use directed testing under worst case conditions with continuous data response to look for margin. This can substantially reduce your sample size as you don’t have to test all conditions. I use 3-5 ‘uses’ or runs as this gets me a good idea of the extreme of the perfromance range. Of course discussion with your reviewer and/or the reviewing statistician is always good practice. Getting agreement on the study design ahead of time can save a lot of time and frustration...
 
Thread starter Similar threads Forum Replies Date
C Medical Device Gamma Irradiation Validation per VDmax25 (ISO 11137) Qualification and Validation (including 21 CFR Part 11) 1
I ISO 2233:2000 Question - Medical Device Shipping/Transportation Validation Other ISO and International Standards and European Regulations 1
Stoic Are any medical device companies using the 2011 FDA process validation guidance instead of GHTF/SG3/N99-10:2004? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
I Medical device Validation/Verification template ISO 13485:2016 - Medical Device Quality Management Systems 1
Z Class 2 medical device - Process Validation - Test sterile or non-sterile units? Qualification and Validation (including 21 CFR Part 11) 9
M Medical Device Process validation, Validation of excel spreadsheets used for process Other Medical Device and Orthopedic Related Topics 4
V Tyvek packaging for Medical device process validation Qualification and Validation (including 21 CFR Part 11) 2
A SOP for software validation of software in medical device IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 5
I Implantable medical device cleaning validation Other Medical Device Related Standards 3
R Determining Sample Size for Medical Device Component Validation Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
M Example Medical Device Validation Plan and Report wanted CE Marking (Conformité Européene) / CB Scheme 4
N China's CFDA - Medical Device Validation Requirements China Medical Device Regulations 8
P Medical Device Validation Protocols and Reports - Acceptance Criteria Other Medical Device and Orthopedic Related Topics 3
T Validation Plan for a new Class II Medical Device (Critical Care Medical Ventilator) IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A Inspection Protocol Validation for a Medical Device Customer Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
D Medical Device Software Tool Validation - Compilers! IEC 62304 - Medical Device Software Life Cycle Processes 7
N When is Medical Device Software Validation required? ISO 13485:2016 - Medical Device Quality Management Systems 6
R Medical Device Packaging Label Validation Other US Medical Device Regulations 8
JoCam Validation of a change to a CE Marked Medical Device EU Medical Device Regulations 4
A FDA Medical Device Reprocessing Guidance - Validation US Food and Drug Administration (FDA) 6
P Validation of Class I Metal Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 3
W Process Validation Training - Custom made Medical Device Design and Development of Products and Processes 4
J Mobile Medical Device App Validation Guidance vs. GPSV? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
Z IEC 62366 and Medical Device Usability Validation IEC 62366 - Medical Device Usability Engineering 10
E FDA Requirements for Implantable Medical Device Software Validation ISO 13485:2016 - Medical Device Quality Management Systems 5
C Thoughts on validation of Legacy Systems for Medical Device Software EU Medical Device Regulations 2
A ClassIIa Medical Device Software Validation Report Format and Content IEC 62304 - Medical Device Software Life Cycle Processes 5
A Medical Device Software Validation Guidance Software Quality Assurance 3
B Validation of a Pressure Decay Leak Tester for a Medical Device Design and Development of Products and Processes 2
K Medical Device Design Validation vs Clinical Evaluation ISO 13485:2016 - Medical Device Quality Management Systems 5
I Medical Device Validation and Pharmaceutical Validation differences? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
T Medical Device Software Verification & Validation for Client Specific Configurations Design and Development of Products and Processes 2
S Medical Device Mechanical Test Method Validation Design and Development of Products and Processes 8
suildur Sterilization Validation of a Unique Medical Device Other Medical Device Related Standards 5
L Validation Plan Requirements when moving to a new Plant - Medical Device Industry Qualification and Validation (including 21 CFR Part 11) 19
K Re-validation of Out Sourced Medical Device Sterilization Process ISO 13485:2016 - Medical Device Quality Management Systems 3
A How to do Medical Device Design Validation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
S Medical Device Software (Firmware) Validation with using IEC62304 IEC 62304 - Medical Device Software Life Cycle Processes 21
A Software Validation Requirements for Class I Active Medical Device EU Medical Device Regulations 4
D Medical Device Design Verification and Validation Differences 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 16
A Software Validation for Class I - Manufacture of a Part for a Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 9
R Medical Device Inspection Sheets - Excel Spreadsheet Validation US Food and Drug Administration (FDA) 2
T Medical Device Manufacturer Document Control Computer System Validation Qualification and Validation (including 21 CFR Part 11) 4
S 21 CFR 820.75 Process Validation for Medical Device Software 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
T New Product Validation Plan example or template for a Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
S Medical Device Software Verification and Validation Results - What is necessary? Software Quality Assurance 4
R Medical Device SPC Software Validation Statistical Analysis Tools, Techniques and SPC 6
Q Medical Device Field Service Performance Inspection Process Validation Qualification and Validation (including 21 CFR Part 11) 1
S Medical Device Design Verification and Validation - info Other Medical Device and Orthopedic Related Topics 1
E ISO 62304 and Validation of Medical Device Software Tools IEC 62304 - Medical Device Software Life Cycle Processes 17

Similar threads

Top Bottom