Validation of outsourced processes - Calibration - How do we validate that?

R

Ron Durham

#1
Our company had a recert audit this month by Orion.
Our auditor stated that outside calibration is an outsourced process and how do we validate that process. That never has been an issue in our previous audits. One of our calbration companies is ISO 9001:2000 certified and 2 of them are ISO 9001:2000 compliant. Is this necessary.

Thanks, Ron
 
Elsmar Forum Sponsor

RoxaneB

Super Moderator
Super Moderator
#2
Ron Durham said:
Our company had a recert audit this month by Orion.
Our auditor stated that outside calibration is an outsourced process and how do we validate that process. That never has been an issue in our previous audits. One of our calbration companies is ISO 9001:2000 certified and 2 of them are ISO 9001:2000 compliant. Is this necessary.

Thanks, Ron
That's never been an issue for us, either. We do, however, have our outside calibration/validation companies on our Approved Vendor List and they subject the rules we've established for that process. How does your company address them?
 
R

ralphsulser

#3
Are any A2LA certified? We considered the outside calibration services as suppliers and they are A2LA certified. We listed them on our Approved Suppliers list. We have passed 2 TS16949 audits and this was never mentioned as anything other than how we document the service.
 
K

Kevin H

#4
Ron, why would you choose an outside calibration service cerified or compliant to ISO 9001, rather than accredited to ISO 17025? The 17025 standard would seem to be a better choice when selecting calibration services. (Note - I "live" in an /ISO/TS system, so we do not really have much choice.)

I'm a bit surprised that your auditor wrote the finding against outsourced processes, rather than 7.6: Control of monitoring & measuring devices. They may have felt that you inadequately defined the requirements for your calibration process, or have seen something that raised an issue. N.B. that is just a WAG. The best bet, would be to discuss with the auditor and try to understand his underlying reasoning.
 
R

Ron Durham

#5
Outsourced processes

Roxane:

We don't list our calibration houses on our approved vendors list. We have been using them for over 10 years and have not ever had an issue with them. What are the rules you established for that process?

Ralph:
None of the calibration companies are certified to A2LA. The only thing we have done outside is our cmm's, electronic height gages, ring gages and thread gages. 95% of our measuring tools are calibrated in house.

One more question on outsourced processes. We send out a lot of parts for anodizing. This is for cosmetic purposes only. We use companies that have been surveyed and recommended by our customers. I stated in our manual that I validate the process by using certs and customers recommended by our customers. That was ok by the auditor but our Purchasing Manager said that if we have a customer that doesn't care who we use then we use someone who has been approved by one of our other customers(Boeing, Storage Tek).
Our auditor said we needed to validate that process by use of a mail in survey or onsite audit. This seems like a lot of work for cosmetic parts. After all we know what our customers require and they have never had an issue with any anodized parts that we send them.
Thanks, Ron
 

Al Rosen

Staff member
Super Moderator
#6
Ron Durham said:
Roxane:

We don't list our calibration houses on our approved vendors list. We have been using them for over 10 years and have not ever had an issue with them. What are the rules you established for that process?

Ralph:
None of the calibration companies are certified to A2LA. The only thing we have done outside is our cmm's, electronic height gages, ring gages and thread gages. 95% of our measuring tools are calibrated in house.

One more question on outsourced processes. We send out a lot of parts for anodizing. This is for cosmetic purposes only. We use companies that have been surveyed and recommended by our customers. I stated in our manual that I validate the process by using certs and customers recommended by our customers. That was ok by the auditor but our Purchasing Manager said that if we have a customer that doesn't care who we use then we use someone who has been approved by one of our other customers(Boeing, Storage Tek).
Our auditor said we needed to validate that process by use of a mail in survey or onsite audit. This seems like a lot of work for cosmetic parts. After all we know what our customers require and they have never had an issue with any anodized parts that we send them.
Thanks, Ron
I'm confused. What you have described, I would not consider validation but control of suppliers. Is that what your auditor was concerned about? What clause did the auditor cite?
 
V

vanputten

#7
Confused by thread

Hello All:

I think the answer to Ron's question may have nothing to do with ISO 17025, 9001, or accredditation.

You asked another organization to perform a process for you. You outsoruced the process. Did you compare what you asked them to do with what they did? Compare the inputs you gave the service provider with the output they gave back to you.

The controls you place over an outsourced process may include a requirement for registration or accreditation. But I am not so sure it has to.

Regards,

Dirk
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#8
Ron Durham said:
None of the calibration companies are certified to A2LA.
No laboratory can be certified to A2LA.....a laboratory CAN e accredited BY A2LA.....

Remember A2LA is one, but only one, of several recognized accrediting bodies operating in North America. The others which are also internationally recognized and EQUALLY acceptable include IAS, NVLAP, SCC/CLAS (Canada), and EMA (Mexico). Nationally recognized bodies include IAS, NVLAP, L-A-B, and ACLASS. Within the U.S., either of these bodies is acceptable.

If you cal provider(s) become accredited by either IAS, A2LA, NVLAP, L-A-B, ACLASS, SCC/CLAS, or EMA, you should have no problem.

Hershal
 
R

ralphsulser

#9
Hershal said:
No laboratory can be certified to A2LA.....a laboratory CAN e accredited BY A2LA.....
Hershal
But they can have a Certificate to show they are Accredited to A2LA and we have those on file.:D
 
C

Craig H.

#10
Ron,

An auditor recommended a mail-in survey? I have a pet peeve about those things. The folks who will take the time to fill one out honestly are the ones you should worry about the least, generally speaking. Fly-by-nighters will send you smoke. How to tell the difference?

We once used mail-in surveys, I hate to admit. The highest self-scorer started having problems, so I/we did a site audit. What a joke that was.

I'd say that you can make the criteria for your system. Why not decide which materials and outsourced work are critical, and put them on a list?

We then decided that an ISO 9001 certificate or a site audit were necessary UNLESS we had a good track record with the supplier. Its in our purchasing procedure.
 
Thread starter Similar threads Forum Replies Date
S Validation Of Outsourced Special Processes under AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 14
T Outsourced Manufacturing - Do we or our manufacturer has to perform the validation? ISO 13485:2016 - Medical Device Quality Management Systems 6
C Spreadsheet Validation and Release Qualification and Validation (including 21 CFR Part 11) 2
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
M Sterile packing validation tests to be performed and protocol Other Medical Device and Orthopedic Related Topics 1
S OEM full range calibrations vs Validation special test points Medical Device and FDA Regulations and Standards News 0
H Software Validation for FFS Packaging Machine Qualification and Validation (including 21 CFR Part 11) 1
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
M ERP / QMS related software standards for Validation IEC 62304 - Medical Device Software Life Cycle Processes 6
C ISO 19227 Validation Cost Other Medical Device Related Standards 2
A CNC Mill and Lathe - Validation Manufacturing and Related Processes 2
N Example for design and development planning,input,output,review,verification,validation and transfer Misc. Quality Assurance and Business Systems Related Topics 4
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
B Oracle Cloud ERP Validation during Quarterly Patch ISO 13485:2016 - Medical Device Quality Management Systems 1
D Software validation team Misc. Quality Assurance and Business Systems Related Topics 3
W LTPD, AQL, Ppk and Cpk validation sampling plan table Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
J Validation Sample Size for Tray Seal Qualification and Validation (including 21 CFR Part 11) 1
F AS9100 - Validation, FAIR's, ITAR and Sub-Contracting AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
R PCBA process validation Qualification and Validation (including 21 CFR Part 11) 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Sterilization validation after changing sterilization process provider Qualification and Validation (including 21 CFR Part 11) 3
B Sterilization Validation Plan Other Medical Device Related Standards 3
D Difference between Test Method Validation and Gage R&R Qualification and Validation (including 21 CFR Part 11) 18
T Laboratory Verification after validation ISO 17025 related Discussions 3
silentmonkey Rationalising the level of effort and depth of software validation based on risk ISO 13485:2016 - Medical Device Quality Management Systems 10
D Questions regarding process validation ISO 13485:2016 - Medical Device Quality Management Systems 6
Y We found out we have been using a equipment without validation for past 4 years Quality Manager and Management Related Issues 6
Z Is IQ necessary for laser marking validation? EU Medical Device Regulations 3
E 13485:2016, Sections 4.1.6, 7.5.6 and 7.6 - Validation of Software - Need some Advice please ISO 13485:2016 - Medical Device Quality Management Systems 3
A Validation of Forced Aeration Process ISO 13485:2016 - Medical Device Quality Management Systems 3
E Mentor for Test Method Validation (TMV) Design and Development of Products and Processes 2
M API 4F/7K/8C Design Package Validation Oil and Gas Industry Standards and Regulations 2
I ISO 2233:2000 Question - Medical Device Shipping/Transportation Validation Other ISO and International Standards and European Regulations 1
T Annual Validation as a detection mode on a PFMEA? FMEA and Control Plans 5
B TMV - Selection of TM's for Validation ISO 13485:2016 - Medical Device Quality Management Systems 5
S Forced ServiceNow validation - No change in our current user and functional requirements IT (Information Technology) Service Management 4
P Human Factors / Usability validation in the time of COVID Human Factors and Ergonomics in Engineering 17
C Template for Excel Validation Reliability Analysis - Predictions, Testing and Standards 6
M IT validation for a paper based MD repair company QMS ISO 13485:2016 - Medical Device Quality Management Systems 6
P Unrealistic Packaging Validation Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 37
D Test summary report example for design validation wanted - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
M Is Validation of Plating Processes required and who is responsible? Qualification and Validation (including 21 CFR Part 11) 11
T ISO 13485 - Process validation at critical suppliers ISO 13485:2016 - Medical Device Quality Management Systems 7
K Software Validation for Measurement Tools used in Process Validation ISO 13485:2016 - Medical Device Quality Management Systems 2
Stoic Manual soldering processes - 100% verifiable, or always requiring validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
P Design verification driven by new equipment. How is this different than process validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Rees System Validation Qualification and Validation (including 21 CFR Part 11) 1
K PQ validation qualification - Asked to write a PQ protocol ISO 13485:2016 - Medical Device Quality Management Systems 6
Stoic Are any medical device companies using the 2011 FDA process validation guidance instead of GHTF/SG3/N99-10:2004? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
K Old medical devices -> 7.3.7. Design and development validation ISO 13485:2016 - Medical Device Quality Management Systems 1

Similar threads

Top Bottom