Validation of processes for production for Test Equipment

A

ariell3

#1
I have a customer who issued a major NC against our test equipment for not meeting the requirements of CFR 21 Sec. 820.75. We are a silicone manufacturer and use a Rheometer to measure product characteristics. These characteristics are neither toleranced, nor are they reported on the COA. The customer claims there is no protocol that outlines the acceptance criteria or a final report the details the results and conclusions.

This test equipment is calibrated by the manufacturer according to ASTM D5289 (Standard Test Method for Rubber Property - Vulcanizing Using Rotorless Cure Meter) as well as ISO 6502 (ibid.) The calibration standards used are NIST traceable and the ASTM standard addresses the correct preparation of the test samples and the precision and bias associated with the testing process. Additionally, there are other ASTM documents outlining the methodology and how the gage performance was calculated. This is a destructive test and a sample can only be tested on time.

I feel this NC is invalid and coming from TS the PPAP manual indicates that gage MSA studies are not required when standardized tests are used. I'm I correct in feeling this also applies to GMP?

Any thoughts?:confused:

:thanks:

Alex
 
Elsmar Forum Sponsor

Marc

Fully vaccinated are you?
Staff member
Admin
#2
Re: Validation of processes for production for Test Eqiupment

Can anyone help with this one?
 

BradM

Staff member
Admin
#3
Alex, I guess I'm still a bit confused at exactly what the customer's complaint is. Is it that there is no validation on the equipment, or the existing validation is insufficient?

Do you have anything in writing where you performed a qualification on the equipment?

Does your organization have a qualification requirement for all equipment, and this equipment was not qualified under that requirement?

I guess I'm trying to gauge where this is. I do think you will need something documented as a qualification on the equipment. However, not sure if the customer is looking for "thickness" :tg: of the qualification, and not it's effectiveness.
 
V

Vicman

#4
Hi, Ariell3
I think what your customer wants is something that is missing.
My understanding of a protocol is – a predefined written procedural method in the design and implementation of experiments.
Normally these protocols are written procedures in order to standardize a laboratory test method to ensure successful “replication” of the results in the same laboratory or others.
Procedures that include detail list of equipment and instruments, sometimes safety precautions and how to calculate the results and its interpretation on how to exclude some results if something undesirable happens like:
Not following calibration procedure, high ambient temperature, high humidity, low barometric pressure, turbulence flow instead of laminar flow that could affect the test results.
You may probably want to include the parameters and conditions in your final report to make the results more reliable.

Vicman
 

BradM

Staff member
Admin
#5
Hi, Ariell3
I think what your customer wants is something that is missing.
My understanding of a protocol is – a predefined written procedural method in the design and implementation of experiments.
Normally these protocols are written procedures in order to standardize a laboratory test method to ensure successful “replication” of the results in the same laboratory or others.
Procedures that include detail list of equipment and instruments, sometimes safety precautions and how to calculate the results and its interpretation on how to exclude some results if something undesirable happens like:
Not following calibration procedure, high ambient temperature, high humidity, low barometric pressure, turbulence flow instead of laminar flow that could affect the test results.
You may probably want to include the parameters and conditions in your final report to make the results more reliable.

Vicman
Good writeup, Vicman. I concur with your comments, here. Hence the reason I was interested in finding out if there is any protocol, a bad protocol, etc.

If this equipment has been in use for a while, I would approach the validation differently than if it is brand new equipment. However, there needs to be some documentation on the equipment, and future scheduled requalifications and such.
 
A

ariell3

#7
Thanks for the feedback! We will look into the way our test process is validated moving forward. However, in this case we just discovered the customer signed-off on the manfuacturing and testing process back in 2003. Which makes us wonder why the validation data was sufficient back then and not now?

This is a standard ASTM test for elastomeric materials and we use the data only for internal adjustments to the product. The results are recorded and not generally shared with the customer since they perform similar tests upon receipt of the material.

Thanks,

Alex
 

BradM

Staff member
Admin
#8
Thanks for the feedback! We will look into the way our test process is validated moving forward. However, in this case we just discovered the customer signed-off on the manfuacturing and testing process back in 2003. Which makes us wonder why the validation data was sufficient back then and not now?

This is a standard ASTM test for elastomeric materials and we use the data only for internal adjustments to the product. The results are recorded and not generally shared with the customer since they perform similar tests upon receipt of the material.

Thanks,

Alex
Thanks for the followup, Alex.

In this case, I am still at a loss as to exactly what the customer is having an issue with. What exactly are they saying is not appropriate; not right; unacceptable, etc.?
 
J

Juan Dude

#9
I have a customer who issued a major NC against our test equipment for not meeting the requirements of CFR 21 Sec. 820.75. We are a silicone manufacturer and use a Rheometer to measure product characteristics. These characteristics are neither toleranced, nor are they reported on the COA. The customer claims there is no protocol that outlines the acceptance criteria or a final report the details the results and conclusions.

This test equipment is calibrated by the manufacturer according to ASTM D5289 (Standard Test Method for Rubber Property - Vulcanizing Using Rotorless Cure Meter) as well as ISO 6502 (ibid.) The calibration standards used are NIST traceable and the ASTM standard addresses the correct preparation of the test samples and the precision and bias associated with the testing process. Additionally, there are other ASTM documents outlining the methodology and how the gage performance was calculated. This is a destructive test and a sample can only be tested on time.

I feel this NC is invalid and coming from TS the PPAP manual indicates that gage MSA studies are not required when standardized tests are used. I'm I correct in feeling this also applies to GMP?

Any thoughts?:confused:

:thanks:

Alex
You mention your NC is against 820.75 which deals with Process validation. Process Validation deals with Validating your Process to be capable of constantly producing product that meets all requirements. Even though your equipment may meet all calibration requirements, this does not mean that your process is validated, which is what your NC seems to be against.
 
M

MIREGMGR

#10
Per your explanation, it sounds as if your product has a significant potential process variability that your production people "know how to" control via this equipment and test method. "No tolerances" means the process is not formally controlled.

Per current FDA interpretations, your customer has no choice but to require you to at least have formally controlled processes. If your materials potentially are significant to their products' safety and effectiveness, they probably must require you to validate their processing.

As to the 2003 signoff, they have no choice as to how they proceed on a forward-going basis, so whatever happened six years ago is irrelevant.
 
Last edited by a moderator:
Thread starter Similar threads Forum Replies Date
D 7.5.2 Validation of Processes for Production and Service Provision AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
L Validation Processes of Production & Servicing - API Q1 clause 5.7.1.5 Oil and Gas Industry Standards and Regulations 2
A 7.5.2.1 Validation of Processes for Production and Service Provision IATF 16949 - Automotive Quality Systems Standard 7
D 7.5.2 Validation of Processes for Production and Service Provision - Distributor AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
Y 7.5.2 Validation of Processes for Production and Service Provision Discussions ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
H Excluding 7.5.2 - Validation of Processes for Production and Service Provision ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
S 7.5.2 Validation of Processes for Production and Service Provision AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 56
B Validation of Processes for Production and Service Provision - 7.5.2 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
A Understanding AS9100 Clause 7.5.2 Validation of Processes for Production and Service AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
B Validation of processes for production & service provision - Service industry ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
V Validation of Processes for Production and Service Provision Procedure Document Control Systems, Procedures, Forms and Templates 8
R Validation of Processes for Production and Service Provision - Clause 7.5.2 IATF 16949 - Automotive Quality Systems Standard 4
R Validation of Processes for Production - How are doing it? Clause 7.5.2 Reliability Analysis - Predictions, Testing and Standards 25
M Examples for Clause 7.5.2 - Validation of processes for production and service ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
S Exclusion from Validation of Processes for Production and Service Provision? Service Industry Specific Topics 17
N Validation of special processes - thread closed Oil and Gas Industry Standards and Regulations 3
M Is Validation of Plating Processes required and who is responsible? Qualification and Validation (including 21 CFR Part 11) 11
Stoic Manual soldering processes - 100% verifiable, or always requiring validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
T ISO 13485:2016 - Processes exempt from process validation ISO 13485:2016 - Medical Device Quality Management Systems 12
J ISO 13485 - 7.5.2 cleanliness , 6.4.2 validation of special processes, and 6.4.2 contamination ability to be exempt? ISO 13485:2016 - Medical Device Quality Management Systems 5
GoSpeedRacer Validation of Processes 7.5.6 - Do cosmetic welds need process validation? ISO 13485:2016 - Medical Device Quality Management Systems 3
P Validation of Processes - What is meant by 7.5.6 (a) - Defined Criteria for Review ISO 13485:2016 - Medical Device Quality Management Systems 7
S Re-validation NCR - In which cases re-validation of processes is required Oil and Gas Industry Standards and Regulations 6
M AS9110C Clause 8.5.1.2 Validation and Control of Special Processes for Repair Station AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
W AS9100D Clause 8.5.1.2 Validation and Control of Special Processes AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 37
I Bracketing for Validation - Processes that share some unitary operations Qualification and Validation (including 21 CFR Part 11) 4
J ISO 9001 Clause 7.5.2 Validation of Processes - How to comply? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 32
M Looking for a simple procedure on section 7.5.2 for Validation of Processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
S ISO/TS 16949 - 7.5.2 - Validation of Processes And 7.5.2.1 All Processes IATF 16949 - Automotive Quality Systems Standard 1
E What are Validation Requirements for Vendor (Supplier) Processes Supplier Quality Assurance and other Supplier Issues 5
K Validation of Processes does not seem to apply - Scientific Research Company Design and Development of Products and Processes 1
F External Storage of Data - ISO 9001 Clause 7.5.2 Validation of Processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
S Validation of Special Processes - Are Casting and Forging Special Processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
N Validation of Special Processes - N/A of ISO 13485 Clause 7.5.2 ISO 13485:2016 - Medical Device Quality Management Systems 12
M Validation of Processes - Is Cl 7.5.2 applicable to bricktile firing in an oven? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
L Outsourcing "passivation" - Clause 7.5.2 Validation of Processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
O Format for Clause 7.5.2 (Validation of Processes) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
T Validation of processes 7.5.2 - We are a label printer ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
S Validation Of Outsourced Special Processes under AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 14
V Validation & which processes are called special processes & why not all the processes Manufacturing and Related Processes 9
C Equipment, Software and Processes Validation for Medical Devices - 21CFR 820.75 (a) ISO 13485:2016 - Medical Device Quality Management Systems 35
L Validation of Processes - Plastic Injection Moulding - ISO 9001 Clause 7.5.2 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
K Validation of Processes - How do I address TS 16949 Sec. 7.5.2.1 IATF 16949 - Automotive Quality Systems Standard 30
L Validation of Processes - Resulting Output can not be Verified - Exclusion to 7.5.2 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 24
R Validation of outsourced processes - Calibration - How do we validate that? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
M Software Validation Processes - Telecommunications software Software Quality Assurance 10
A Validation of Processes - How to meet requirement of TS 16949 para. 7.5.2.1? IATF 16949 - Automotive Quality Systems Standard 12
F Process Validation of 'Special Processes' - How do I? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
J 7.5.2.1 Validation of processes - supplemental requirement IATF 16949 - Automotive Quality Systems Standard 11
S How to cope with the validation of processes - Services Service Industry Specific Topics 6